Ew, yeah, it doesn't define probes, and it doesn't use LISNs for injection and measurement (at least most of the time, at a glance). I guess you depend on the probes being within calibration. So, Pearson and Tektronix (or other suitable) AC and DC current clamps are called for.
Pearson CTs aren't especially remarkable, they're just well made and calibrated. The key is a toroidal winding with as few turns as possible -- 100 or less, say, so the resonant frequency is high. Toroidal resonant modes are mostly shunted away, by enclosing the winding in a metal case, a torus that is slitted in the plane (but only the inner ring). They also use nanocrystalline core material, to extend LF bandwidth.
You can do much the same by winding a single layer on a high-mu ferrite toroid, and wrapping that with foil tape so that the outside is a solid conductive layer, but a gap is left in the middle. Connect the tape to the winding so it's grounded. Ideally you'd have a balanced output (so, the tape would be grounded to a center tap, and the winding connects to shielded twisted pair cable), but that sucks, so you'd probably just ground one end of the winding instead, and use coax. (This suggests possible improvements, like a balun, which might be as simple as a CMC ("current balun"), or a full on transformer.) Also, don't forget the burden resistor, and an attenuator if needed.
To calibrate such a beast, you need a consistent current source. You might set up a large coax or stripline model, where a conductor is surrounded on all sides, sizing things to get a 50 ohm characteristic impedance. Use tapered sections to go between the input connector and the large model. Put the current probe on that. Like the, oh, I ran across that, uh, 5.2.8.3, make that thing and calibrate your probe against a 50 ohm termination on one end and a known source on the other.
Tim