Tektronix 465B - help needed -> fixed!
« on: May 24, 2013, 04:25:35 pm »
Hi,

First up, a disclaimer, I am an electronics beginner so please don't assume too much!

I have recently acquired a Tektronix 465B and HP 8011A pulse generator as a package deal. All seemed to be working well and I was playing around with the scope learning how it all worked. All the functions that a new, inexperienced in electronics, user would be expected to find seemed to be fine. Then I forgot to turn off the signal generator before I turned the scope back on while it was hooked up. What I observed is as follows:
1. The power light comes on
2. The lights behind the volts/div, indicating the selected setting, also came on for a short period before immediately going out
3. The screen then proceeded to light up slowly with a very diffuse glow on the right hand side of the screen

The fan -may- have also stopped although this could just be due to the auto start up sequence... although as this is analog I expect the fan is based of a thermistor of some sort? so as it had been running and was already warm it should have kept going?

Following this tragedy the last step is all that happens on a restart with no input. The focus does seem to have some effect (very minimal) and other lights come on when the corresponding buttons are pushed.

No magic smoke was seen/smelt escaping.

I have attached two images below showing the current state of the scope (you -may- be able to see that it is faintly glowing on the rhs of the screen) and the settings for the pulse generator when it all went south.





As I said before, I am a novice, although the reason I got hold of this scope is to learn electronics and this may be a good (or bad) way to dive into the deep end straight up. I am very willing to try fix it myself with a little advice from the gurus here.

Essentially I am asking very humbly if those of you with experience in these matters think that I have a chance of fixing this and if so how do I go about it?
« Last Edit: May 29, 2013, 11:31:01 am by nomad86 »
 

Offline PaulAm

  • Frequent Contributor
  • **
  • Posts: 939
  • Country: us
Re: Tektronix 465B - help needed!
« Reply #1 on: May 24, 2013, 05:02:48 pm »
Forgetting to turn off the signal generator has nothing to do with your scope problems.

The 465s have a few common failure points.  Two of the most common are filter caps that fail open and bridge rectifiers that fail.  The very first thing to do is to check the power supply voltages.  You can get a service manual free off the net and there are test points located on the interface board (the large board on the bottom).  Check the +55V, +110,  +15, +5 and -8, they should all be very close (within 0.1V).  If you can check it, ripple should be in the mv range.  If the +55 is off, nothing else will be correct since everything uses that as a reference.  If the voltages are out of spec, or zero, you get all manner of weird behavior.  Fix the PS before doing anything else.

If all the voltages are correct, there's likely a problem in the horizontal amp somewhere.  Unfortunately, it's helpful to have another scope when you need to fix a scope.  It's fairly straightforward to track back until you find the bad section.  If you don't have another scope, check the wires on the horizontal deflection plates; it's possible one might have come off.  Otherwise, you might get an idea what's going on using a DMM on the horizontal output.  You might have a bad transistor there somewhere.  I think they may have started soldering in the transistors on the 465b; on the 465 they were in sockets and it was trivial to pull one to test it (although the sockets are a big failure point as well).
 

Offline kfitch42

  • Frequent Contributor
  • **
  • Posts: 300
  • Country: us
Re: Tektronix 465B - help needed!
« Reply #2 on: May 24, 2013, 05:33:33 pm »
What happens when you hit "Beam Find" ?

Do any other knobs (besides focus) make a difference in the 'glow'? In particular horizontal or vertical position. What about intensity?

Knowing which controls seem to effect the output can help in narrowing down where to look.

 

Offline Paul Price

  • Super Contributor
  • ***
  • Posts: 1433
Re: Tektronix 465B - help needed!
« Reply #3 on: May 24, 2013, 07:24:09 pm »
I am quite a fan of Tek scopes.

I think your fan should start from the get-go. Power supply failure?

You've already  probably got past these simple settings  below but.....

Have you set the Timer Base to something abut 1 mSec/div (Is the Horiz possibly set to Ext. Inp.??)
Have you set the Vertical controls and Horiz. Pos controls to middle of their range?
Have you set the Trigger to "Auto"
« Last Edit: May 24, 2013, 07:29:19 pm by Paul Price »
 

Offline nomad86Topic starter

  • Contributor
  • Posts: 11
  • Country: au
Re: Tektronix 465B - help needed!
« Reply #4 on: May 25, 2013, 12:46:21 am »
Thank you all for your replies, they are greatly appreciated.

Before I take the case off and start poking around I will try to answer the operational questions.

I have the scope set to the following settings:
- all position knobs in the middle of their range
- both volts/div = 1V
- time/div = 1ms
- trig mode = auto
- trig source = norm
- ch1 & ch2 buttons depressed
- Horiz display A button depressed

(hope I haven't missed something important)

Q: What happens when you hit "Beam Find" ?
A: Nothing... (this was working before)

Q: Do any other knobs (besides focus) make a difference in the 'glow'? In particular horizontal or vertical position. What about intensity?
A: All position knobs have no visible effect. Intensity causes the glow to change in intensity very slightly but only in certain parts of the knob sweep and no change is visible between these little 'flickers'.

---

Something else that I forgot to add previously is that after the scope stopped working properly and I was fruitlessly poking buttons and turning dials the following occurred. When rotating the time/div knob a single very bright blob appeared at the far left of the screen, like the entire output of the electron beam was loosely focused just off the screen. With further rotation of the same knob this blob moved from left to right sometimes with the standard sweeping movement of a trace and other times it just stopped in one place. Meanwhile the indicator lights on the volts/div knobs occasionally lit up as I fiddled with time/div. This occured for about two or three back and forths of the time/div knob before it reverted to the 'diffuse glow on the right'. I have not been able to replicate this since.

---

PaulAm, thank you for your detailed instructions. I am also grateful to hear that the signal generator is not the cause, I can now tell the wife that I didn't kill it, it just 'happened'!

I have gotten hold of a service manual and will attempt to trace the fault. I will report back soon!

Many thanks once again.
 

Offline nomad86Topic starter

  • Contributor
  • Posts: 11
  • Country: au
Re: Tektronix 465B - help needed!
« Reply #5 on: May 25, 2013, 02:30:51 am »
I have checked the regulated voltages and they have values as follows:
(checked at location marked with circle in image below, ground used is the one to the left of the circle)

+55 = +46.6
+110 = +101.4
+15 = 0.0
-8 = +0.6

So it would seem there is definitely something wrong with the PS.

Following the power supply isolation flow diagram in the manual I am now looking to check the un-regulated +55V line. Unfortunately I can't find anything in the manual that specifically states where this is on the board. I am assuming that it is the point marked in a square in the below image. I just want to make sure that this is the correct test point as this point is in an area marked "DANGEROUS VOLTAGES" and I like being alive.

Now, maybe getting head of myself, if this point is low then capacitor C4429 and C4439 need to be checked. The layout in the manual is a bit confusing so I thought I would ask here. In my picture I have labelled two double solder points with what I think is on the under side of the board, am I right with this?



Looking forward to more good advice.
 

Offline ivan747

  • Super Contributor
  • ***
  • Posts: 2046
  • Country: us
Re: Tektronix 465B - help needed!
« Reply #6 on: May 25, 2013, 02:39:06 am »
A good way to measure things around high voltage equipment is to clip the negative lead of your multimeter to a ground point on the scope and only use the positive lead. Keep a hand out of the place (maybe in your pocket or on your back) and keep the instrument disconnected.
 

Offline Anson

  • Regular Contributor
  • *
  • Posts: 135
  • Country: us
Re: Tektronix 465B - help needed!
« Reply #7 on: May 25, 2013, 02:58:04 am »
Check these two caps. The plus signs I drew are the positive leads.
« Last Edit: May 25, 2013, 03:00:01 am by Anson »
 

Offline nomad86Topic starter

  • Contributor
  • Posts: 11
  • Country: au
Re: Tektronix 465B - help needed!
« Reply #8 on: May 25, 2013, 03:17:32 am »
Anson,

forgive the complete novice question here, but when you say "check these two caps" I interpret this as check the voltage from ground to the positive leads you have marked, or should I be checking the capacitance somehow?

Thanks for your input.

UPDATE: I went ahead and tested the voltage from ground to the centre point of all groups of 3 solder blobs and got the following voltages:

-8 = +15.4
+15 = +25.2
+55 = +80.3
+110 = +163.5
+5 = +10.5

hope this helps...
« Last Edit: May 25, 2013, 03:28:40 am by nomad86 »
 

Offline Paul Price

  • Super Contributor
  • ***
  • Posts: 1433
Re: Tektronix 465B - help needed!
« Reply #9 on: May 25, 2013, 03:30:03 am »
Check the caps means see if they open or shorted.  You can bridge a cap with a similar voltage rating and >= capacity. If you have a multimeter you can see the cap charge on the ohms scale if it is disconnected from the circuit.
If the circuit is power, you can use the AC scale on your MM and read the AC voltage. The voltage across an electrolytic cap should be close to zero, if it isn't you've probably found the problems.
 

Offline Rerouter

  • Super Contributor
  • ***
  • Posts: 4700
  • Country: au
  • Question Everything... Except This Statement
Re: Tektronix 465B - help needed!
« Reply #10 on: May 25, 2013, 03:51:29 am »
-8 = +15.4

???

if you meant a negative there, that would imply the caps are atleast not shorted or open circuit, could still be cruddy from age, but points the finger towards the regulation stage,
 

Offline nomad86Topic starter

  • Contributor
  • Posts: 11
  • Country: au
Re: Tektronix 465B - help needed!
« Reply #11 on: May 25, 2013, 04:14:23 am »
-8 = +15.4

???

if you meant a negative there, that would imply the caps are atleast not shorted or open circuit, could still be cruddy from age, but points the finger towards the regulation stage,

Nope, I get +15 as shown in pic. It should be some negative value right? Am I using incorrect measuring technique? That is why I posted a pic so you can tell me if I am doing it all wrong.



I have checked the voltages from the centre post to the outer post, with the positive (red) DMM lead on the centre post and get the following readings when using the AC scale:

-8 = +31.8
+55 = +176
+110 = +184.2
+15 = +54.3
+5 = +22.2

and with the DC scale (just for completeness)

-8 = +14.7
+55 = +79.6
+110 = +83.0
+15 = +24.9
+5 = +10.5
 

Offline nomad86Topic starter

  • Contributor
  • Posts: 11
  • Country: au
Re: Tektronix 465B - help needed!
« Reply #12 on: May 27, 2013, 02:38:59 pm »
Ok, so I have placed an order for a FX-888 solder station and UNI-T UT61E multimeter so I can de-solder and test some of those caps. In the meantime I thought I would test a few voltages on the +15V rail as it shows an approximately correct unregulated voltage (test = +25.2V, spec = +21.4V) but I get 0.0V on the regulated test point. The following pic is what I have measured with my $20 DMM.



Now, my knowledge of circuit theory comes from a (10yrs ago) 1st year general electronics uni. course and so does not quite cut it for interpreting what these values really mean. However, I am suspicious of U4411B and Q4401, as the resistors seem to check out and so do the removable transistors. I tested the transistors with the diode function on my DMM according to http://www.wikihow.com/Test-a-Transistor. The op-amp is a plug in unit and so should be testable once I have other supplies I am waiting on (breadboard, assorted resistors, caps, etc). The transistor, Q4401, on the other hand is soldered in so I have to wait for my soldering iron. Also, I am really puzzled by the change in voltage from R4403 to C4331 from 0.01V to 5mV, or is this probably just the accuracy of my DMM?

My question is, am I barking up the wrong tree? Also, should I be disconnecting all the other boards from the interface board according to the isolation procedure in the manual? I have gotten the impression from forum posts that this is not always necessary and I would like to leave as much alone as possible.

I am trying not to be a "please spoon feed me" type so if I am coming across that way please tell me!

EDIT: If this topic should be in the "Beginner" section then please feel free to move it!
 

Offline Anson

  • Regular Contributor
  • *
  • Posts: 135
  • Country: us
Re: Tektronix 465B - help needed!
« Reply #13 on: May 27, 2013, 03:19:51 pm »
What voltage do you have on the 8 pin of the opamp U4411? I suspect you have low voltage into the amp.
 

Offline SeanB

  • Super Contributor
  • ***
  • Posts: 16362
  • Country: za
Re: Tektronix 465B - help needed!
« Reply #14 on: May 27, 2013, 03:37:02 pm »
C4431 or something on the output is dead shorted. Q4401 should be getting rather warm in this condition. Check the resistance across C4431 with the power off, anything other than a reading below 100R is OK.
 

Offline nomad86Topic starter

  • Contributor
  • Posts: 11
  • Country: au
Re: Tektronix 465B - help needed!
« Reply #15 on: May 28, 2013, 12:39:40 am »
Thanks very much guys.

Anson:
The 8pin shows a voltage of +6.4V so is definitely low. Should be +22V if I am reading the schematic right. Looks like we are pulling a bit too much current through R4412 (off the +55V line) to get this voltage drop. I am assuming this is related to the problem with C4331 discussed below.

SeanB:
If you were meaning to say C4331 then I think you may have something. The resistance across C4331 is 0.7R which is exactly what I get if I simply cross the probes on my DMM. Q4401 also heats up quite quickly. I'm not not sure how hot it is supposed to get but I think it is too hot, good thing it is mounted directly to the chassis.

It looks like you have guided me into finding at least one fault, greatly appreciated!
 

Offline SeanB

  • Super Contributor
  • ***
  • Posts: 16362
  • Country: za
Re: Tektronix 465B - help needed!
« Reply #16 on: May 28, 2013, 04:50:22 am »
So now disconnect the output from that rail and measure again. If still a short C4431 is shorted, otherwise it is fine and the fault lies on something powered from the supply rail, probably another tantalum capacitor.
 

Offline nomad86Topic starter

  • Contributor
  • Posts: 11
  • Country: au
Re: Tektronix 465B - help needed!
« Reply #17 on: May 28, 2013, 09:16:08 am »
Ok, disconnected everything specified in the isolation procedure (P4061, P4060, P4057, P4058, P4398, W7292, W4481, W4440) and I see the same values as before. So it is pretty much definite that C4331 is shorted. I have on hand a 47uf +16V tantalum cap from jaycar but the one on the board has written on it K47 +20V. Should this be a problem? It should only be seeing +15V so I am assuming the +20V spec on the existing cap is for a safety factor? They also look somewhat dissimilar with the one on the board being about twice the size as the jaycar one. Maybe I should just order another cap (maybe http://au.element14.com/kemet/t350k476k020at/capacitor-tant-47uf-20v-1-2-ohm/dp/1171967) and be patient...
 

Offline Rerouter

  • Super Contributor
  • ***
  • Posts: 4700
  • Country: au
  • Question Everything... Except This Statement
Re: Tektronix 465B - help needed!
« Reply #18 on: May 28, 2013, 10:48:18 am »
higher voltages only, a 1V margin with a tant on a supply rail is not a nice mix,
 

Offline Anson

  • Regular Contributor
  • *
  • Posts: 135
  • Country: us
Re: Tektronix 465B - help needed!
« Reply #19 on: May 28, 2013, 03:07:38 pm »
Couldn't he use a mica or ceramic cap here? Haven't studied up on capacitor differences yet. I know some are polarized like electrolytic's and some are not like ceramics. Are tantalum's polarized? I haven't done anything with them yet I here they are horrible caps.
 

Offline SeanB

  • Super Contributor
  • ***
  • Posts: 16362
  • Country: za
Re: Tektronix 465B - help needed!
« Reply #20 on: May 28, 2013, 04:08:44 pm »
Replace it with a 100uF 35V 105C low ESR capacitor instead. Do the same to the others in the power supply of the same value. Simpler than looking for a 47u 35V unit, and probably cheaper as well.
 

Offline nomad86Topic starter

  • Contributor
  • Posts: 11
  • Country: au
Re: Tektronix 465B - help needed!
« Reply #21 on: May 28, 2013, 11:22:04 pm »
So I should try to stay away from the tantalum caps?

I looked up 100uF 35V caps on element14 (hoping for quick postage in Aus) and found the following:

CAP, ALU ELECT, 100UF, 35V, CAN

(there were others similar to the above link but it was the smallest)

I am assuming that the polarization due to being an electrolytic (thanks Anson) shouldn't be an issue as long as I align it correctly on the board?

Otherwise I am back to the tantalum caps:

element14 tantalum caps

From what I have gathered from reading around is that the values of the capacitors in the power supplies of these old scopes are not too important i.e. as long as it is equal or higher then it should be good. So I should be looking for a relatively cheap option that is slightly over-specced compared to the original part?

Many thanks for this advice. I am on a real learning curve at the moment and this help is greatly appreciated.
 

Offline chulomex3

  • Newbie
  • Posts: 3
Re: Tektronix 465B - help needed!
« Reply #22 on: May 28, 2013, 11:24:55 pm »
Did you get this fixed?  let me know if you need anything as far as documentation.
 

Offline Anson

  • Regular Contributor
  • *
  • Posts: 135
  • Country: us
Re: Tektronix 465B - help needed!
« Reply #23 on: May 28, 2013, 11:55:30 pm »
So I should try to stay away from the tantalum caps?

Totally up to you. Some people say they are great and so do some books and papers on the net. Others say they are horrible and have a tendency to short and burst into flames. Never seen that  personally yet and so far in the last few weeks as I've been working on my own tek scope I haven't had a problem with them.

I am assuming that the polarization due to being an electrolytic (thanks Anson) shouldn't be an issue as long as I align it correctly on the board?

Most definitely.

From what I have gathered from reading around is that the values of the capacitors in the power supplies of these old scopes are not too important i.e. as long as it is equal or higher then it should be good. So I should be looking for a relatively cheap option that is slightly over-specced compared to the original part?

I think as close as you can get to the original is best. Be sure you only do this in the power supply part though. If you have issues on the vertical or horizontal sections you will want to try and stick to the originals and even stay with the same types. Different types of caps react differently to different frequency oscillations.

 

Offline nomad86Topic starter

  • Contributor
  • Posts: 11
  • Country: au
Re: Tektronix 465B - help needed!
« Reply #24 on: May 29, 2013, 11:30:03 am »
Well, a very big thank you for all those who have given me input! I can report a working Tektronix 465B thanks to eevblog.com and its members.

I couldn't hold out any longer and decided to temporarily solder in my replacement of C4331 with my tantalum cap from Jaycar which is only rated for 16V. I have another on order that is rated for 35V but the 16V one was burning a hole in the desk just looking at me. Soldered it in and success! Immediately had the single trace that you get with no input, ch1 & ch2 and a few buttons didn't work but I was very happy regardless.

Then I realised I hadn't reattached the plugs on the interface board that I had disconnected as part of the power isolation routine. Plug those in and even better success, everything works!! Checked the regulated voltage test points and they are all within approx 0.2V with most within about 0.1V.

At first I thought there must be other things wrong as the trace was very 'wiggly' until I realised my fluoro desk lamp was polluting the signal something chronic.

Here is a pic proving that it works. If it looks a bit weird that is because the CRO is upside down on the desk and I have rotated the photo.



I am planning on replacing the 16V cap asap with the 35V one on order. Overall I am very surprised that I only needed to replace the one cap. I am half expecting something else to release the magic smoke any time now.

Once again, many many thanks to all those who have supplied advice and comments, it is very greatly appreciated.

Now for a more mundane, although related, question. One of the volts/div knob seems to give me a noisy trace when I have it on 0.2. I am assuming this is due to cruddy contacts in the selector switch mechanism. The manual says to only use isopropyl alcohol and that is about all they say. I am wondering if there are any experienced hands out there who can give me an idea of procedure for cleaning these switch assemblies?
 

Offline Rerouter

  • Super Contributor
  • ***
  • Posts: 4700
  • Country: au
  • Question Everything... Except This Statement
Re: Tektronix 465B - help needed -> fixed!
« Reply #25 on: May 29, 2013, 12:46:52 pm »
not so much for the procedure, but isocol branded isoproply you get at chemists has so far never left any residues for me,

from memory it involves wetting a sheet of something with it, then sliding in and out underneath the contacts to clean them, perhaps someone else can fill that in...
 

Offline nomad86Topic starter

  • Contributor
  • Posts: 11
  • Country: au
Re: Tektronix 465B - help needed -> fixed!
« Reply #26 on: May 29, 2013, 03:30:50 pm »
Yep, read around a bit more and that sounds like the accepted process. A bit of paper with isopropy alcohol on it GENTLY rubbed between the contacts. I have a some isocol on hand so I will give it a go. Thanks Rerouter.

Another question, if I do the following:
- hook up the calibration signal output into the A trigger input
- set the A trigger switch to EXT
- set the VERT MODE switch to A TRIG VIEW
what should I see on the trace?

I am currently seeing just a plain line. Note that I can get the square wave from the calibration output to display without any worries when input into CH1 or CH2. I then leave the time base and the volts/div at the same settings when I follow the above steps.

My understanding was that it takes the input given by the trigger selector and displays it without any amplification so what I see (just a line) is correct. But the operators manual procedure for Basic Timing Check makes me think I should see the same signal as if i plugged it into CH1 or CH2. I am confused...

EDIT: Figured it out. I will write up my findings just in case someone else needs it. The input into A TRIGGER EXT should be displayed the same as if it is fed into CH1 or CH2 except with a constant vertical scale of 0.1V/DIV (as stated on the front panel). You can't adjust this scale although you can adjust the overall vertical position with the A TRIGGER SLOPE knob. This is probably obvious to anyone who has a working 465B. The reason I was only getting a single straight line for a trace was that the two connectors (J7535 & J7524 on schematic for A & B trigger generators, this is missing from most manuals off the net, find it at http://bama.edebris.com/manuals/tek/465b/ under "465bMissingPages.pdf") were not seated correctly. The wire end was out to the side of the connector assembly rather than in the socket and so no signal was getting through. Simply re-seat these connectors and you're good to go!
« Last Edit: May 30, 2013, 01:41:00 pm by nomad86 »
 

Offline scatterandfocus

  • Regular Contributor
  • *
  • Posts: 165
  • Country: us
Re: Tektronix 465B - help needed -> fixed!
« Reply #27 on: September 19, 2019, 08:33:11 pm »
I got a 465B recently with exactly the same symptoms, showing similar voltages, and sure enough C4331 is shorted  just the same.  When the scope arrived it showed a trace, although not a good one.  After the next power cycle it showed exactly as nomad86 described.  Good work  nomad86, SeanB!  I was holding off a bit on ordering a bunch of components until after I got a chance to open this thing up and look at the voltages just in case the problem was only a failed component that I could track down.  And then I ran into this thread while searching on the low voltage at the +55V test point and other off voltages at the nearby test points.  So hopefully swapping out that cap will get the 465B back up.
« Last Edit: September 19, 2019, 08:49:56 pm by scatterandfocus »
 

Offline scatterandfocus

  • Regular Contributor
  • *
  • Posts: 165
  • Country: us
Re: Tektronix 465B - help needed -> fixed!
« Reply #28 on: September 19, 2019, 09:43:23 pm »
I borrowed a 100uF 20V electrolytic cap from a broken battery charger, and I got a beam and a trace!  Woo ha.

I think this thing is going to have some other issues though.  The display seems to have alot of flicker, the beam is a little off kilter, and the beam trace seem a little fuzzy and noisy.

Edit:  The flicker was due to the holdoff setting.  The kilter seems ok after adjusting the trace rotation.  Maybe the slightly fuzzy/grainy trace is normal on this scope?  I have seen some people mention that the 465B doesn't have the sharpest trace.
« Last Edit: September 19, 2019, 09:56:33 pm by scatterandfocus »
 

Offline james_s

  • Super Contributor
  • ***
  • Posts: 21611
  • Country: us
Re: Tektronix 465B - help needed -> fixed!
« Reply #29 on: September 19, 2019, 11:02:29 pm »
Does the fuzzy trace clean up if you set the inputs to Gnd? My 465B doesn't have the sharpest trace I've seen but I wouldn't call it fuzzy either.
 

Offline scatterandfocus

  • Regular Contributor
  • *
  • Posts: 165
  • Country: us
Re: Tektronix 465B - help needed -> fixed!
« Reply #30 on: September 20, 2019, 12:53:28 am »
Does the fuzzy trace clean up if you set the inputs to Gnd? My 465B doesn't have the sharpest trace I've seen but I wouldn't call it fuzzy either.

It doesn't change anything when doing that.  Rather than fuzzy, I would better describe it more as the trace being thick and having soft edges, in comparison to a Hitachi that I have.  I think it is functioning as it should be, and it is not a big deal.  There was some good discussion on trace sharpness here:  https://www.eevblog.com/forum/repair/tektronix-465b-repair-97327/

This is a hell of a cool old scope for a beginner to tinker with.
« Last Edit: September 20, 2019, 04:25:22 am by scatterandfocus »
 

Offline james_s

  • Super Contributor
  • ***
  • Posts: 21611
  • Country: us
Re: Tektronix 465B - help needed -> fixed!
« Reply #31 on: October 15, 2019, 05:15:09 pm »
I can pull my 465B out and get a few pictures of the trace later if it would help. I agree it's a lovely old scope and an iconic classic, first "modern" scope I owned and I still hold onto it because the XY mode blows away that on my DSO. Also sometimes it's just fun to use an old analog scope. The fact that it's almost as old as I am and still a useful instrument says something about the design.
 