Products => Test Equipment => Topic started by: Jay_Diddy_B on May 30, 2015, 02:23:43 am
Hi Group,
Time to share a repair that I made to a Tektronix 494AP Spectrum Analyzer.
A friend gave me the SA to repair, it would not work at all, it would blow the main line fuse at the back of the unit.
Schematic
(https://www.eevblog.com/forum/testgear/tektronix-494ap-spectrum-analyzer-repair/?action=dlattach;attach=154384;image)
I did some poking around with a meter and I found that one the rectifiers in the bridge was short-circuit (CR3096 or CR4094). In addition one the main electrolytic capacitors was short-circuit (C6101 or C6111).
The rectifier was easy to repair, I used a 1N5408 3A diode.
Electrolytic Replacement
The original capacitor in the spectrum analyzer were manufactured by Aerovox. They have a 5 pin base. This was probably chosen by Tektronix for mechanical stability.
(https://www.eevblog.com/forum/testgear/tektronix-494ap-spectrum-analyzer-repair/?action=dlattach;attach=154386;image)
I decided that I would make some adapter boards, so that I could use two pin snap-in electrolytic capacitors as a replacement.
(https://www.eevblog.com/forum/testgear/tektronix-494ap-spectrum-analyzer-repair/?action=dlattach;attach=154388;image)
(https://www.eevblog.com/forum/testgear/tektronix-494ap-spectrum-analyzer-repair/?action=dlattach;attach=154390;image)
I went to all this trouble with the adapter boards, because the end of the capacitor pokes through a hole in the chassis. The replacement capacitor has to be in the same place as the original.
(https://www.eevblog.com/forum/testgear/tektronix-494ap-spectrum-analyzer-repair/?action=dlattach;attach=154392;image)
(https://www.eevblog.com/forum/testgear/tektronix-494ap-spectrum-analyzer-repair/?action=dlattach;attach=154394;image)
Working again
The Spectrum Analyzer works fine now that the capacitors and the rectifier have been replaced.
(https://www.eevblog.com/forum/testgear/tektronix-494ap-spectrum-analyzer-repair/?action=dlattach;attach=154396;image)
I found the Service manuals at:
http://www.ko4bb.com/manuals/ (http://www.ko4bb.com/manuals/)
Regards,
Jay_Diddy_B
Sweet!
Very cool.
Thanks for posting. I always find such repairs very instructive.
Very cool, those are the same style of caps used in the 3456A's infamous power supply. I just used a radial cap, and bent the pins for best fit.