Tektronix 7844 traces fade with scan speed
« on: October 28, 2022, 03:27:50 pm »
Hi,
I have a 7844 oscilloscope that has two problems:
1) traces fade with scan speed, until they are no longer visible at 50ns/div; if I raise the intensity, I only lose the focus. The problem is exactly the same on both beams.
2) the readout "dances". I have already tried changing the whole board with another that is certainly good, but nothing changed. All voltages seems to be right and clean.
I have a video but I am not able to upload it (which format should I use?)

Unfortunately I have not a high voltage probe, but in 7844 I do not find a test point for it however.

Can anyone please give me some direction where to investigate?

Thanks
 

Offline perdrix

  • Frequent Contributor
  • **
  • Posts: 640
  • Country: gb
Re: Tektronix 7844 traces fade with scan speed
« Reply #1 on: October 28, 2022, 08:49:24 pm »
Which timebase are you using?  What frequency is the signal - if you're using a 1kHz signal you should not expect to be able to see it at 50nS/div
 - think about it for a while and you should be able to work out why.

D.
 