smaultre:
I have RSA5126A, that needed calibration.  :-BROKE
I have some calibration menu in Tek software, but i didn't understand how it works.
Maybe someone have such experience, and can help me with calibration?
Also i have some parts, or partially worked/untested modules for RSA6100 series.
And yes, i'm looking for wide bandwidth module for 5100A series, and a way to generate some new options ;)

squadchannel:
The adjustment procedure is in the service manual, or am I mistaken?

page 19.

mod:
Wait, are you talking about this?

--- Quote ---
Alignments Are Not Calibrations

Alignments are adjustment procedures run by the instrument using internal reference signals and measurements.
Calibrations can only be performed at a Tektronix service center and require the use of traceable test equipment (signal sources and measuring equipment) to verify the performance of the instrument.


--- End quote ---

Maybe that's not what I had in mind.
By calibration, do you mean you want to adjust the internal oscillator?
If so, it may be impossible as the manual says.

The adjustment that can be controlled from the settings menu is probably a common alignments in SA.
It would not be an calibration.

smaultre:
Taking about calibration..

squadchannel:
"Calibration" has two meanings: "Performance Verification" and "Adjustment". SA also includes "Alignment".

"Peroformance Verification", it is described in the 077-0520-05 PV manual.
"Alignment", it is done automatically or manually during SA operation.
"Adjustment", maybe done as described in the SM, perhaps with dedicated software. It would be impossible.

don't know what the "Calibration menu" is. Maybe it can be adjusting, but if you mess with it badly, you might be in big trouble. :scared:

smaultre:
I'm own something like prototyping device it's full of special features on menu and additional diagnostic apps..