Products > Test Equipment
Tektronix TDS1000B and TDS2000B series hacks
66bono:
TDS1001B
I read the batch and soldered the S29JL064H70TFI00 again. The oscilloscope works as before.
No idea about programming :)
I wanted to get the highest possible bandwidth.
I also experimented with OpenChoice Talker Listener
I established communication, failed to break the oscilloscope :)
Fieroluke:
The pitbull hack doesn’t work with the B-series scopes.
Check location 7F0007 of the file you programmed into the firmware. Which value does it have?
If the py-script didn’t recognize your scope, the file didn’t get modified and you programmed the original file again.
Which model does the startup screen report? (The „set time and date“-screen)
66bono:
I didn't change the original bin file. I just read it with the programmer and soldered the again
Fieroluke:
Don’t look for content 7F0007, look at ADDRESS 7F0007! The part on the left (blue) should be 7F0007 or 7F0000.
66bono:
7f0007 not found is 7f0008
Navigation