I've got a TDS2014 fixed and as a post-repair check, I had the frequency response measured. The measurement setup is the following: Agilent E4432B signal generator --> RG58 coaxial cable --> BNC 50-Ohm feed-through termination --> TDS2014. The measured frequency response is attached. However, the response does not look Gaussian to me and in-band roll-off starts too early at around 10 to 20 MHz. The measured 3-dB bandwidth is 100 MHz and rise-time is 2 ns. I was wondering if the 109-step calibration procedures in the service manual (with DC levels and 5 kHz, 50 kHz, 1 MHz, 20 MHz and 100 MHz sine waves) would correct for the frequency response. Thank you.
UPDATE:
The culprit was the poor RG58 cable. After replacing that with a Pomona 2249 BNC cable, the frequency response looks Gaussian.