Hello again: Many thanks for the fast reply!
NOT A TANT!
1/ all other power is OK! Had assumed the very slow appearance of the bug meant thermally related.
2/ Instead it is related to time/div cam switch and variable: just touching the var knob would cause delaying and/or delayed traces to slowly shrink towards the left.
3/ I tried to clean and exercise time/div using only 100% isopropylene alcohol, injected into slots. No go.
4/ Removed board on opposite side of time/div cam switch, check for loose screws on switch and injected alcohol thru holes. No go.
5/ Finally realized the var switch in center red knob of time/div may be the culprit.
6/ var Switch is at the back of long shaft, has a plastic actuator moving fingers on PCB.
Used same alcohol/exercise, IT WORKS!
The cause must be a marginal contact, opening after some time or movement, with timing Rt open circuit, the trace moves and shrinks slowly left.
Doing this repair, and removing the board on 6 precision standoffs, with about 30 0.025 long pin interconnects, I could only marvel at the electronic and mechanical engineering in the plugin! Fortunately I have the original TEK service manulas for both 7B92 and 7B92A.
Really a great plugin especially with 500 pS/div and extensive triggering capabilities. I will compare to 7B92A.
I welcome comments from you TEK 7Bxx mavens!
Cheers,
Jon