Author Topic: Tektronix VX1410A  (Read 521 times)
Offline MynameisntdjTopic starter
Tektronix VX1410A
« on: November 15, 2024, 12:49:57 am »
Well I wanna know if anybody knows a company that can repair the Tektronix VX1410A cages? I got in contact with Tektronix and they mention that it's obsolete and there is no replacement. So now I'm looking at other repair sources. Any help pointing me in the right direction would help a lot.
Offline calibrationfixture
Re: Tektronix VX1410A
« Reply #1 on: November 18, 2024, 07:40:42 am »
Hi Mynameisntdj,
When the Power Supply isn't working you can use the one of a Standard TLA7016. Both use the same Power Supply Module (119-5553-00). See also Page 9-2 of attached Document. I said Standard, because there are also High Power TLA7016's which have another Mains Connector and Partnumber.
Calibrationfixture
« Last Edit: November 18, 2024, 07:50:32 am by calibrationfixture »
« Reply #2 on: November 18, 2024, 04:41:05 pm »
Thanks a lot for that info. I'll search for one to purchase. I have about 3 or 4 of these I've accumulated at work over the years with the same problem.
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #3 on: November 18, 2024, 04:57:24 pm »
We used Tektronix logic analyzer and CPU plugins in a HP VXI crate successfully. I mean if you can get any working VXI crate, that should work as well
Regards, Dieter
Offline gslick
« Reply #4 on: November 18, 2024, 09:57:30 pm »
I have some TLA711/720/721 VXI mainframes that I don't use. I'm not sure if they all have the same 119-5553-00 power supply. I would have to check. I might consider parting the mainframes out for just the power supplies. They are easy to remove. The power supplies are heavy and shipping probably wouldn't be cheap, but still reasonable. Shipping an entire VXI mainframe would probably be too expensive to be reasonable.
The following users thanked this post: Mynameisntdj
« Reply #5 on: November 20, 2024, 12:26:38 am »
Thanks I'll keep that in mind. If I can prove that it truly is a power supply issue I can probably convince some of the higher ups to look at purchasing your stock. I currently have the unit open on my work desk picking apart the supply
« Reply #6 on: November 20, 2024, 04:04:50 am »
So power supply was not the issue. Found a faulty 555 timer on the backplane PCB. Replaced and fixed. 1 fixed 3 or 4 more to go.