Thanks David for looking the socket up. So it looks like it's yet another pain in the butt of a socket then.... here comes the IEC socket....
I don't have a passion for a million different vintage cables with impossible to find power cords. I rationalize on the more modern IEC 3 prong.
Well I don't know yet if I am gonna keep this one or sell it, so it's way too early to talk about details like power socket replacement.
Anyway, I worked some more on the thing. Seeing as contact cleaner in the tube sockets brought the power rails back to life, I did the same treatment to all remaining tubes in the instrument.
That would be 5 tubes. x4 EF86 and one E80F.
Each time, I pulled the tube, gave it a good clean because I like sparkling glass.
Then tested the heater on the lab supply to see if it works, and how much current it draws versus what the datasheet says.
Then some contact cleaner in the tube sockets and exercise the tube in its socket for like 20/25 cycles.
Applied power....tested the thing, using my lab power supply to sweep voltage, and my chinese Vref to make more accurate assessments and.... it works !!!
Yes... there is now life in the needle. it moves smotthly as I sweep the input voltage, and is stable, no jerkiness like the Ballantine TRMS VTVM.
In the lower ranges, the most sensitive ones, 10 and 30mV, it does what you would expect : needle goes haywire just by touching the input jacks with my fingers.
So that's good.
See ? Those blue axial caps were not at fault, told ya !
So it's now back to life, yeah !
But it's not perfect of course. We are now in the second pass of the troubleshooting process.... now that we got the big problems out of the way, there is now the second wave of problems, the less severe ones....
So, I see two main problems that need addressing now :
1) The needle reads way too low. At first it read 15% low, but then I let it cook for 30 minutes and when I came back, it had worsened and was now 25% off ! Grrr... EDIT : 10 minutes later it improved a bit and is now 20% off...
Way too much to be a calibration issue, especially since it moved in the course a few minutes, so dramatically.
I maybe have an idea / suspect : look at the schematic. In green I circled the amplifier section, the part that does the hard work. 4 tubes. x3 EF86 whose filaments drew 200mA as per the datasheet. However there is also an E80F, coloured in yellow. This one, its filament was drawing only 200mA but datasheet says it should be 300mA, 50% more ! That's quite significant. So maybe that tube is tired and has low emissions, and since it's the front-end tube so to speak, the first to see the input signal... I guess having a low gain here is not good. Might explain my very low reading ?!
However this tube is quite expensive. Not crazy expensive, but expensive enough to wonder if it's worth investing that money in the old boat anchor, especially since I am skint.
So I would rather want to troubleshoot the problem more precisely to be really sure that the tube is the problem. I wish I had a tube tester...
Unfortunately to do more detailed troubleshooting, the schematic is not enough... I now need to be able to read and UNDERSTAND what the manual says, theory of operation etc.... and it's just not possible with this bloody manual that's in GERMAN !!!!
I am not so good at German...
IF someone can find me that manual in French or English, that would be soooo helpful....
2) Problem #2 : the two indicator tubes on the front panel (coloured in magenta at the bottom of the schematic), that indicate the polarity of the input signal... neither of them ever lights up, it's so sad.
Might have a clue here as well : there is a tube / triode nearby (coloured in blue) that maybe is related to these indicator tubes. The filament of this tube is strange, it draws MORE current than expected.... it's an EF86 again, so should be 200mA like all the others, but this one happens to draw.... 300mA, 50% more ! By what phenomena is that even possible ?!
MAybe this is my problem... but just a shot in the dark and again, I need the freaking manual ot help me understand how this circuitry works so I can troubleshoot it precisely and avoid wasting money on a tube that does not need changing...
That's it for today... made good progress just this afternoon, quite happy. Now hoping I can find the root cause of these two remaining problems...