A bit more progress with the HP 180A, trying a set of untested plug-ins I still had the trace stuck off to the right, some voltage checks found yet another bad 4-071 transistor Q205 (completely open), replacing this moved the trace slightly to the left
but not enough.
More voltage checks I found 100V on both sides of diode CR209, the diode was fine, but the 22.1k resistor R271 was nearer to 1.5M, the one for the opposite amplifier circuit R234 was also bad, measuring at 33k.
Once these were replaced the trace was able to fill the screen. These are both 1/2W 1% metal film resistors, being run at 0.4275W, not much derating, similar size resistors seem to fail often in the PSUs of some HP scopes.
-group-therapy-thread/?action=dlattach;attach=2076197)
I didn't manage to get a picture of the trace with the focus set correctly, as it's now got a intermittent problem where the trace disappears once everything has warmed up. The trace is not good either, it's not flat, this could either be the plug-ins, or maybe the plates in the CRT are bent, it may have been dropped at some point, the plastic carrying handle on older HP 180As in known to snap without warning, one of the feet is broken from the too.
-group-therapy-thread/?action=dlattach;attach=2076203)
David