Here's a better picture of the 465 channel 2 leaf switches (Sorry, but my phone does not do good close ups so bear with me). The pair of leaf switches under that shield with the one on the right is the one that is open.

On the schematic with the coupling switch in DC and the Volts/Div switch set to 5mv the attenuators are completely bypassed and there should be continuity (0 ohms) between R51 and R65. There wasn't. It was open. Position 6 is the culprit.

I decided to go for broke and carefully pry the leaf switch from under the shield. Here is what I found. Not burnt but completely mangled!

How in the hell did that happen?

That leaf switch is completely protected by that shield. The only thing that makes some sense is that it was damaged at the factory, QC missed it because it's under the shield, and at some point it did make contact but no longer does.


I carefully straightened out the 2 prongs. The center prongs must be completely folded under or are missing. I couldn't tell. Carefully put it back under the shield.

Result after reassembly of the switch deck. It worked! Now have continuity. And I verified that it drops in and out with the rotation of the switch the same as channel 1. I think I finally nailed all the issues with this Vertical board.

Here's the Vertical board and the Switch board assembled and ready to install. I also fixed the BNC's. When it's complete I'll show the results. If all that's left is to completely calibrate the vertical section because I jack assed so much stuff around I'll be happier than a pig in shit.