Today I stuck an unterminated piece of wire up its input, and used that to measure the frequency of the 50Hz mains. The values agreed with those on
http://gridwatch.templar.co.uk/I also found the cal out on my HP8562 spectrum analyser is out by 100Hz in 300MHz. That's 0.3ppm - absolutely disgusting, if you ask me. I ought to dig out my uBlox GPS receiver, and have a look at that.
Good catch, look like it's in good shape
For good measure, you should double check the power supply. In most 53310A the caps in the power supply end up dying and leaking all over the place, causing a lot of damages.
Haven't spotted any problems, yet.
IC datestamps contain 1996, so 25 years old.
I've had a look at the PSU caps. None are bulging, there's no sign of leakage, and generally the inside and outside is very clean; even the white gunk is intact. While I've recapped many PSUs, I'm reluctant to remove all the caps and gunk without a reason. I ought to look at the ripple on the +5.10V line, but I can't see a spec for it.
Nice! The 53310A is really a sleeper instrument. You wonder how the heck you ever lived without it.
One of the things mine gets used the most for is as a counter. This might sound silly, but it makes a really excellent counter. With an external GPSDO ref, it makes tuning other frequency references a snap. And, unlike a normal counter, you can zoom way in on the signal for effective minute adjustments.
As an aside, if yours doesn't have the deep memory option installed, this is easy to add (you can still buy the SRAMs). You just populate the empty DIP sockets and move a couple jumpers. It noticeably improves the performance of the panorama function; also I believe it increases the depth of memory available for panorama, so you can use a larger time window.
I'd heard about the tuning ability, and it makes sense.
It does have the deep RAM; the only thing it doesn't have is opt 031, the high resolution variant of opt 030 2.5GHz RF channel C.
All in all it looks like a good buy.