502A
Started probing around the thing. Printed the jigsaw schematics, taped them back in one piece so I can actually read them... then did the same digitally so I can show it to you. Just because I am nice of course. Been reading a bit the theory of operation chapters in the service manual as well.
3AM so am tired, will try to make it short(er than usual).
First tested the trigger section. AUTO MODE somehow doesn't work, but I don't care for now because NORMAL mode works : I fed the calibrator signal to the lower beam, the one I suspect kinda works. Trigger circuitry can pick up on it no problem, it produces lovely text book pulses in return, I am very happy.
So these lovely trigger pulses then go to the " Time Base Generator " as Tek calls it. Schematic below. Checked the output for a sawtooth, no luck , zero signal, just DC. Then checked the input of the circuitry at the other end... trigger pulses go through an RC differentiator to turn the square pulses into pin sharp pulses/spikes. That part works fine. Then these sharp pulses are applied to the grid of the first tube in there, V135A, the input of multivibrator.
checked the output of this thing, ie plate of V145A / R148 as highlighted on the schematic. I get no signal, just DC.
Now, this multivibrator can be turned of if you flick the big XY switch on the side of the cabinet, to " XY" mode instead of "swept" mode. In swept mode, what I want to see working, the "tail" of the multivibrator, the two cathodes, are connected via this XY switch, to the -150V rail. So I operated that switch to make sure I got a solid -150V, and I do. So no problem there.
So that's good news I guess, I narrowed it down. I have trigger, a good signal going to the sweep circuitry.. but no sweep.
I also noticed that the RECURRENT mode does not work either. This mode is new to me, never seen it in any other Tek scope, must be a 502A special...
From what I understand from the schematic and the Theory of operation, it's supposed, in the absence of a trigger signal, to make the multivibrator free run, just like it does in the Trigger section when you select AUTO.
When I switch to RECURRENT mode,, as you can see on the schematic I am supposed to get a triangular periodic signal at the input of the multivibrator... but all I get is DC.
So either another problem, or just another hint to steer me in the direction of a common problem.
Happy for now, more tomorrow....
MISC, I noticed that one of the 3 HV tube rectifiers in the CRT section, has been replaced, no doubt about that... looks newer / cleaner than the other two, and the soldering job on the ceramic strips is hardly factory grade...
Good night.