Making progress !!
Before resorting to the risky business of pulling the CRT out, I tried one thing : part of the left side of the scope is accessible via a removable access panel that's there to give access to a board that's there.
Removed that panel, shined some light in there, and immediately found the rattling screw, got lucky !!!
So, I could then power up the scope at last ! And I immediately got traces on the screen !
Not familiar with 7000 scopes so I played with all the knobs and eventually managed to display a sine wave on all 4 channels yeah !
Readout works as well
So it's functional, but not working right at all though... look at the video clip I just made for you !
1) At the bottom of the screen near the center, there is a sloped straight line.... appears to be part of the readout (reacts to the readout brightness control).
2) The first trace at the top just can't trigger, it's moving all the time, and the shape f the signal is no correct at the very start, there is a sharp bend to it somehow
3) Last / fourth trace at the bottom spans only a small part of the screen, on the left, and the span varies from zero to XX % of the width, all the time. It comes and goes, as you can see in the video.
4) Not seen in the video, but the signals are only visible on a particular sweep speed. For most sweep speeds, I get nothing at all on the screen .....
I could try the other TB plugin though, see if that is the problem or not.
BUT.... I am not worried at all by any of this, at least not at this point because.... I already found a major issue that sure could cause all sorts of weird problems !
Checked all voltage rails. The reference rail, -50V, is spot on at -50.0V. All 5 analog rails are spot on and very low ripple (2 or 3 mV AVG on the DMM, good enough for now at least).
BUT.... the digital +5V rail is kaput ! DMM said 4.1Vdc and 1.4Vac !!
So I scopes it, see below. As you can see, I do get a nice stable 5V but.... every half cycle / 10ms, I can a huge voltage drop that goes almost all the way to the ground ! Surely the TTL chips in the scope can't like that....
So first thing now is to investigate and fix this 5V rail, then resume testing.
The fan is indeed loud and sounds horrible, just like in the video ! At first I thought maybe the damaged metal work in this area is hitting the fan blades, like always happens in my Tek 500 series hollow state scopes but... no, not so here. 100% of the noise is due to the fan an nothing else. Sounds like the bearing is shot. will need a new fan, will look into that.
But that will be for tomorrow... 00H30 here, tired and time to got to bed.