I've just picked up pre-loved Agilent U1253A multimeter for just over a hundred quid and, surprisingly, it came with a brand new cal certificate from what I consider to be a reputable test house (they shall remain nameless for the time being to protect the innocent / guilty!).
I've been playing around with this thing and, when in resistance mode and shorting the supplied test leads, I get a reading of -0.04?... Now forgive my rubbish physics, but this sounds a little strange...
(https://www.cantanko.com/wp-content/uploads/2014/10/IMG_1406-1.jpg)
My 87V, with the same leads, gives a (far more plausible) reading of 0.10? - I'm assuming this probably means the cal cert is worth naff all?
Any opinions welcome ;D
Cheers,
Harry
-
You have discovered the source of dark energy! :scared:
No seriously, the calibration is off IMHO.
-
Yeah - figured as much... Incidentally that was supposed to be a unicode ohm symbol after the readings - seems the act of posting just made me look like I'm a fan of the interrogative ;D
-
I doubt it's anything to worry about. I have a Fluke 289 here that will read -0.011 in low ohms mode with a shorting plug on the jacks. Right or wrong, I always attributed it to software calibration (stored offsets). Tens of milli-ohms is down in the noise for a handheld DMM using 2-wire method with low current, and you're probably going be using relative mode here anyway.
-
The meter should be reading the test leads resistance, so it is out by a significant amount. I would send it back to the calibration house, if at all possible. If they screwed that one, what else is possibly out of specs?
-
I'm not sure I understand the problem here. The manual for the U1253A states for the 500R range "0.05% + 10 counts". So the 10 counts alone are 0.1ohm for that range - meaning it can be as high as 0.14R while you read 0.04R on the screen. So even if its 0.1R as the Fluke says (and who knows if its really that accurate), its still within spec.
-
To be right on the limits of the 1-year accuracy specs, the test leads resistance should be 0.06 Ohm or less. Possible, but unlikely.
-
Just check that this isn't one of those meters that allow you to null out the test leads, then helpfully keep nulling after you turn it off and on again. I have seen some like that and it is a pain until you figure it out.