Author Topic: U1272A update  (Read 31373 times)

0 Members and 1 Guest are viewing this topic.

Offline PetrosATopic starter

  • Frequent Contributor
  • **
  • Posts: 625
  • Country: us
U1272A update
« on: May 26, 2011, 01:37:33 am »
I've been using mine at work and also (while at work...) playing around with the switch to see if I could reproduce any of the issues. Some but not all the halfway positions give screwy readings, which may only show up while turning in one direction and not the other - for instance:

Moving from OFF to ACV will give a 0.L readings (no mode symbols) and secondary display flashes between 0L and 301(?) - ACV to OFF just shuts the meter off;

Moving from ACV to AC mV will change my secondary reading from temp to DCV and locks up the volt reading at 7.5V AC, while moving back from AC mV to ACV will make the mV reading jump around until I move the dial far enough that it switches back to ACV;

Between mVAC and VDC both directions seem ok, secondary display remains fine.

Between VDC and mVDC in either direction the display goes wonky until it settles on 0.L, secondary display remains fine;

From mVDC to Ohms it goes into the 0.L state in volts mode - from Ohms to mVDC the reading goes from 0.L Ohms to DCV mode with 21 MOhms displayed (both the DCV symbol and MOhms symbols are displayed), secondary display remains fine;

Ohms to diode test also shows both DCV and Ohms symbols with 21 MOhms. Move it slightly further and the voltage ramps up, the meter beeps a few times and finally displays . 0L V (additional space between decimal and 0), DCV symbol and diode symbol are active, secondary display remains fine;

From diode check to cap check the meter will start a steady beep and displays both DCV and diode symbols with a 0.0004V reading. Move further it will do the same kind of ramp up as between ohms and diode. The opposite direction will activate the DCV symbol and give a dead reading for about 10 seconds, after which it goes to .0L

Between A and uA the display goes wonky again, with readings as high as 12A shown. From uA to A, the display will go to 0L. and the backlight stays on (this seems to be connected to the lead sensors).

The bigger problem for me with the two current settings has to do with the alarm. Going from Off to A, both inputs work fine with no alarm. Once the meter has been in uA range, going back to mA/A setting will produce alarm beeps in one, the other or both inputs until the meter "gets the message" (no standard time) that it's ok for the lead to be in either of the two holes. This continues until I shut the meter off again (see video). It will also sometimes beep a warning in volt mode with the probe in the volt terminal after being in uA mode (I wasn't able to capture this for the video :( )

I miss my home I miss my porch, porch
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #1 on: May 26, 2011, 02:08:10 am »
Nice find PetrosA , I will start also to believe what some one else wrote ,
in another similar thread : That looks like that some one loaded windows Vista , on a windows XP hardware.  
 

Offline PetrosATopic starter

  • Frequent Contributor
  • **
  • Posts: 625
  • Country: us
Re: U1272A update
« Reply #2 on: May 26, 2011, 12:26:04 pm »
I forgot to add that I haven't had any issues during normal use with my 1272 except for the probes alarm, and that's been rare since I'm normally using a clamp meter (U1211A) for current. I just happened to be measuring current on a SMPS feeding a gang of LEDs at >25kHz and had to use this meter for that.
I miss my home I miss my porch, porch
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #3 on: May 26, 2011, 04:40:17 pm »
Just received  an reply from Agilent , about an question that I did related to firmware updates .

It is confirmed that both U1253B OLED DMM  and U1270 series ,
could receive firmware updates, by the use of the IR-USB cable .

For the U1253B there is all ready one update available .
(resolves an issue about the start-up screen design )

Additionally the users evaluating program about HH of Agilent is on Hold ,
and my own demo unit too.

Lets hope that their technicians works faster and coordinates better,
than the postman that they use, for delivering the mail .
  
« Last Edit: May 26, 2011, 04:44:19 pm by Kiriakos-GR »
 

Offline PetrosATopic starter

  • Frequent Contributor
  • **
  • Posts: 625
  • Country: us
Re: U1272A update
« Reply #4 on: May 28, 2011, 04:42:19 pm »
I'm getting a response ready for someone at Agilent who I contacted about this. Can I assume from the lack of responses that no one else has been able to reproduce the mA/A-uA jack warning alarm?
I miss my home I miss my porch, porch
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #5 on: May 28, 2011, 05:09:52 pm »
I'm getting a response ready for someone at Agilent who I contacted about this.

PetrosA , as Greek to Greek , I will say to you " to hold an small basket " about the Agilent response.

Personally I have start to have doughs.

 

Offline saturation

  • Super Contributor
  • ***
  • Posts: 4787
  • Country: us
  • Doveryai, no proveryai
    • NIST
Re: U1272A update
« Reply #6 on: May 28, 2011, 05:18:02 pm »
The errors occurring are fairly diverse, it appears to depend on how much the firmware is confused by some unstable connection; the worse is a full lockup, with no ranges working, the best is the meter reads inaccurately but its readings are mostly reproducible.

You can get a series of repeatable errors after one glitch, then if you shut it off, and wait 1/2 minute, it will be fine until the next set glitch sets it off.  If you wait for only a few seconds after off, the glitch will not reset.

I've so many various types of errors happening on several 1272a but regardless, they can all be reset once the meter is off for some time.

I have not experienced an error once the DMM boots properly and is working on any range: that is if you are already working on e.g. VDC, then switch to ohms, VAC, mV, temperature etc., all is well, so long as you do not power OFF the unit through the rotary switch.  If the machine goes to sleep, its fine, if you wake it by pressing buttons.

I am about to disassemble one of the meters and examine the switching area and PCB to see if the problem can be user resolved.
« Last Edit: May 28, 2011, 05:22:05 pm by saturation »
Best Wishes,

 Saturation
 

Offline ejeffrey

  • Super Contributor
  • ***
  • Posts: 3722
  • Country: us
Re: U1272A update
« Reply #7 on: June 01, 2011, 09:41:14 am »
Can someone with one of these meters check something I just discovered on the U1253A?

When measuring on the 10 nanofarad (smallest) capacitance range, and measuring a capacitance below 100 pF, the meter is extremely slow to respond.  The display update rate is like 1/s, but if you zero the reading and then hook up a 33 pF capacitor it will take 10 seconds or more to climb to the appropriate value.  Disconnect the leads and it will take another 10+ seconds to fall back to near zero.  This makes it very difficult to get a good reading as you never know if it has actually reached the target value.

A 330 pF capacitor on the same range is no problem.  It gives a good reading immediately and has the normal update rate of several Hz.  I checked on a couple of other varieties of (less expensive) meters and they had no problem with the small capacitance.  Anyway, I am curious if the 127X series shows this behavior as well.
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #8 on: June 01, 2011, 10:10:46 am »
Hi ejeffrey .

It would be nice to post this question at the Agilent forum too.

https://forums.tm.agilent.com/community/viewforum.php?f=552&sid=ac542c0140e499099b6d22a1931f405e

By letting them know about it , they will investigate if there is a problem.
 

Offline saturation

  • Super Contributor
  • ***
  • Posts: 4787
  • Country: us
  • Doveryai, no proveryai
    • NIST
Re: U1272A update
« Reply #9 on: June 01, 2011, 04:12:13 pm »
I have a 1252a, it easily measures 2.4pF quickly and easily, just rechecked.


Can someone with one of these meters check something I just discovered on the U1253A?

When measuring on the 10 nanofarad (smallest) capacitance range, and measuring a capacitance below 100 pF, the meter is extremely slow to respond.  The display update rate is like 1/s, but if you zero the reading and then hook up a 33 pF capacitor it will take 10 seconds or more to climb to the appropriate value.  Disconnect the leads and it will take another 10+ seconds to fall back to near zero.  This makes it very difficult to get a good reading as you never know if it has actually reached the target value.

A 330 pF capacitor on the same range is no problem.  It gives a good reading immediately and has the normal update rate of several Hz.  I checked on a couple of other varieties of (less expensive) meters and they had no problem with the small capacitance.  Anyway, I am curious if the 127X series shows this behavior as well.
Best Wishes,

 Saturation
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37742
  • Country: au
    • EEVblog
Re: U1272A update
« Reply #10 on: June 03, 2011, 05:48:27 am »
Just got off the phone with Agilent who have gotten back to me after investigating this in their multimeter group in Penang (they took it very seriously).
They have found the problem and confirmed it is the ADC result locking up during certain range switch conditions.
They can fix it in firmware and are doing this now, or have already done so.
They will provide me with a formal email outlying the exact company position on this, so will keep you posted.

Dave.
 

Offline saturation

  • Super Contributor
  • ***
  • Posts: 4787
  • Country: us
  • Doveryai, no proveryai
    • NIST
Re: U1272A update
« Reply #11 on: June 03, 2011, 10:28:45 am »
This is fantastic news.  No previous reviewer caught this issue until it was reviewed on eevblog.  I recall the safety problem you found with a tweezer DMM that was of equal concern.  Likewise, after your Atten 858 hot air reworkstation review, other owners found issues with its build quality and posted on the forum.

It shows the eevblog reviews are more than just entertainment: they have been more thorough whether its by you or your audience.

There are reviews on the 1272a on element14.com that preceded eevblog by 6 months or more, and were trite and elementary regardless of the length of them in words.  Reviews in trade magazines are nearly worthless, they all read like advertising for a product. 


Just got off the phone with Agilent who have gotten back to me after investigating this in their multimeter group in Penang (they took it very seriously).
They have found the problem and confirmed it is the ADC result locking up during certain range switch conditions.
They can fix it in firmware and are doing this now, or have already done so.
They will provide me with a formal email outlying the exact company position on this, so will keep you posted.

Dave.
Best Wishes,

 Saturation
 

Offline PetrosATopic starter

  • Frequent Contributor
  • **
  • Posts: 625
  • Country: us
Re: U1272A update
« Reply #12 on: June 03, 2011, 11:50:23 am »
That is great news! I agree that this is a superb place to figure out issues like this and kudos to Agilent for keeping their head in the game. I wouldn't have figured out what was happening with my probe alarm if I hadn't tried to recreate the other on-off-on switch problems that I read about here.
I miss my home I miss my porch, porch
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37742
  • Country: au
    • EEVblog
Re: U1272A update
« Reply #13 on: June 03, 2011, 11:53:26 am »
Thanks.
Often I don't actually get to play with a product much before a review, so real time spent actually using the instrument can be rather scarce.
In this case I didn't even do a review, just a teardown, and it just happened to play up. Luck, or the EEVblog review curse?  ;D
There is always great follow up stuff from viewers after my reviews, finding and pickup stuff I overlooked. So the whole community is working well toward making better instruments.
Agilent actually commented on this to me today, and as many big companies are, are a bit unsure what to make of the whole new social media thing. But they do love and appreciate the feedback and like the way it's all working to produce better product.
After all, I think we all understand issues can slip out with gear getting more complex these days and design/release cycles getting shorter.
And it's the smart companies that know the ones who are seen to respond and fix things when problems arise, will be the ultimate winners.
Some of the design and management teams from various companies (not just Agilent) do actually read this, and presumably other forums, so they are listening!

Dave.
 

Offline saturation

  • Super Contributor
  • ***
  • Posts: 4787
  • Country: us
  • Doveryai, no proveryai
    • NIST
Re: U1272A update
« Reply #14 on: June 03, 2011, 02:04:04 pm »
Yes, that's good to know and I hope it also helps in keeping you self employed now.

With the fixes, the 1272a can indeed challenge the Fluke 87V easily, and after some weeks of added evaluation, its both easy to use ergonomically, and tough.  I'd surmise a users only concern is the digits are less easier to see on the 1272a vs the Fluke 87V, the Fluke has a 'lifetime' warranty vs 3 years on the Agilent, but otherwise, the Agilent is superior as a measuring instrument in all functions that the Fluke supports, and has more.

Just to add, added items were found at eevblog that have made a difference:

sleep power consumption bug in the 80 series III Flukes, that went unreported for almost 10 years
GSM interference bug in the Fluke 87V unreported since the series V was released in 2006

Unique items that were born in the forum and that no forum has been unable to match [ I know, I check!]:

The Rigol 1052e hack and support threads
Hantek/Tekway scope threads

Just the big one's that come to mind.

Here are some samples of the 1272a after a temperature test.  I took it down to 0F then up to 120F, and these are sample images of the DMM warming and cooling down.  RH% exposure to 100% caused the amp alarms to turn on on the 1272a that stopped once all the condensation dried up.  Rechecking the DMM shows all functions well.  During the temp stress the meters are tested for VDC, DCA, and ohms.  Since the meters are often very wet, I don't do any AC tests as it a shock hazard for me.




« Last Edit: June 03, 2011, 02:07:26 pm by saturation »
Best Wishes,

 Saturation
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #15 on: June 03, 2011, 03:57:14 pm »
I can confirm too , what Dave said about Agilent and the U1272A firmware update,
but even if they have everything under control, they need some time to prepare their strategy,
about making public the update.

Some times we forget the size of this companies, and the paper work,
that they need for every single step of action.

Agilent are a bit unsure what to make of the whole new social media thing.
Their confusion are totally justified , they play the game with the rules of the market,
and we ( reviewers and  testers )  some times we exaggerate.

1) Dave started to call the device as piece of Shit.
And kicking it around as to was a football.    :(

2) Saturation, instead of examining if this product had useful functions,
or if it is practical to use, or if it needs improvements ..
He started to analyze the DNA of the unit, and did blood exams to it,
and at the end I am very hard pressed to believe that he acted wisely so far.  :(

By evaluating our behaviors, we become better characters,
and if we maintain one positive climate with out exaggerating,
this positive climate it will raise the needed trust,
that those cooperation's or team works needs so to grow.

According to my own information, one evaluation unit shipped today ( Friday 3)  to me,
directly from the factory, with the latest firmware on it.

Even so, I am hopping,  Dave to receive an similar unit even faster than me.
So to finish the review, and restore the negative climate, if the product proved to be now as spotless one.

About Saturation : Now that there is a new firmware, he must get in the line and wait so to load the new one .
The firmware in his samples, today are called as outdated.
( at list you will have the joy to re do all the testing from the start )  :)

Smack me in the face if you dislike what I have write so far.  ;D
I do not care ..

 

 
« Last Edit: June 03, 2011, 04:12:58 pm by Kiriakos-GR »
 

Offline PetrosATopic starter

  • Frequent Contributor
  • **
  • Posts: 625
  • Country: us
Re: U1272A update
« Reply #16 on: June 04, 2011, 12:16:50 am »
I don't have any controlled way to test it, but after reading Saturation's post, I suspect the issues I've had with the probe alarms are related to humidity, and that less than 100% will cause the problem. From what Agilent has explained to me, the split plates in the socket are used to sense the presence of a probe in the jack by measuring for a short between the two plates. I speculate that depending on how that's done exactly, if more than one socket is detecting low enough resistance under high humidity levels, that could cause the sensor circuits to go whacky. Whatever the cause, it needs to be ironed out since many regions in the world have regular seasonal high humidity levels and the meter becomes unusable when the alarm is going off.
I miss my home I miss my porch, porch
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #17 on: June 04, 2011, 12:45:14 am »
My opinion is that there is no need for further speculations,
until the arrival and testing of the new firmware.

Your own video and thread , plus the findings of Dave, was known to Agilent.
And hopefully this update will cure all the known bugs.

About the split plates in the socket, lots of DMM are made by the same way and from other brands,
never ever had be reported an issue about humidity problems with this design.
Even the Fluke 87V uses this design !!

 

Offline PetrosATopic starter

  • Frequent Contributor
  • **
  • Posts: 625
  • Country: us
Re: U1272A update
« Reply #18 on: June 04, 2011, 01:21:34 am »
About the split plates in the socket, lots of DMM are made by the same way and from other brands,
never ever had be reported an issue about humidity problems with this design.
Even the Fluke 87V uses this design !!

Yes, but do they implement the same probe sensor solution?
I miss my home I miss my porch, porch
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #19 on: June 04, 2011, 10:32:55 am »

Yes, but do they implement the same probe sensor solution?


Personally I think that they do.
But due the fact that the U1272A uses an secondary PCB for the probe connections,
its easier to see the traces on the PCB.

Only the Fluke 28II is an exception about the leads sensing circuitry, it uses optical sensors,
because the leads holes they must be waterproof too.

For the Fluke 87 ( all generations ) it is known, that if it enters in leads holes humidity or oils or dirt,
you just have to clean the split plates in the socket, with a bit of cotton and alcohol.

 
 
  
 
« Last Edit: June 04, 2011, 10:35:43 am by Kiriakos-GR »
 

Offline PetrosATopic starter

  • Frequent Contributor
  • **
  • Posts: 625
  • Country: us
Re: U1272A update
« Reply #20 on: June 05, 2011, 03:59:14 am »

...you just have to clean the split plates in the socket, with a bit of cotton and alcohol.

More shit to carry on the truck ;)
I miss my home I miss my porch, porch
 

Offline toli

  • Frequent Contributor
  • **
  • Posts: 313
  • Country: il
Re: U1272A update
« Reply #21 on: June 05, 2011, 05:31:59 am »
Can someone with one of these meters check something I just discovered on the U1253A?

When measuring on the 10 nanofarad (smallest) capacitance range, and measuring a capacitance below 100 pF, the meter is extremely slow to respond.  The display update rate is like 1/s, but if you zero the reading and then hook up a 33 pF capacitor it will take 10 seconds or more to climb to the appropriate value.  Disconnect the leads and it will take another 10+ seconds to fall back to near zero.  This makes it very difficult to get a good reading as you never know if it has actually reached the target value.

A 330 pF capacitor on the same range is no problem.  It gives a good reading immediately and has the normal update rate of several Hz.  I checked on a couple of other varieties of (less expensive) meters and they had no problem with the small capacitance.  Anyway, I am curious if the 127X series shows this behavior as well.

I did notice the same thing a few days after I got it. I've tested a 33pF cap and it took a long time for it to settle. For 220pF it was way faster. I'm not sure what caused it, but I didn't pay much attention as I almost never use such small caps in my projects. Anyway, posting on the Agilent forum as Kiriakos  suggested could be a good idea.
My DIY blog (mostly electronics/stereo related):
http://tolisdiy.com/
 

Offline firewalker

  • Super Contributor
  • ***
  • Posts: 2450
  • Country: gr
Re: U1272A update
« Reply #22 on: June 05, 2011, 07:42:56 am »
After all, I think we all understand issues can slip out with gear getting more complex these days and design/release cycles getting shorter.

Dave.

We should also keep in mind that companies use peoples. Peoples like us, electronics engineers. We are not robots we make mistakes.
« Last Edit: June 05, 2011, 06:56:38 pm by firewalker »
Become a realist, stay a dreamer.

 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #23 on: June 05, 2011, 10:56:41 am »

...you just have to clean the split plates in the socket, with a bit of cotton and alcohol.

More shit to carry on the truck ;)

Lets hope that you have an exclusive space for the flammable materials.
Like the plastic compartment that you store your comb - cologne - etc etc   ;D  ;D  ;D
« Last Edit: June 05, 2011, 02:56:33 pm by Kiriakos-GR »
 

Offline PetrosATopic starter

  • Frequent Contributor
  • **
  • Posts: 625
  • Country: us
Re: U1272A update
« Reply #24 on: June 05, 2011, 02:52:16 pm »
Actually I'm thinking of trying a dielectric grease in the jacks. That should keep any moisture to a minimum.
I miss my home I miss my porch, porch
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #25 on: June 05, 2011, 03:27:05 pm »
Well personally I do have an question that bothers me some how.
And probably only Agilent has the answer.

The U1272A will enter in the Greek market with three years warranty ..
And this is an very positive policy.

Even so, what bothers me is the total period, that Agilent plans to support the U1270 family.
Example:
Spare parts availability ? Software support ? and for how long ?
This sort of answers, they will totally kill any doughs that blows away confidence.  :)







 

Offline saturation

  • Super Contributor
  • ***
  • Posts: 4787
  • Country: us
  • Doveryai, no proveryai
    • NIST
Re: U1272A update
« Reply #26 on: June 06, 2011, 11:50:30 am »
Hi PetrosA,

So far I've taken it to 100% RH without a problem, the problem is only if there is condensation during the extreme testing, just wait for the socket to dry or dry it with compressed air or tissue.  It is water tight and O ring sealed.


I don't have any controlled way to test it, but after reading Saturation's post, I suspect the issues I've had with the probe alarms are related to humidity, and that less than 100% will cause the problem. From what Agilent has explained to me, the split plates in the socket are used to sense the presence of a probe in the jack by measuring for a short between the two plates. I speculate that depending on how that's done exactly, if more than one socket is detecting low enough resistance under high humidity levels, that could cause the sensor circuits to go whacky. Whatever the cause, it needs to be ironed out since many regions in the world have regular seasonal high humidity levels and the meter becomes unusable when the alarm is going off.
Best Wishes,

 Saturation
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37742
  • Country: au
    • EEVblog
Re: U1272A update
« Reply #27 on: June 09, 2011, 11:20:18 am »
Agilent have now updated the firmware for the meter to fix the problem:
http://www.home.agilent.com/agilent/editorial.jspx?id=2055659&ckey=2055659&nid=-34618.956189.02&lc=eng&cc=AU

Looks like they fixed a bunch of other stuff too:
http://www.home.agilent.com/upload/cmc_upload/All/version2_00_history.txt

Quote
1) “-“ sign intermittently show up in ACV for ZLow function while pressing Trig Hold/Auto Hold button
2) No showing “OL” when overloaded in Manual Range while Auto Hold is enabled
3) Frequency range does not change range accordingly
4) Bar graph does not show voltage reading in Peak Hold function
5) Showing invalid reading while pressing Trig Hold in Peak Hold function
6) Incorrect reading after pressing “Hz%ms” in MaxMin mode
7) In ACDC mA function, “OL” is shown even change to higher range in MaxMin mode
8) In Capacitance function, “OL” or incorrect reading is shown when change to higher range in MaxMin mode
9) Bar graph does not reflect the real measurement while in Continuity 300ohm range
10) Secondary display changed while pressing “Dual” in Auto Hold mode
11) Auto Hold function is disabled when single press one “AutoHold” button in Auto Hold mode
12) Temperature unit F-C or C-F display should show “F” only or “C” only in both 1st and 2nd display
13) Null not functioning together with other Math function
14) Continuity flash alert as fast as beep alarm
15) “ZLOW” display control by EEPROM location F4
16) Remove Frequency filter code for LPF
17) Add TRIG SCPI command
18) Issue fixed for ADC Initialising Failure

For those with effected units, they offer two options:
1) A replacement unit with the updated firmware
2) A FREE USB interface cable (to download the firmware)
This applies until Nov.

They are also fixed all their stock on the shelves with the authorised reps as quickly as possible, who have all been altered to the issue. And all factory stocked units have already been fixed.

And I just noticed there is a service guide too:
http://cp.literature.agilent.com/litweb/pdf/U1271-90020.pdf

Dave.
« Last Edit: June 09, 2011, 11:27:18 am by EEVblog »
 

Offline saturation

  • Super Contributor
  • ***
  • Posts: 4787
  • Country: us
  • Doveryai, no proveryai
    • NIST
Re: U1272A update
« Reply #28 on: June 09, 2011, 11:26:01 am »
Awesome.  Again, it shows the value of a more insightful reviewer, as there were reviews elsewhere far earlier on other forums than on eevblog that did not discover any issues.

I've never seen a DMM with a user based firmware upgrade, this is a first for me, but its very welcome.  Have you?

The problems I've been experiencing do point to firmware, because the stability, casing and build are fairly solid .  I've recent had issues with its capacitance range, but will redo the accuracy checks with the new firmware.  I'm impressed how fast this has occurred from the onset of the complaints to the fix.


Agilent have now updated the firmware for the meter to fix the problem:
http://www.home.agilent.com/agilent/editorial.jspx?id=2055659&ckey=2055659&nid=-34618.956189.02&lc=eng&cc=AU

Looks like they fixed a bunch of other stuff too:
http://www.home.agilent.com/upload/cmc_upload/All/version2_00_history.txt
....

And I just noticed there is a service guide too:
http://cp.literature.agilent.com/litweb/pdf/U1271-90020.pdf

Dave.
« Last Edit: June 09, 2011, 11:31:59 am by saturation »
Best Wishes,

 Saturation
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37742
  • Country: au
    • EEVblog
Re: U1272A update
« Reply #29 on: June 09, 2011, 11:28:31 am »
I've never heard of a DMM with a firmware upgrade, this is a first for me, but its very welcome.  Have you?

I just updated my Gossen Energy firmware.
But it does require a $450 USB cable!  :o

Dave.
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37742
  • Country: au
    • EEVblog
Re: U1272A update
« Reply #30 on: June 09, 2011, 12:46:25 pm »
Here is the official Service Note for the problem, and details on how users can contact Agilent for their FREE USB cable!

http://cp.literature.agilent.com/litweb/pdf/U1272A-02.pdf

Could Agilent have possibly given a better response in a more timely manner?

Compare that with the response we got from Fluke on the GSM phone problem...  >:(

Dave.
 

Offline johnboxall

  • Supporter
  • ****
  • Posts: 652
  • Country: au
  • You do nothing, you get nothing.
    • Books, services and more:
Re: U1272A update
« Reply #31 on: June 09, 2011, 01:05:55 pm »
Fantastic news, and thanks to you Dave for being a squeaky wheel  :)

Offline Mechatrommer

  • Super Contributor
  • ***
  • Posts: 11653
  • Country: my
  • reassessing directives...
Re: U1272A update
« Reply #32 on: June 09, 2011, 01:31:59 pm »
1) A replacement unit with the updated firmware
2) A FREE USB interface cable (to download the firmware)
number 2 looks like a clear winner to me if i own one agilent. later when they have rev3 fw, no need to send back. and can play around with the cable :P

Compare that with the response we got from Fluke on the GSM phone problem...  >:(
a silence may means "oh we cannot fix that hardware problem unless you send back the unit and repaired by us, which is uneconomical, from our side point of view" :P
« Last Edit: June 09, 2011, 01:35:19 pm by Mechatrommer »
Nature: Evolution and the Illusion of Randomness (Stephen L. Talbott): Its now indisputable that... organisms “expertise” contextualizes its genome, and its nonsense to say that these powers are under the control of the genome being contextualized - Barbara McClintock
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #33 on: June 09, 2011, 02:49:07 pm »
squeaky wheel  :)

Yes he use to do that quit often   :D ( we have many solid proofs about that ..)
If he continue doing that, eventually he will enter and in the Guinness Book too.  ;) 

About Agilent I can not hide how happy I am about them.
And about Fluke multimeters, they do have very respectable people working for them,
in their development department, but they are not magicians.  :)

I feel privileged and happy, that due my activity at the EEV forum,
I managed to have direct communication paths with people who care about my opinion,
in both castles.  ;)


About the Agilent move, like free  PC connectivity modem, for every DMM with older firmware,
this is an true surprise !!
I had never see something like that, ever.

Call me romantic, but it feels to me like that there is a new sheriff, in multimeters town.  :) LOL
 
 

 
 





 

Offline Floyo

  • Regular Contributor
  • *
  • Posts: 90
  • Country: nl
Re: U1272A update
« Reply #34 on: June 09, 2011, 03:26:07 pm »
It is good to hear that Agilent have fixed this problem so quickly. And it's a very nice gesture of them to give all the affected people the option to receive free means of resolving the problem, service as it should be.

Now just some extra "beep" options so I don't disturb the whole house whilst working with my dmm, and possible even an pdf describing the serial commands so people can write their own logging applications (maybe even in embedded systems). If Agilent makes these things available as well, they might have a winner on their hands, both for the professional and the advanced hobbyist market. 
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37742
  • Country: au
    • EEVblog
Re: U1272A update
« Reply #35 on: June 09, 2011, 10:29:56 pm »
1) A replacement unit with the updated firmware
2) A FREE USB interface cable (to download the firmware)
number 2 looks like a clear winner to me if i own one agilent. later when they have rev3 fw, no need to send back. and can play around with the cable :P

Yeah, unless you have a beat up meter you want replaced, the USB cable is by far the better deal!

Dave.
 

Offline mikeselectricstuff

  • Super Contributor
  • ***
  • Posts: 13748
  • Country: gb
    • Mike's Electric Stuff
Re: U1272A update
« Reply #36 on: June 09, 2011, 11:16:21 pm »
Could Agilent have possibly given a better response in a more timely manner?
Compare that with the response we got from Fluke on the GSM phone problem...  >:(
Dave.
And the response of Rigol to..... well pretty much everything :D
Youtube channel:Taking wierd stuff apart. Very apart.
Mike's Electric Stuff: High voltage, vintage electronics etc.
Day Job: Mostly LEDs
 

Offline elCap

  • Regular Contributor
  • *
  • Posts: 109
  • Country: jp
Re: U1272A update
« Reply #37 on: June 10, 2011, 04:33:09 am »
This great news! Very well done, Agilent! This is how the customers should be handled. I don't mind a few bugs if I get a free gift (of value!) when they are fixed :)

Here is the official Service Note for the problem, and details on how users can contact Agilent for their FREE USB cable!

http://cp.literature.agilent.com/litweb/pdf/U1272A-02.pdf

Could Agilent have possibly given a better response in a more timely manner?
 

Offline Freighternut

  • Regular Contributor
  • *
  • Posts: 54
  • Country: nz
Re: U1272A update
« Reply #38 on: June 10, 2011, 11:14:53 am »
Great news indeed, and it is certaily great to see a responsive supplier at last.

Dave....how far of is the review. I am hanging off from buying one until I get your good oil on the usability and
any quirks.

Save the earth - Travel the world by Container Ship! Almost zero carbon foot print
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37742
  • Country: au
    • EEVblog
Re: U1272A update
« Reply #39 on: June 10, 2011, 11:28:17 am »
Dave....how far of is the review. I am hanging off from buying one until I get your good oil on the usability and
any quirks.

As usual, not much of an idea, sorry. There are at least couple of other theory/project blogs I want to do first, a Tek scope teardown, FX-888 review too.
I need a USB cable first, but can likely get one from Triosmartcal around the corner.
But as always, I work on whims!

Dave.
 

Offline zaoka

  • Frequent Contributor
  • **
  • Posts: 376
  • Country: us
Re: U1272A update
« Reply #40 on: June 10, 2011, 09:36:01 pm »
Any chance to get new Hameg scope?
 

Offline Freighternut

  • Regular Contributor
  • *
  • Posts: 54
  • Country: nz
Re: U1272A update
« Reply #41 on: June 11, 2011, 12:44:18 am »
Hi dave,

Whim's are good...in lieu of a full review, any likes/dislikes with the U1272A so far?

Keep up the good work, I am enjoying the blogs and forum immensely.

Roger...
Save the earth - Travel the world by Container Ship! Almost zero carbon foot print
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37742
  • Country: au
    • EEVblog
Re: U1272A update
« Reply #42 on: June 11, 2011, 01:04:18 am »
Whim's are good...in lieu of a full review, any likes/dislikes with the U1272A so far?

There is a lot to like (assuming the firmware does fix the issues as claimed)
I like the 7/times sec updating and all the extra features over the Fluke.
I like the feel and the construction, esp holding in the hand.
I like the extra digit of resolution on mV.
The display digit size isn't as good as the fluke, but you do get all the extra display functions, and the contrast is better than fluke, particularly from above the display.
I don't like the Off position not being at the end, can't just slam the knob around to shut off.
The backlight is very nice, brighter than fluke.
I hate the LOUD power on beep which can be disabled, but then you don't get button beeps which I like.
Continuity speed is slower than the fluke, enough to be a slightly annoying, but much faster than many other meters.
I like the battery choice and fuse compartment, better than fluke.
I don't like the removable probes tips, they are annoyingly long IMO.

Hard to compare the two really, as the Agilent obviously kills the Fluke for functionality and bang-per-buck, but it hasn't proven itself long term yet.
If you are after just a basic meter that you could bet your life on, and will likely still work in 15-20 years, the fluke would still be my pick. The Fluke has a lifetime warranty, bigger and less "busy" display, removable rubber holster, and proven reliability and parts support.

If it was my own money on the line though, I'd buy the Agilent I think, just for the bang-per-buck...  :-\

Dave.
« Last Edit: June 11, 2011, 01:25:54 am by EEVblog »
 

Offline Freighternut

  • Regular Contributor
  • *
  • Posts: 54
  • Country: nz
Re: U1272A update
« Reply #43 on: June 11, 2011, 02:07:48 am »
Thanks Dave!

Very helpfull!

Is the dual display any good for viewing DC voltage and ripple simultaneously. Most supplies that I work on are low voltage Dc
(Max 40 Volt) but I need to observe ripple down to about 5 - 10 mv.

Roger...
Save the earth - Travel the world by Container Ship! Almost zero carbon foot print
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37742
  • Country: au
    • EEVblog
Re: U1272A update
« Reply #44 on: June 11, 2011, 04:14:27 am »
Is the dual display any good for viewing DC voltage and ripple simultaneously. Most supplies that I work on are low voltage Dc
(Max 40 Volt) but I need to observe ripple down to about 5 - 10 mv.

Yes, the secondary AC display can display 0.0000V whilst displaying autoranged DC volts.
But the 2nd AC display does seem to get fooled initially and autoranges to 0.000V display and does not range back down to give you the extra digit (when probing a 9V battery).

Oh, and the autoranging is not very fast either.

Dave.
 

Offline ruessel

  • Newbie
  • Posts: 6
Re: U1272A update
« Reply #45 on: June 11, 2011, 11:02:07 am »
Thanks Dave!

Very helpfull!

Is the dual display any good for viewing DC voltage and ripple simultaneously. Most supplies that I work on are low voltage Dc
(Max 40 Volt) but I need to observe ripple down to about 5 - 10 mv.

Roger...


Yes, it does that. But it takes almost forever until you get a stable reading for the AC ripple.
My beat up, 20 year old Fluke 45 dual display bench multimeter is way faster!
Also have a new Agilent U3401, bought as a replacement for the old Fluke 45 -but this one is almost as slow as the U1272.
I guess in this respect the old Fluke 45 is hard to beat.

Helmut
 

Offline saturation

  • Super Contributor
  • ***
  • Posts: 4787
  • Country: us
  • Doveryai, no proveryai
    • NIST
Re: U1272A update
« Reply #46 on: June 11, 2011, 11:46:49 am »
New buyers assume no further bugs exists after this firmware release, but that remains to be seen.
« Last Edit: June 11, 2011, 12:45:30 pm by saturation »
Best Wishes,

 Saturation
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37742
  • Country: au
    • EEVblog
Re: U1272A update
« Reply #47 on: June 11, 2011, 12:00:51 pm »
New buyers assume no further bugs exists after this firmware release, but that remains to be seen.

They fixed an awful lot of bugs in this release!
You gotta wonder what else is lurking in there...

Dave.
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #48 on: June 11, 2011, 01:40:04 pm »
New buyers assume no further bugs exists after this firmware release, but that remains to be seen.

They fixed an awful lot of bugs in this release!
You gotta wonder what else is lurking in there...

Dave.


I like this hypothesis, if it turns as true,
that would translate to, that I will have my chance to find it.  ;D  ;D

And another thing, Agilent had proven with their actions,
that they do listen.
Lets help them out with full thrust.

Lets give a lesson to the big boys who hides behind their big castles. 
Primarily the Internet it is an communication tool, and marketing and advertising comes second or last to me.

Who cares about robustness and fame, when one DMM has twenty functions,
and the selection of those options was wrongly made ?

Some people they seek even today the champion DMM in aspects like speed at AC voltage or Continuity speed,
I care for the productivity factor mostly.
 
     

 

Offline PetrosATopic starter

  • Frequent Contributor
  • **
  • Posts: 625
  • Country: us
Re: U1272A update
« Reply #49 on: June 12, 2011, 03:05:50 pm »
Has anyone actually been able to update their firmware for the 1272 yet? I keep getting a "file not found" screen every time I click on the Firmware Update Software link (top link of the four). The other three files are there, but not that one :(
I miss my home I miss my porch, porch
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #50 on: June 12, 2011, 03:11:42 pm »
I test it, and download  it before 5 seconds .

http://www.home.agilent.com/upload/cmc_upload/All/U1272A_V200_0525.zip
 

Offline PetrosATopic starter

  • Frequent Contributor
  • **
  • Posts: 625
  • Country: us
Re: U1272A update
« Reply #51 on: June 12, 2011, 03:18:16 pm »
That file just contains the firmware itself (U1272A_V200_0525.AG), not the updater software. There's supposed to be an .exe file with the program that actually loads the firmware that I can't get.
I miss my home I miss my porch, porch
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #52 on: June 12, 2011, 05:08:07 pm »
Ok, if so you need the firmware utility  U1272A_Firmware_Update_Software_20100908.zip   26MB.

Yes it looks down for the moment, and by looking the time zone of Malaysia,
it will take at list 12 hours until to be restored.

I would like to help, but I have no bandwidth so to upload one file like that. ( GSM internet)
Sorry.   
 

Offline saturation

  • Super Contributor
  • ***
  • Posts: 4787
  • Country: us
  • Doveryai, no proveryai
    • NIST
Re: U1272A update
« Reply #53 on: June 13, 2011, 02:20:01 pm »
There is some weirdness in the upgrade process, so maybe Agilent is fixing the installer.  You may want to wait for them to repost it.   If you want it now pm me, the installer is 26MB.

If you already have the serial port driver installed, its fine, but the upgrade installer does not install the driver, and appears to hangs if it can't find it, nor gives you a message it can't find it or its not installed.

Once the serial port driver is properly installed the upgrade software finds the DMM in a snap.   The .net framework based software is slow, is > 10 minutes for the PC side on Win7, but once OK its very fast for the firmware upgrade to the DMM, even with the default baud rate.   The install software or the driver have no bugs, otherwise.

The meter works more reliably but there are residual issues, none of them are critical, that remain unchanged.  For example, The artifact 1kVAC appears by manipulating the switch with 15VDC being measured, then slowly disappears.  Other minor issues that exists in this model aren't in the 1252a or the 87V.




That file just contains the firmware itself (U1272A_V200_0525.AG), not the updater software. There's supposed to be an .exe file with the program that actually loads the firmware that I can't get.
Best Wishes,

 Saturation
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #54 on: June 15, 2011, 10:33:41 pm »
 

Offline Floyo

  • Regular Contributor
  • *
  • Posts: 90
  • Country: nl
Re: U1272A update
« Reply #55 on: June 21, 2011, 09:31:40 am »
I just got my cable in the post this morning, DHL high priority air mail. I sent the email June 11th, package was sent the 17th. and it got here today. Not bad especially becouse I'm used to waiting a month or more for stuff to arrive from Asia.

Updating the firmware was a piece of cake, so now the hunt for new bugs is on :P
 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: U1272A update
« Reply #56 on: June 21, 2011, 09:55:20 am »
Floyo did you use the first software updater utility ? or the latest ?  and what windows do you use ?

( The truth are that both versions was identical , the latest for some reason it uses an older driver than the first,
but comes uncompressed , and you can see the driver uninstaller as separate file)

I did not had to use the update,  my DMM came all ready updated from the factory.   
Welcome on the data logging club  ;)
« Last Edit: June 21, 2011, 10:46:31 am by Kiriakos-GR »
 

Offline Floyo

  • Regular Contributor
  • *
  • Posts: 90
  • Country: nl
Re: U1272A update
« Reply #57 on: June 21, 2011, 10:01:38 am »
I think its the latest one, it has to separate folders with the drivers one normal and one for 64 bit vista. But I didn't install the drivers because I already have an usb-rs232 converter for µC projects that uses the same chip. I use win7.

 

Offline saturation

  • Super Contributor
  • ***
  • Posts: 4787
  • Country: us
  • Doveryai, no proveryai
    • NIST
Re: U1272A update
« Reply #58 on: June 23, 2011, 06:39:18 pm »
Accuracy tests on the 1272a are summarized on these graphs.

The test 1272a unit meets or exceeds the Fluke 87V in all functions except capacitance.  The minimum capacitance it reads accurately is 1nF despite what the manual says; beyond it the accuracy drops markedly and one 1272a , refuses to read below 1nF.  For comparison, other known DMM of equal or lesser accuracy are also co-measured to show the tests accuracy is expected of each DMM's published specification.



1272aC is factory calibrated.  1272a was shipped off cal, and was adjusted by me in DC ranges.  However, further tests show its also off in AC ranges compared to the factory calibrated unit.

This chart is a rank scale of accuracy for each basic DMM function, smaller bars are better.  The Agilent 1252a is surprisingly the most accurate of all the DMMs tested, in all ranges ~ almost a factor of 2, see the raw data chart.  The 1252a's one weakness is noise in high megaohms scale reduces accuracy reading.

The 'busy' chart is a cumulative error for every tested scale; shorter bars are better.  The lines follow 2 primary sources of inaccuracy, the capacitance scale on the 1272a and the Vac scale for all DMMs.

« Last Edit: June 23, 2011, 06:47:53 pm by saturation »
Best Wishes,