Hello John , thanks for posting your findings ..
With your findings , my speculation looks that it does work,
and this is a great find .
Well yesterday , I did an very in-depth reading of all the manual.
( so to find every detail about the U1272 and compare it with what I own )
I did check your graph (No1), and it looks that you got there as Max, three sample rates per second.
At this point I like to say, that I do
Not see things like to be a race or something.
The data logging are a similar process with the Min/Max/Average function.
And found lots of details
mostly from Fluke users manual about it .
It looks that Fluke and Agilent agree , that in the given time of 250uS ( between of any taken sample),
the result are reliable.
In faster sample rates , the possibility of an error grows by some percentage .
Fluke are very clear that at Min/Max/Average there is an 100mS rate, and 250uS at
Peak/Min/Max/Average,
(I just found how to get in there yesterday )

And now I know that I must use
Peak , when I need maximum reliability .
Unfortunately Agilent has no details at all about the Min/Max/Average sample rate ,
All that it says is that at their own
Peak function, the sample rate are also 250uS.
There is no need to say , that your findings about the Com port speed,
its an great info .. 100mS .. that's great , and so the connection speed gets out ,
from the equation of the negative factors. ( Does not slowdown things )