| stj:
it doesnt have anything to do with ajustment protocol. why would you mess up the calibration if it may just be a cracked solder joint? changing the factory calibration should be the very last option, not the first! |
| Bobson:
What made you believe it was the first option? |
| stj:
it's the first thing people think of when they get something. :( |
| sstenks:
I have the same problem with UT181a with AC voltage it measures 205V on other multimeters 225V ..etc I tried to calibrate the AC voltage but it does not accept the calibration, everything else works DC, A, mA ... how to solve this? Thank you in advance |
| sstenks:
Hello, Just to let you know that I found a way to successfully calibrate the AC voltage on the UT-181A I will post the details soon and hope it helps whoever needs it. I still can't calibrate the 1000V range but I'm working on it. Thank you all once again for the constructive consideration of the problem, they really complicated it 🙂😊😁 |