Products => Test Equipment => Topic started by: AwaisN on August 23, 2018, 06:10:24 pm
-
Hi,
I have UNI-T UT202 AC clamp meter. Few days back when I tried to measure AC voltages and current, it gave wrong readings as compared to Fluke clamp meter. I need calibration procedure for this UT202 and will appreciate if get schematic also. Temperature readings also need to be calibrated. DC voltages and resistance measurements seems ok.
Thanks and regards,
Awais
-
Are you sure you are using the meter correctly?
There are some issues with thermocouplers and multimeters: https://lygte-info.dk/info/DMMThermocouplers%20UK.html
The clamp do not measure true rms, could that be the problem?
-
Yes. Before it was giving correct readings.
-
How far off are the readings? Provide examples please.
-
Here are some measurement results for 50 Hz line voltages and AC Induction Motor Current.
From Fluke 322 Clamp Meter: 226.7 Vac, 116.5 Vac and 3.25 Amperes
From UNI-T UT202: Fluctuate b/w Overflow and random reading against 226.7 V from Fluke,
106.7 Vac against 116.5 Vac from Fluke
1.889 A against 3.25 A from Fluke