Author Topic: UNI-T UT61E Multimeter teardown photos.  (Read 766111 times)
Offline Nisei
Re: UNI-T UT61E Multimeter teardown photos.
« Reply #850 on: August 23, 2015, 02:20:44 pm »
Correct, TL175
I got them with my Fluke 177. Sold the 177 but kept the TL175 leads. I really like them.
Offline mos6502
  • Country: aq
« Reply #851 on: August 23, 2015, 09:54:33 pm »
Only issue is if you have pitted and/or dirty leads, which they usually are even when new out of the box. Cleaning them with contact spray helps a lot.
Had the same problem. Tried all kinds of chemicals. Acetone. Alcohol. Window cleaner. It helped a little, but it wasn't 100%. Then I tried this:
(A small brass wire brush.)
Scrub scrub scrub. Result: PERFECT. When I rub the probes together in buzzer mode, the resistance never even goes above 0.2 Ohms. Before, when did the same thing, the resistance would spike to hundreds of ohms. The formerly shitty Uni-T probes are a joy to use now.
« Last Edit: August 23, 2015, 09:56:16 pm by mos6502 »
for(;;);
Offline naughtilus
  • Country: mk
  • If it ain't broke, fix it anyway.
« Reply #852 on: August 24, 2015, 12:06:22 pm »
Great idea with the brush.  :-+ The stock leads are not great, but certainly not terrible like the ones found on 830 DMMs.
...or is it?
Offline ocw
« Reply #853 on: September 07, 2015, 02:59:05 pm »
I haven't seen much information on the sensitivity of the frequency counter in the UT61E other than it's not the best.  The manufacturer rating of 220 MHz along with the meter's banana plug input is a bit of a oxymoron.  220 MHz and the normal banana plug leads don't get along.  So, I made a banana plug to BNC shielded cable.  I fed the output of a Ramsey COM3010 service monitor directly to the meter when its 0 dBm output level was adequate.  When needed, I added a 16 dB amplifier.  It's 40 mW output was inadequate above 101 MHz.  Even starting below that frequency, greater frequency resolution is generally required dictating use of a frequency counter anyway.
I measured the sensitivity of the frequency counter portion of the UT61E as:
        UT61E
FREQ    COUNTER SENSITIVITY
1 Hz            >3.43 V   > +24 dBm
2 Hz           171 mV   -2.3 dBm
10 Hz   105 mV   -6.6 dBm
60 Hz   105 mV   -6.6 dBm
1 kHz   105 mV   -6.6 dBm
10 kHz   117 mV   -6.0 dBm
100 kHz   117 mV   -6.0 dBm
0.5 MHz   165 mV   -2.6 dBm
1 MHz   165 mV   -2.6 dBm
4 MHz   200 mV   -1.0 dBm
12 MHz   243 mV   +0.7 dBm
25 MHz   243 mV   +0.7 dBm
30 MHz   243 mV   +0.7 dBm
35 MHz   243 mV   +0.7 dBm
40 MHz   267 mV   +1.5 dBm
50 MHz   385 mV   +4.7 dBm
60 MHz   385 mV   +4.7 dBm
65 MHz   554 mV   +7.9 dBm
70 MHz   777 mV   +10.8 dBm
75 MHz   873 mV   +11.9 dBm
80 MHz   873 mV   +11.9 dBm
85 MHz   1.10 V   +13.8 dBm
100 MHz   1.20 V   +14.6 dBm
101 MHz   1.35 V   +15.6 dBm
Given the low resolution of the counter I didn't go into detail confirming its accuracy, but I never saw a sign of any significant error.
I also measured the frequency response of the AC voltmeter portion of the meter.  Since I'm more concerned about its validity for frequency response measurements, I measured the response +/- 1 dB.  I found that to be 5 Hz to 68 kHz.  The response was down 1 dB at 5 Hz and up 1 dB at 68 kHz.  At 60 Hz I measured its AC voltage accuracy as +0.155% (well within its 0.8%+10 rating).
Offline mos6502
  • Country: aq
« Reply #854 on: September 07, 2015, 03:57:55 pm »
ocw, that's very interesting. I think above a few MHz, the capacitance of the leads becomes a big issue. Have you tried using extremely short leads? Or a BNC to banana Adapter?
Here's a schematic of the UT61E: http://docdro.id/EbNqMPt
As you can see, the input circuitry of the frequency counter is very simple (the stuff around R35).
I'm having some issues with the frequency input of my UT139C, but it did measure up to 27MHz correctly, even though it's only spec'd to 10MHz.
Have you tried the duty cycle measurement? It would be interesting to know up to what frequency that would be halfway accurate.
for(;;);
Offline ocw
« Reply #855 on: September 07, 2015, 05:39:54 pm »
I must have made a good banana plug to BNC cable--I didn't see a significant change when instead of using the cable, making it clumsy and using a banana to BNC adapter instead.
I'm not sure what duty accuracy you are looking for, but this is what I measured on the UT61E:
Freq    Waveform           D Cycle
100 MHz   Sine           5.0%
90 MHz   Sine           6.7%
80 MHz   Sine           15.3%
70 MHz   Sine           25.0%
60 MHz   Sine           30.7%
50 MHz   Sine           34.9%
40 MHz   Sine           41.3%
30 MHz   Sine           42.1%
20 MHz   Sqr 50%   44.8%
15 MHz   Sqr 50%   45.9%
15 MHz   Sqr 32%   35.0%
15 MHz   Sqr 68%   58.6%
10 MHz   Sqr 50%   46.9%
5 MHz   Sqr 50%   48.0%
1 MHz   Sqr 50%   49.4%
The duty cycle reading being displayed on the UT61E depended on the input voltage.  If that voltage was just enough for an accurate frequency reading, the duty cycle reading would be low.  Double the voltage and the duty cycle reading would be much more accurate.  Since I was unable to increase the RF voltage for the higher frequency readings, the accuracy of their duty cycle readings seen was compromised by that.
Offline mos6502
  • Country: aq
« Reply #856 on: September 07, 2015, 06:10:29 pm »
Great job, thanks! Of course, the measurment of the duty cycle depends on the trigger level, so it's really only meaningful for square wave signals. But it looks like the duty cycle measurement is useful to at least 1MHz.  :-+
Duty cycle measurement is really useful for a lot of things. E.g. if you wrote a multitasking scheduler for your microcontroller, you can set a pin at the beginning of your task and reset it at the end of your task, and by measuring the duty cycle you can directly see how much of the CPU time that task consumes in percent.
for(;;);
Offline ocw
« Reply #857 on: September 08, 2015, 02:33:55 am »
I forgot about the higher output power available from my HP 3200B VHF generator.  Taking it to a metered output of +23.5 dBm/224 mW/3.346 V (1.7 dB beyond its rating) the UT61E was able to measure its output frequency up to 225 MHz.  Uni-Trend hasn't over-rated it.  But, there's a reason why their specs don't show its counter's sensitivity.
Offline xZero
« Reply #858 on: September 11, 2015, 04:34:15 pm »
Is this one still worth consideration in 2015? Or is there something better for same price?  :P
Thanks in advance!
I was on cheap one, and finally gonna move to something better with PC connection.
I work primarily with electronics and informatics (hardware).
Offline mos6502
  • Country: aq
« Reply #859 on: September 11, 2015, 05:02:12 pm »
Thanks in advance!
Do you need the PC connection? If not, I'd go with the newer UT139C. Much better construction, faster display, actually has input protection, temperature measurement, real fuses and has a backlight. Only 6,000 counts, but unless you have specific requirements, that's perfectly adequate.
for(;;);
Offline xZero
« Reply #860 on: September 11, 2015, 05:37:39 pm »
Thanks for the reply.
But yes, I do need PC connection, and that's the feature I've been missing on all multimeters I had (cheap ones).
Looking from my point of view, by my needs, UT61E seems like perfect choice, I've just wondering can I get something better/newer with same capabilities for the same price?
Since UT61E is few years old, I expected price to be lower as well, however it seems that it holds it price since then.
Offline mos6502
  • Country: aq
« Reply #861 on: September 11, 2015, 06:50:31 pm »
22,000 counts and PC interface is pretty much unparalleled at this price point. So yeah, buy it. Forgot to mention, the burden voltages are also worse on the UT61E vs. the UT139C. Really, I would get the UT139C as well. Each has features that the other one is missing. I'm thinking of getting a UT61E myself to complement my UT139C.
for(;;);
Offline xZero
« Reply #862 on: September 11, 2015, 07:45:57 pm »
Since I'm limited with price at the moment I'm gonna buy UT61-E as it have what I need and later I may go for UNI-T UT71D. Having spare multimeter can't be bad.
Or, what if I buy Mastech MY64 as first hand, since I'm out of any right now... It's pretty cheap, and seems to be decent temporarily.
« Last Edit: September 11, 2015, 07:54:28 pm by xZero »
Offline mos6502
  • Country: aq
« Reply #863 on: September 11, 2015, 07:59:50 pm »
Wat. That one doesn't have a PC connection. ???
Anyway, this one is cheaper and better than the Mastech you mentioned:
for(;;);
Offline Lightages
  • Canadian po
« Reply #864 on: September 11, 2015, 08:02:23 pm »
Just remember that the UT61E,and the UT71D do not meet their CAT ratings. Use them for electronics use and they are are good buys. The UT139C is a much better built and safer meter but has less features.
I have zero confidence in anything made by Mastech and would recommend the Uni-T models before anything made by Mastech.
« Last Edit: September 12, 2015, 03:26:39 am by Lightages »
Offline mos6502
  • Country: aq
« Reply #865 on: September 11, 2015, 08:14:12 pm »
Bingo!  :-+
Another option would be the DT-4000ZC:
http://www.ebay.de/itm/Digitek-DT-4000ZC-TekPower-TP4000ZC-Data-Logging-Multimeter-with-Temperature-/200912705694
Pretty decent meter, very cheap and it has RS232.
for(;;);
Offline xZero
« Reply #866 on: September 12, 2015, 01:29:01 am »
Yeah I noticed that, but for price I didn't expected it.
Thank you for your effort guys, I'm realy happy to see al that responses and I had took a look at every of DMM recommended, and so I decided to go for UT61E for now, and after when I get some more money for some better one. It's good to have more than one.
As I said I'm electronic hobbyist, and sometimes I "mess" up with household installations and such voltages, but my very first generic DMM (DT-SOMETHING) had cost me only 10USD and It lasted for 5 years, and I did use it to measure amperage up to 20A, measure DC/AC voltage up to 200/500V for electronic tasks, for car electronics etc and it did all pretty well. It had bad probes, and unusable capacitance tester. I killed it by measuring voltage from DC/AC converter (12VDC -> 90VAC) because I forgot to change from DC to AC. That one was almost visually/feature identical as Mastech MY64 with exception that protection cover was yellow.
« Last Edit: September 12, 2015, 01:41:10 am by xZero »
Offline niclasf
    • foogadgets
« Reply #867 on: September 22, 2015, 11:12:51 am »
I have done a combination of many of the UT61E modifications.
I have used a PIC16F688 to add more modes to the yellow and blue function buttons.
I have also added a LPF mode, but I am not sure how I can verify that it works. ::)
Offline NicolasCalisto01
  • Country: cl
« Reply #868 on: October 10, 2015, 02:59:31 pm »
Yes, is a nice multimeter hack
Offline groff
  • Country: cs
« Reply #869 on: October 28, 2015, 06:09:39 pm »
Yes, very good work. :-+  I found your c file on foogadgets but can I get hex file? Can you attach it on foogadgets or on this forum?
Offline niclasf
    • foogadgets
« Reply #870 on: October 29, 2015, 06:35:45 am »
Now you can download it in the download section in bitbucket, https://bitbucket.org/foogadgets/ut61e-modification/downloads/ut61e-mod.hex
Enjoy!
Offline groff
  • Country: cs
« Reply #871 on: October 29, 2015, 05:51:13 pm »
Thanx  :clap:
Offline jbuszkie
« Reply #872 on: October 30, 2015, 09:35:59 pm »
Do you have pictures of the mods?  I tired to go to the forum post in the foogadgets blog but it doesn't seem to work.
Thanks!
Jim
Offline niclasf
    • foogadgets
« Reply #873 on: November 09, 2015, 11:26:51 am »
I have uploaded the pictures from the modification, https://bitbucket.org/foogadgets/ut61e-modification/downloads
Offline MrMacman2u
« Reply #874 on: November 15, 2015, 05:07:01 am »
I have used a PIC16F688 to add more modes to the yellow and blue function buttons. (popped out the link -MM2U)
Excellent... no... Outstanding work niclasf! :-+
I just bought this meter and this is, literally, an even better modification than the one I was in the middle of making using an ATTiny 85 since I wasn't going to bother breaking out the LPF!
Now I need to step it up and figure out how to hack in the temperature function using the ADF pin since that the chip supports temp too.
I'm working on other hardware upgrades as well including setting in metal inserts to use machine screws for the battery compartment and case, solid input jack upgrades to replace the cheapo splits, and installing the missing MOV's.
I'll document all the changes and post them here.
« Last Edit: November 15, 2015, 06:03:47 pm by MrMacman2u »