Hello, EEVblog forum users,
I acquired UNI-T UT81C scopemeter some time ago. It served me well so I decided to do a teardown.
Other forum members have done a teardown/review of a more popular version UT81B (lower bandwith and sample rate spec.) - search forum if you would like to compare.
TeardownI know that picture quality is not the best. If forum members are interested in something specific, I could redo the pictures.
UT81C scopemeter before teardown.

Teardown was an easy process - unscrew few screws and open the bottom of the meter.



Meter is constructed from 3 PCBs (buttons PCB + DMM PCB + scope PCB) connected via 0.254 mm headers and display board.

Overall, there is lack of shielding - see huge holes on top/bottom. Case sides are high and they are not shielded at all. Shielding contact is too stiff/sharp metal piece, which made a hole in the shield.



Input jacks are soldered directly to the PCB at one point. They are split type, bent metal. Certainly not as great as solid jack construction with stress relief.

Note that UT81C has not the best input protection (glass fuses, overvoltage protection could be improved).

DMM board has five crappy adjustment pots, processing PCB has one adjustment cap. I do not know exact purpose of these, so left those unadjusted.

Soldering quality of SMD parts are OK, but though hole stuff is messy (unclean, but solder joints are good enough). Note that processing board has 2 ICs that were manually soldered (resoldered) - maybe these chips differ between UT81A/B/C models.




DC input jack is recessed very far from the edge of the PCB, presumably to limit a leverage of DC jack and increase robustness.

Scopemeter works as before after assembly.
ReviewScopemeter specifications
- Bandwith: 16 MHz, 1 channel, 80 MS/s, no AC coupling (only DC coupling), 50ns-5s (1-2-5 steps)
- Voltage: 20mV-500V, current: 200µA-1mA/20mA-100mA/2A-5A
- Trigger: auto/normal/shot, raising/falling (auto only with no trigger from 200 ms and up)
- Screen: 160*160 monochrome LCD with backlight
- Probes: Multimeter probes or 1x/10x probe with adapter
- Other: 10 screenshots memory, isolated USB serial via IR adapter, 4 AA batteries or DC adapter.
Notes:
- 4*AA batteries are very hard to insert - space for batteries seems to be 0.5-1.5 mm too short
- Different time settings seems to have different performance (in terms of frequency, level response - possible mismatch between ranges).
- Input impedance is all over the place - check table below. Almost impossible to cal. probe and make absolute measurements if signal is in a wide range and your switch between DMM/scope modes.
- Oscilloscope range impedance, equivalent DC/AC range on multimeter:
- 5.017MOhm 200V-500V -> 1000V/750V DC/AC
- 5.017MOhm 20V-100V -> 400V DC/AC
- 5.043MOhm 2V-10V -> 40V DC/AC
- 5.410MOhm 200mV-1V -> 4V DC/AC
- 20.525MOhm/7.215MOhm 20mV-100mV -> 400mV DC/AC
- Multimeter, frequency/duty mode impedance:
- 10.507MOhm frequency/duty -> 1V
- Double impedance (~10MOhm 200mV-500V, 20.525MOhm 20mV-100mV)(similar to DC range) on multimeter mode, but significant capacitance added
- Voltage mode has good rise and fall time on "Voltage mode", but "Frequency" mode has terrible rise and fall time.
- Little shielding - mains 50Hz noise are easily visible. See unshielded areas and hole in the back shield. Poor shielding overall - no shield on the sides or battery compartment. Mains hum easily couples in.
- UNI-T multimeter-BNC plastic holder has intermittent connection with UNI-T scope probe (ground doesn't connect properly to outer ground ring).
- DMM side of things is OK - accurate enough, but slow update on screen and slow continuity.
Advantages:
- USB interface works, basic functionality (streaming data only). Live screen viewing though USB communication, can save data, show current and saved waveforms.
- 80 MS/s is true spec. Can test high voltages on scope. Scopemeter is useful for a single channel low freq. measurements.
- Killer feature - you can measure current waveform directly with DMM probes only! On µA/mA/A ranges!
Disadvantages:
- Issues with input impedance and absolute measurements.
- Not enough shielding - mains hum shows up if you are not careful.
- Probe adapter can be better (try not to lose this small piece).
I could add additional comments or pictures if requested.