Products > Test Equipment
Upgrading Mainboard in Lecroy DDA-3000 (aka WavePro 7300a?)
darkstar49:
--- Quote from: BeckGyver on August 30, 2020, 12:41:09 pm ---I searched for the "Demo Mode" but could not locate. There are some Scope-Settings and I reset to those but had the same waveforms.
I assume these demo waveforms are stored data and not a generated waveform fed to the inputs. Therefore not a real test of the frontend.
--- End quote ---
The 'demo mode' gets activated when the XStream software can't access the acquisition board (when the interface board can't be accessed, the software won't even start).
To my knowledge, the only way to deactivate it, is via the service menu. Some have said that the demo mode might also be linked with the DEV option, but I've never seen anyone elaborate on this.
Mechatrommer:
--- Quote from: georgo on August 31, 2020, 09:09:07 pm ---Another thing I want to ask if the gerber/schematics for SIL151 hdmi to lvds boards from wldshy are available?
--- End quote ---
cant PM you further georgo, your inbox is full, here is the msg... (i think you need to post more to get more priviledge such as inbox space, not sure)
--- Quote from: PM me ---
--- Quote from: georgo on September 03, 2020, 02:31:28 pm ---good to know thanks for the tip - are the gerbers for the SIL151 PCB available on the Forum? - i did search but I most have overlooked them? Do you know on which page they are?
--- End quote ---
to be frank, i dont know what is SIL151 :-\ until you mentioned it, searching in eevblog also returns none. so i guess you have to study that outside eevblog, ie google, sorry.. looking back at your post, you've been sent chapter 3 and 4 whereas me, they only sent me chapter 5 (and lots of *lss files), but i guess chap 3 and 4 only the usual user's manual and device specification thats available in the net right? not really crucial for verification process (chapter 5) i guess. cheers.
--- End quote ---
BeckGyver:
Thanks for this input, that fits in the picture of missing/wrong voltages at the acquisition board. I will have a peak after i finish moving in the new house with a bigger bench and lab to get all tools on the table.
I will keep you in stream when I start working on this beauty again. I also got the adapters today and will be able to feed son signals as soon as I get this thing back working.
I also got an Error log of the Communication Test from the service menu:
--- Code: ---0 1.9.2020 19:43:29 Informational CommTest ------------------ Starting Tests -------------------
1 1.9.2020 19:43:29 Informational CommTest Processor Speed : 2870456832.000000
2 1.9.2020 19:43:29 Informational CommTest Device Driver Build # : 1605
3 1.9.2020 19:43:29 Informational CommTest PCI Card Revision : 16
4 1.9.2020 19:43:29 Informational CommTest ------ Checking Gigabit Links -----
5 1.9.2020 19:43:29 Informational CommTest Gigabit Signal detected on AAC Link
6 1.9.2020 19:43:29 Informational CommTest Gigabit Signal detected on CH1 Link
7 1.9.2020 19:43:29 Informational CommTest Gigabit Signal detected on CH2 Link
8 1.9.2020 19:43:29 Informational CommTest ----- Starting Readout Tests -----
9 1.9.2020 19:43:29 Informational CommTest Testing Readout on Channel 1
10 1.9.2020 19:43:34 Warning CommTest Channel 1 Readout Failed ! Error - 79
11 1.9.2020 19:43:34 Informational CommTest Testing Readout on Channel 2
12 1.9.2020 19:43:39 Warning CommTest Channel 2 Readout Failed ! Error - 79
13 1.9.2020 19:43:39 Informational CommTest Testing Readout on Channel 3
14 1.9.2020 19:43:44 Warning CommTest Channel 3 Readout Failed ! Error - 79
15 1.9.2020 19:43:44 Informational CommTest Testing Readout on Channel 4
16 1.9.2020 19:43:49 Warning CommTest Channel 4 Readout Failed ! Error - 79
17 1.9.2020 19:43:49 Informational CommTest ----- End Readout Tests -----
18 1.9.2020 19:43:49 Informational CommTest ----- Starting DMA Tests -----
19 1.9.2020 19:43:49 Informational CommTest Testing DMA Readout on Channel 1
20 1.9.2020 19:43:55 Informational CommTest DMA Readout Error. Error - 79
21 1.9.2020 19:44:00 Informational CommTest DMA Readout Error. Error - 79
22 1.9.2020 19:44:05 Informational CommTest DMA Readout Error. Error - 79
23 1.9.2020 19:44:10 Informational CommTest DMA Readout Error. Error - 79
24 1.9.2020 19:44:15 Informational CommTest DMA Readout Error. Error - 79
25 1.9.2020 19:44:15 Informational CommTest Transfered: 0.00MB at -1.#IND00MB/s
26 1.9.2020 19:44:15 Informational CommTest Testing DMA Readout on Channel 2
27 1.9.2020 19:44:20 Informational CommTest DMA Readout Error. Error - 79
28 1.9.2020 19:44:25 Informational CommTest DMA Readout Error. Error - 79
29 1.9.2020 19:44:30 Informational CommTest DMA Readout Error. Error - 79
30 1.9.2020 19:44:35 Informational CommTest DMA Readout Error. Error - 79
31 1.9.2020 19:44:40 Informational CommTest DMA Readout Error. Error - 79
32 1.9.2020 19:44:40 Informational CommTest Transfered: 0.00MB at -1.#IND00MB/s
33 1.9.2020 19:44:40 Informational CommTest Testing DMA Readout on Channel 3
34 1.9.2020 19:44:45 Informational CommTest DMA Readout Error. Error - 79
35 1.9.2020 19:44:50 Informational CommTest DMA Readout Error. Error - 79
36 1.9.2020 19:44:55 Informational CommTest DMA Readout Error. Error - 79
37 1.9.2020 19:45:00 Informational CommTest DMA Readout Error. Error - 79
38 1.9.2020 19:45:05 Informational CommTest DMA Readout Error. Error - 79
39 1.9.2020 19:45:05 Informational CommTest Transfered: 0.00MB at -1.#IND00MB/s
40 1.9.2020 19:45:05 Informational CommTest Testing DMA Readout on Channel 4
41 1.9.2020 19:45:10 Informational CommTest DMA Readout Error. Error - 79
42 1.9.2020 19:45:15 Informational CommTest DMA Readout Error. Error - 79
43 1.9.2020 19:45:20 Informational CommTest DMA Readout Error. Error - 79
44 1.9.2020 19:45:25 Informational CommTest DMA Readout Error. Error - 79
45 1.9.2020 19:45:30 Informational CommTest DMA Readout Error. Error - 79
46 1.9.2020 19:45:30 Informational CommTest Transfered: 0.00MB at -1.#IND00MB/s
47 1.9.2020 19:45:30 Informational CommTest ----- End DMA Tests -----
48 1.9.2020 19:45:31 Warning CommTest Cannot Read Micro Version. Communication to the AAC Link is broken!
49 1.9.2020 19:45:31 Informational CommTest AAC FPGA Version : 5.2
--- End code ---
Converter:
BeckGyver,
Can you show in which section of the system menu you receive this log?
BeckGyver:
As I am currently moving to a new house I can not fire up the scope to take some pictures from the Menu, but I will do so as soon as I have a bench available.
If I am right I used "Production Access" and was able to get access to the event logs. You can export the Logs to a thumbdrive and analyze them on a real PC. In addition I noticed that some logs were overflown (in 2013) and so I exported them and cleared it afterwards.
--- Quote from: nctnico on December 08, 2019, 03:33:39 pm ---You don't need the D drive at all. And you don't need to backup anything. The factory cal data is stored on the acquisition board itself en copied onto the hard drive automatically. The missing cal data is automatically created from the data gathered at the first time the oscilloscope is started.
A self calibration can be started from the service menu. Service center access is probably enough.
9472 Service Center
19641 Production Access
11011969 Kernel Access
22872201 Developer Access
--- End quote ---
I just wanted to share some news regarding possible adapters for the BMA adapter on the front. There are some adapters available from Aliexpress which cost around 10 bucks and look pretty nice. I ordered two of them:
https://de.aliexpress.com/item/32959937555.html
At the first glance they look pretty well made and I will have a test with the network analyzer at the new lab in some weeks.
I also attatched a picture of an original adapter LPA-SMA-A for reference ;-)
If someone is interested in building our own adapters please let me know. I also have some kind of galvanic isolation amplifier for low bandwidths in mind as well in the first place ( based on AMC1211x).