Use of Oscilloscope Probe x100 Vs x1000
« on: April 03, 2024, 09:12:58 pm »
Hello,
I want to buy one of the below oscilloscope probes in order to measure 230Vac and the output voltage of backlight inverter for driving CCFL lamps (~1000Vrms) but I am little confused when I am looking their specs.
I beileived that the difference between x100 Vs x1000 probe is that with x1000 probe can be measured higher voltages in respect with x100 probe, but on the specifications (page 7) of the two below probes the probes rating voltages are same. So, what is the purpose of the x1000 probe?

x100 probe
R&S®RT-ZH10   
1409.7720.02

x1000 probe
R&S®RT-ZH11   
1409.7737.02

« Last Edit: April 03, 2024, 09:24:40 pm by 2X »
 

Offline DaneLaw

  • Frequent Contributor
  • **
  • Posts: 950
  • Country: dk
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #1 on: April 03, 2024, 11:58:23 pm »
Why have you singled out these two x100 & x1000 probes from R&S?
Im not familiar with them, just briefly looked them up. they both seem to be 400MHz X100 (ZH10) around 740$ and the X1000 (ZH11) around 1300$ [Farnell, though incl. VAT.]
 
The following users thanked this post: egonotto, 2X, KungFuJosh

Offline Psi

  • Super Contributor
  • ***
  • Posts: 11327
  • Country: nz
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #2 on: April 04, 2024, 01:12:24 am »
There's a few things to consider with high voltages and probes.

There is the x10 x100 x1000 etc. Which controls what voltage the scope input will see, and source loading.   
If your scope input can handle 200V peak to peak then using a x10 probe means you scope wont see 200V until the probe is at 2000v p-p.  But obviously you have to take into account spikes on the signal you're measuring, as they will be above your signal voltage. So actually putting 2000V p-p on that x10 probe would be stupid, any spikes would overvolt your scopes 200v max input. So maybe you decide to only use it for measuring 1000V p-p so you have 1000V extra for spikes without damaging your scope. 

Or you might choose to use a 100x probe just because your signal is very sensitive to external loading. So you use a 100x probe even through the voltage of your circuit is low enough for even a 1x probe. In this case you may not be dealing with high voltages at all, so you don't want a probe with a huge amount of insulation that is big and cumbersome to use.

The second thing is the probes size/shape. How safe is it to use it on your circuit.
If you're working on tiny 0.5mA high voltage supply then maybe a typical looking small 10x probe would be fine for that at 1000V.   On the other hand, if you are measuring 1000V that can supply 100A then using that same probe would be crazy. You will see 10x or 100x probes that are huge with large red plastic insulation and others that look like normal probes.  It's not just the voltage the probe and scope can handle, it's the dangers from what you are measuring.

Just because a probe says 10x, 100x or 1000x you can't assume it's ok to use that probe to measure high voltage.
It maybe ok for the scope (voltage wise), but may not ok for you (safety wise)  :)
« Last Edit: April 04, 2024, 11:20:57 am by Psi »
Greek letter 'Psi' (not Pounds per Square Inch)
 
The following users thanked this post: 2X, KungFuJosh

Offline bdunham7

  • Super Contributor
  • ***
  • Posts: 9305
  • Country: us
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #3 on: April 04, 2024, 01:33:01 am »
So, what is the purpose of the x1000 probe?

It's a good question.  Both are 50M 7.5pF input and the only difference seems to be the final divider value.  They have peak or pulse ratings of 6kV so perhaps if you don't want to deal with the resulting 60V peak with the 100X probe (most scopes can't display that), you'll use the 1000X version instead. 
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 
The following users thanked this post: 2X

Offline Circlotron

  • Super Contributor
  • ***
  • Posts: 3498
  • Country: au
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #4 on: April 04, 2024, 02:27:31 am »
The voltage rise time of the circuit under test is something to watch out for too. I've blown x10 probes that are rated at 400V with a voltage that jumps to 400V in 1uS. Fast rise times can be a probe killer.
 
The following users thanked this post: 2X

Offline 2XTopic starter

  • Frequent Contributor
  • **
  • Posts: 554
  • Country: gr
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #5 on: April 04, 2024, 08:07:11 am »
Can you see what is the purpose of existing 2 probes x100 & x1000 with same characteristics? There is a positive advantage of x100 in contrast to x1000?
 

Offline Phil1977

  • Frequent Contributor
  • **
  • Posts: 769
  • Country: de
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #6 on: April 04, 2024, 10:35:48 am »
Please don't get me wrong, but as long as you need help to understand the probe characteristics I wouldn't recommend connecting the scope to mains or to any inverter...

What you probably need to do it safe is a high voltage differential probe. With that utility, you do much harder to destroy anything just by its connection. And though it costs around €200,- it seems cheaper than a decent scope or the hospital bill.

Trying to be constructive: Do you know "high voltage differential probes"? Do you want more information about them?

The thing with mains voltage and non differential probes just is that grounding the probe easily causes catastrophy. Even with an isolation transformer, sometimes even with a galvanically isolated scope. With the HV differential probe you really do not need to worry about ground, and even better their inputs are symmetrical. It´s the only way to measure common mode signals, and mains and inverters are always good for (unexpected) common mode distortions.
 
The following users thanked this post: egonotto

Offline bdunham7

  • Super Contributor
  • ***
  • Posts: 9305
  • Country: us
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #7 on: April 04, 2024, 01:58:39 pm »
Can you see what is the purpose of existing 2 probes x100 & x1000 with same characteristics? There is a positive advantage of x100 in contrast to x1000?

The 100X probe shows small signals 10X higher at the scope.  So if you needed the probe to work for 2kV and for 2V, at 2V the 100X probe gives you 20mV at the scope vs 2mV (which may not look as clear) for the 1000X.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 
The following users thanked this post: 2X

Offline 2XTopic starter

  • Frequent Contributor
  • **
  • Posts: 554
  • Country: gr
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #8 on: April 04, 2024, 07:26:26 pm »
Do you suggest this probes for measuring the output voltage of backlight inverter for driving CCFL lamps where is ~1000Vrms? Both probes they have Maximum Input Voltage 1000Vrms, so I beileive it is on the probe limits. If you agree on that and you know any other good quality brand with probes x100 & x1000 please suggest me.
 

Offline 2XTopic starter

  • Frequent Contributor
  • **
  • Posts: 554
  • Country: gr
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #9 on: April 04, 2024, 07:28:56 pm »
Can you see what is the purpose of existing 2 probes x100 & x1000 with same characteristics? There is a positive advantage of x100 in contrast to x1000?

The 100X probe shows small signals 10X higher at the scope.  So if you needed the probe to work for 2kV and for 2V, at 2V the 100X probe gives you 20mV at the scope vs 2mV (which may not look as clear) for the 1000X.

For similar reason because my oscilloscope when I bought it had only two probes x10 and I want to see a small signal without noise from the x10 attenuation I bougth x1 probe.
« Last Edit: April 04, 2024, 08:02:37 pm by 2X »
 

Offline DaneLaw

  • Frequent Contributor
  • **
  • Posts: 950
  • Country: dk
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #10 on: April 04, 2024, 09:40:27 pm »
For similar reason because my oscilloscope when I bought it had only two probes x10 and I want to see a small signal without noise from the x10 attenuation I bougth x1 probe.

What oscilloscope do you have?
 
The following users thanked this post: 2X

Offline 2XTopic starter

  • Frequent Contributor
  • **
  • Posts: 554
  • Country: gr
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #11 on: April 04, 2024, 09:48:55 pm »
For similar reason because my oscilloscope when I bought it had only two probes x10 and I want to see a small signal without noise from the x10 attenuation I bougth x1 probe.

What oscilloscope do you have?


https://www.rohde-schwarz.com/ae/products/test-and-measurement/oscilloscopes/rs-rtc1000-oscilloscope_63493-515585.html


 

Offline 2XTopic starter

  • Frequent Contributor
  • **
  • Posts: 554
  • Country: gr
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #12 on: April 05, 2024, 10:37:02 pm »
I found the below probe where have 2000V@3MHz & 1000V@7MHz (DC + AC Peak) max input voltage. Have anyone used it on 1000Vrms? I beleive his specs are good enough for these that I want to measure. Whatever product I have used from BK it was good quality.

BK Precision PR2000B 200 MHz 2 kV Oscilloscope High Voltage Probe
https://www.instrumentation2000.com/b-k-precision-pr2000b.html
https://www.instrumentation2000.com//media/pdf/bk-pr2000b-manual.pdf

Also, I found and the below but I beileve is too much.
https://www.instrumentation2000.com/cal-test-ct4432-oscilloscope-probe.html
« Last Edit: April 05, 2024, 10:43:52 pm by 2X »
 

Offline bdunham7

  • Super Contributor
  • ***
  • Posts: 9305
  • Country: us
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #13 on: April 05, 2024, 10:46:46 pm »
I found the below probe where have 2000V@3MHz & 1000V@7MHz (DC + AC Peak) max input voltage. Have anyone used it on 1000Vrms? I beleive his specs are good enough for these that I want to measure. Whatever product I have used from BK it was good quality.

That should work, but since you're in the EU, the very similar looking Testec TT-HV 250 might be easier to come by.

https://www.testec.de/en/products/product-categories/standard-probes/

How much current loading can your CCFL driver take?  These probes have good input impedances--100M and 4 or 5pF--but you need to make sure that even the small 5pF won't cause too much loading.  You'd need to know approximately the frequency and current capability of your CCFL system to calculate whether loading will be an issue.
« Last Edit: April 05, 2024, 10:49:15 pm by bdunham7 »
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 
The following users thanked this post: 2X

Offline 2XTopic starter

  • Frequent Contributor
  • **
  • Posts: 554
  • Country: gr
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #14 on: April 05, 2024, 10:57:58 pm »
I found the below probe where have 2000V@3MHz & 1000V@7MHz (DC + AC Peak) max input voltage. Have anyone used it on 1000Vrms? I beleive his specs are good enough for these that I want to measure. Whatever product I have used from BK it was good quality.

That should work, but since you're in the EU, the very similar looking Testec TT-HV 250 might be easier to come by.

https://www.testec.de/en/products/product-categories/standard-probes/

How much current loading can your CCFL driver take?  These probes have good input impedances--100M and 4 or 5pF--but you need to make sure that even the small 5pF won't cause too much loading.  You'd need to know approximately the frequency and current capability of your CCFL system to calculate whether loading will be an issue.



I found the BK precision probes on mouser electronics so it is not a problem the shipping. I have never used Testec probe or any other product of this brand... It has better, same or lower quality and safety from BK precision?
From searching on the internet and from the below waveform I found that the frequency is approximatelly ~50KHz and the current ~10mA, so I beleive the loading from the probes will be imperceptible.

https://www.researchgate.net/figure/Measured-CCFL-voltage-05-KV-div-and-current-33-mA-div-when-driven-by-a-PT-based-HF_fig2_3281120
« Last Edit: April 05, 2024, 10:59:52 pm by 2X »
 

Offline bdunham7

  • Super Contributor
  • ***
  • Posts: 9305
  • Country: us
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #15 on: April 05, 2024, 11:05:58 pm »
The BK/Cal Test probes are made in Taiwan, the Testec say they're made in Germany and I haven't heard anything bad about either.  Others have recommended the Testec TT-HV 250 but I haven't tried it.

At 50kHz and 1000VRMS, you'll have an input impedance that is dominated by the capacitor at about 636k and a resulting current of about 1.5mA (90 degrees leading).  Note that this does not mean the probe is dissipating 1.5 watts!  This may have some effect on the circuit but it isn't too bad and it may not mind at all since it gets most of the energy back--the 1.5mA is mostly recirculating.  Circuit loading can make it very difficult to accurately scope HF/HV signals! 
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 
The following users thanked this post: 2X

Online KungFuJosh

  • Super Contributor
  • ***
  • Posts: 5730
  • Country: us
  • TEAS is real.
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #16 on: April 05, 2024, 11:09:58 pm »
I have the BK probe, and it's good, but I've never gone over ~500V with it.
"Experience is something you don't get until just after you need it." - Steven Wright
Best Continuity Tester Ever
 

Offline 2XTopic starter

  • Frequent Contributor
  • **
  • Posts: 554
  • Country: gr
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #17 on: April 05, 2024, 11:19:31 pm »
The BK/Cal Test probes are made in Taiwan, the Testec say they're made in Germany and I haven't heard anything bad about either.  Others have recommended the Testec TT-HV 250 but I haven't tried it.

At 50kHz and 1000VRMS, you'll have an input impedance that is dominated by the capacitor at about 636k and a resulting current of about 1.5mA (90 degrees leading).  Note that this does not mean the probe is dissipating 1.5 watts!  This may have some effect on the circuit but it isn't too bad and it may not mind at all since it gets most of the energy back--the 1.5mA is mostly recirculating.  Circuit loading can make it very difficult to accurately scope HF/HV signals!

Thanks for the reply. One thing I didn't catch... Why the probe will not dissipating 1.5 watts ... it gets most of the energy back--the 1.5mA is mostly recirculating?
 

Offline 2XTopic starter

  • Frequent Contributor
  • **
  • Posts: 554
  • Country: gr
 

Offline bdunham7

  • Super Contributor
  • ***
  • Posts: 9305
  • Country: us
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #19 on: April 06, 2024, 12:00:04 am »

Testec TT-HVP-40
https://www.batronix.com/shop/measurement/probes/Testec-TT-HVP-40.html

Testec TT-HVP-2739
https://www.batronix.com/shop/measurement/probes/Testec-TT-HVP-2739.html

The first is useless for you (it is only a low BW tool) and the second is massive (literally) overkill.  I would use my Tek P6015s for this, but I happen to have them--I wouldn't buy them for this occasion.

As far as recirculating or reactive currents, that's just basic AC circuit theory that you should get acquainted with.  When you put a capacitor across a sinusoid AC source, there is in fact a current and it is also a sinusoid, but it is out of phase by 90 degrees (leading the voltage) and therefore half of the time energy is being transferred to the capacitor and half them time it is being transferred from the capacitor, less any losses of course. If losses are low, there is very little energy actually lost in the process.  If your signal is not sinusoidal, things get more complicated but the same basic principles apply.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 
The following users thanked this post: 2X

Offline 2XTopic starter

  • Frequent Contributor
  • **
  • Posts: 554
  • Country: gr
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #20 on: April 06, 2024, 07:46:17 am »

Testec TT-HVP-40
https://www.batronix.com/shop/measurement/probes/Testec-TT-HVP-40.html

Testec TT-HVP-2739
https://www.batronix.com/shop/measurement/probes/Testec-TT-HVP-2739.html

The first is useless for you (it is only a low BW tool) and the second is massive (literally) overkill.  I would use my Tek P6015s for this, but I happen to have them--I wouldn't buy them for this occasion.

As far as recirculating or reactive currents, that's just basic AC circuit theory that you should get acquainted with.  When you put a capacitor across a sinusoid AC source, there is in fact a current and it is also a sinusoid, but it is out of phase by 90 degrees (leading the voltage) and therefore half of the time energy is being transferred to the capacitor and half them time it is being transferred from the capacitor, less any losses of course. If losses are low, there is very little energy actually lost in the process.  If your signal is not sinusoidal, things get more complicated but the same basic principles apply.

I didn't combine the 90 degrees out of phase with you wrote. Thanks for your reply. Also, I found and the below that you told me.

https://www.batronix.com/shop/measurement/probes/Testec-TT-HV-250.html

 

Offline Phil1977

  • Frequent Contributor
  • **
  • Posts: 769
  • Country: de
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #21 on: April 06, 2024, 07:47:28 am »
That´s exactly the reason why I think you can get good results only with a differential probe.

With high voltage and high frequency, you have capacitors literally everywhere. If you use your probe without grounding, you have no idea about what your reference plane is doing - except you use two identical probes and do the differential math in the scope.

If you use probe grounding, you ether blow up your circuit or *massively* change the circuit by inducing a new low-z-connection where there is none in the original circuit. Galvanic isolation works only for DC or very low frequency AC, never for HF.

Please appologize if the first answer to your question was not very nice, but - without knowing special details - you just can not meassure with non-differential probes. And please don't think I didn't try to... :-X
 
The following users thanked this post: egonotto, tooki, 2X

Offline Phil1977

  • Frequent Contributor
  • **
  • Posts: 769
  • Country: de
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #22 on: April 06, 2024, 08:20:39 am »
PS: A possible answer why x1000-probes exist: If you work a lot on HV-stuff you may also have unintentional discharges. A x100-probe transfers by design 10 times as much of the transient to the scope than a x1000-probe, so if the datasheet values are exceeded anyhow, then it´s of advantage to protect the scope instead of the probe.

Maybe there are also some instruments that just don't have a -/-50 V input range.

But as long as you use everything in spec, I can't see a good reason why the probe attenuation factor should make a great difference.
 
The following users thanked this post: 2X

Offline 2XTopic starter

  • Frequent Contributor
  • **
  • Posts: 554
  • Country: gr
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #23 on: April 06, 2024, 08:25:21 am »
That´s exactly the reason why I think you can get good results only with a differential probe.

With high voltage and high frequency, you have capacitors literally everywhere. If you use your probe without grounding, you have no idea about what your reference plane is doing - except you use two identical probes and do the differential math in the scope.

If you use probe grounding, you ether blow up your circuit or *massively* change the circuit by inducing a new low-z-connection where there is none in the original circuit. Galvanic isolation works only for DC or very low frequency AC, never for HF.

Please appologize if the first answer to your question was not very nice, but - without knowing special details - you just can not meassure with non-differential probes. And please don't think I didn't try to... :-X

Thanks for you replies. I have made quite experiments with oscilloscope but fortunately I hadn't blow up something yet and because I know all the danger with the grounding I ask to many things here. In same cases the the groung was not so clear how is arranged on the PCB I have used two identical probes and I 've done the differential math in the oscilloscope but I had noise on the measured signal because the ground clips are not connectioned. I have used differential probe but I would like to risk and measure with "ordinary" probes. Always there is something to new even with risk.

I am close to select the two below for buying. They have good specs for what I want to experiment/measure.

Testec TT-HV-250
https://www.batronix.com/shop/measurement/probes/Testec-TT-HV-250.html

Testec TT-HVP-2739
https://www.batronix.com/shop/measurement/probes/Testec-TT-HVP-2739.html



« Last Edit: April 06, 2024, 08:29:40 am by 2X »
 

Offline tooki

  • Super Contributor
  • ***
  • Posts: 14716
  • Country: ch
Re: Use of Oscilloscope Probe x100 Vs x1000
« Reply #24 on: April 06, 2024, 09:50:40 am »
I have never used Testec probe or any other product of this brand... It has better, same or lower quality and safety from BK precision?
I have used many Testec probes and quite like them. Great quality at reasonable prices. The grips are a bit fatter than many probes, which I happen to like. I have some Testec at work and some at home.

Hameg (the test equipment company R&S purchased in order to break into the general test and measurement industry) shipped rebadged Testec with their scopes. (R&S now ships probes from another German manufacturer, PMK, which is more focused on high-bandwidth probes.)
 