bdunham7:

--- Quote from: tautech on July 16, 2021, 02:43:35 am ---Knock yourself out looking at properly terminated sources for square waves here:
https://www.eevblog.com/forum/testgear/show-us-your-square-wave/

--- End quote ---

I have my own--SDG2042X with PP215 probe and BNC adapter, Leo Bodnar fast edge with the same plus a BNC-to-BNC extension.  Each has issues, but not ringing.

You need to terminate the characteristic impedance of a transmission line, not a source.



Orion33:
So, I got my Siglent 1104X-E under the AKIP brand. Success :)
However, I can not calibrate the scope qualitatively. After each iteration, I get different DC offset values. Moreover, these values change over time.
What deviation is acceptable? Now after 2 hours of warm-up I get 400uV on the first channel
Is it possible to calibrate each channel individually using service commands?
If I activate options, for example, AWG, how do I then turn it back off for a warranty claim?

tautech:

--- Quote from: Orion33 on October 02, 2022, 12:22:46 pm ---So, I got my Siglent 1104X-E under the AKIP brand. Success :)
However, I can not calibrate the scope qualitatively. After each iteration, I get different DC offset values. Moreover, these values change over time.
What deviation is acceptable? Now after 2 hours of warm-up I get 400uV on the first channel
Is it possible to calibrate each channel individually using service commands?
If I activate options, for example, AWG, how do I then turn it back off for a warranty claim?

--- End quote ---
Which version firmware is installed ?

We can not see what you do, please post some screenshots with relevant menus visible, preferably in English.
Also, describe your signal connections if any and be sure to run the Self Cal only after ~30 minutes of warm up time.

Orion33:

--- Quote from: tautech on October 02, 2022, 07:27:59 pm ---Which version firmware is installed ?

We can not see what you do, please post some screenshots with relevant menus visible, preferably in English.
Also, describe your signal connections if any and be sure to run the Self Cal only after ~30 minutes of warm up time.

--- End quote ---
V6.1.37R8
All day long I just run self-calibration. The device worked for almost 9 hours. The result of the last calibration in the photo. For some reason, the device does not recognize the flash drive (exFAT). During the day, the best result was +/-80uV. After that, it was not possible to repeat it, the calibration of one of the channels is always unsatisfactory, as in the photo. The device works in a separate room where other devices are turned off.

tautech:

--- Quote from: Orion33 on October 02, 2022, 08:45:01 pm ---[For some reason, the device does not recognize the flash drive.

--- End quote ---
It needs be 8GB max FAT32 and 4k clusters.
Some brands of 16GB sticks work.

Your stick might be faulty or formatted incorrectly.
Get this fixed first.