Author Topic: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes  (Read 10416 times)

0 Members and 1 Guest are viewing this topic.

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Another new topic? I think it is good to separate out different topics so that we don't go too much off-topic.

This is about using (not buying, choosing, or comparing the SDS1000X-E Series Oscilloscope with other scopes).

It is about owning and living with a Siglent Scope for Home/Hobby/Enthusiast (HHE) Users e.g. someone who isn't necessarily owning/using the scope to derive their income. Other scopes are availableTM...

If most HHE OWners/Users are like me, they will evaluate their needs and buy only one of the scopes based upon how they want to use it. I would be the first to admit that I sometimes overestimate my needs and I'm fine with that - so I don't need to be reminded that I perhaps "got it wrong" and could have gone in a different direction by buying several pieces of equipment and "saving money". That's not to say that I'm not interested in how it goes with one of the other scopes in the series or how to do things in different ways. Because they are different and I can't afford to just buy them all, I am very interested in knowing more about all the scopes in the series that can be used (owned and lived with) and how they differ slightly, although I suspect that they are more similar than different.

The version I have is the SDS1000X-E 4-channel, 100MHz, Dual 1 GSa/s, Quad 500 Ms/s scope. Other versions have variations on this, with fewer channels or higher bandwidth, and these are the first (entry) series of digital oscilloscopes from Siglent. I'm not starting this topic because I have an issue with the scope and I don't have buyer's remorse. My impression is that Siglent does a nice job and creates an excellent piece of kit that will be very useful.

With the scope, you also get four 100MHz probes, the UK distributor includes a UK power cord and there is also a European cord in the box. There is a simple Quick Start Guide and on the web other resources available e.g. https://www.siglenteu.com/resources and https://www.siglenteu.com/service-and-support/ . The manufacturer's production (certification) date is from a few months ago and not surprisingly it has the latest firmware already installed. The products were initially released about three years ago and more recent versions have been released including the newer SDS1000X-U. I like the fact that the Firmware also has Release Notes, so you can decide if an update is really required and also learn about additional features they added. For my series of scope, there are 15 versions of Firmware (current and historical) but closer inspection shows that there are the current and four previous versions for this 4-channel version. There is also one version of the Operating System separately downloadable. The Firmware for the 2-channel was separated out and there is also Firmware for some of the Series' optional extras. As a "newbie" I don't want too much choice, but I do like to see the history, so what's there (in one place) is OK (good).

The main downloadable resources are a User Manual covering 1000X-E and 1000X-U and a Service Manual covering the 2-channel and 4-channel scopes. There is also a Programming Guide and some software which I haven't looked at. Being a bit "nerdy" I read the spec sheet in conjunction with the downloadable User Manual and the Service Manual before deciding to buy the scope. If you aren't "nerdy" then there is no need to look at the Service Manual but it has the advantage of letting you "look inside" without needing to remove the cover - mine had a holographic sticker that prevents prodding around! at first glance, the scopes all look similar but the position of the I/O in the User Manual hints at differences and the Service Manual shows that the layouts are a bit different e.g. placement of the PSU, the fan, and the mainboard being horizontal (2-channel) or vertical (4-channel).

Setting up the scope means plugging it in, attaching the four probes, and calibrating the probes to the scope. each probe comes with light-colored markers for the BNC and the probe to match the yellow, magenta, green, and blue controls, and display trace color. I calibrated the four probes after the unit was powered up with each probe attached and with the colored bands swapped on three of the probes (it may not matter) and with 10X selected on probe and scope. Worth noting is that the 4-channels are matched with four identical probes, so there is no immediate need to go out and (be compelled to) buy additional "stuff", at least when starting out.

Scope Calibration
The scope calibration is something that makes sense to do after the scope has reached a stable operating temperature and before the probe calibration is finalized. It isn't something to do last minute and the whole procedure is time-consuming. It's not simply a "press the button" and wait two minutes. It does a good job of calibrating, although it takes some time - best get a coffee or something.

Probe Calibration
The calibration waveforms to me look either square (when correctly setup) or needing adjustment with overshoot/undershoot (as supplied) that is symmetrical e.g. needing the same correction for the top or bottom. The "nerdy" part of me says this needs to be "spot on", so careful adjustment and close inspection (this might not be absolutely essential). I used the Auto Setup for each probe in turn and adjusted the Horizontal Timebase to show one cycle and a bit. Pressing the Channel button brings up soft menus above the horizontal buttons with the option to set the scope to 10X but also to invert the waveform. It may be obvious, but inverting the waveform also means that the Vertical adjustment can be used to zoom in and for the bottom of the waveform to remain in the same position. Put simply, there is no need to reposition the trace vertically. If all the channels are active, each channel overlays the previous one, which might be confusing, and "zooming in" is possible because there is no detectable DC offset. 

Attention - read what is on the screen
It may sound obvious, but it is worth reading what is displayed on the screen. An example is using the Utility Menu to check the Firmware level. When it says:
Quote
Press the "Single" button to exit
That is exactly what it means. Pressing buttons and rotary controls have no effect until you do press the "Single" button that is with the Trigger Controls on the RHS. Pressing any other button isn't the same! So, the scope hasn't "locked up" and there is no need to cycle the power button unnecessarily. Pressing and holding a button in for two seconds will bring up the Help. It's still a good idea to read the manual e.g. to know that this Help "feature" exists, but this helps too...

With audible feedback (or not)
Perhaps one of the first tasks on getting a new scope is to decide if the "audible" beep is something you want. I decided I could do without it. Putting the scope back to "defaults" can be changed as well. The scope can be restored to "defaults' which can be controlled by the User and there is also a factory default as well.  Because I want to return to "my default", I have the sound turned off and all four channels identically setup, 10X scope/probe, non-inverted DC - that's what made sense to me.

There have been previous topics for this family of scopes, so a lot has been said about it in the past and no need to repeat all that. For a new user, I thought it has some value to look at what you get today when starting out. FWIW, I'm a "newbie" to DSOs. So that's it for now.

Simon
 
The following users thanked this post: Performa01, tautech, TheBay

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #1 on: June 20, 2021, 01:36:47 pm »
This is about using (not buying, choosing, or comparing the SDS1000X-E Series Oscilloscope with other scopes).

It is about owning and living with a Siglent Scope for Home/Hobby/Enthusiast (HHE) Users e.g. someone who isn't necessarily owning/using the scope to derive their income. Other scopes are availableTM...

USB ports (front) and (back)
It's the advancing years - I may be getting a bit slow (and gray-haired), but...

My 4-channel SDS1104X-E scope has two USB ports (for memory sticks) one on the front and also one on the back. The internals (according to the Service Manual at least) have internal shielding that will (probably) affect the WiFi direction/range if I plug the (EEVblog suggested TL-WN725N) dongle into the rear USB port. It had me thinking that, although the two USB ports are intended to be used for a USB stick (front) and WiFi dongle (back), they are probably the same and are "interchangeable".

Whilst it makes sense to put the USB stick in the front so that it can be easily removed to take to a computer, for me it makes more sense to keep a USB stick in the back. So that it's always "ready" and I haven't got to search for one. Normal operation is to transfer info back to the computer... I have a small USB stick that is a similar size to the Siglent Wifi dongle, so I doubt the airflow from the fan is altered by switching the use around. So now, when I have something interesting to record, I don't have to hunt down a FAT32 formatted stick - because it's already with the scope. With some of the longer USB sticks, I know that having one in the front will someday be a mechanical (leverage) issue, and being "tucked away" around the corner helps.

LAN cable
I do have a 25' long LAN cable and I wanted to connect the scope (in one room) to my fixed computer (in another room across a hallway) using the cable. Whilst it works, I'm mindful that it's not exactly HSE compliant and the last thing I want is for someone else to nudge the scope or computer onto the floor (or trip up, btw). There may be arguments for using a LAN cable and not WiFi but for this, it makes sense to at least try the WiFi option before it's a trip hazard. If finding a small USB stick is a problem, trying to remember where a WiFi dongle is (they seem designed to be lost), means that I ordered a 2nd (inexpensive) TL-WN725N from Amazon.

Operation with a PC
I have an iMac for normal use and a dedicated laptop with Windows for another use. I don't like to mix the two. I will be using the Windows software for the scope and I use VMWare Fusion to run Windows Applications alongside Mac Apps. That's one of the things on my list to try and experience says it should work just as well on the Mac. Having access to the machine remotely via a web browser is a "neat trick" because it's not operating system dependent, but for using the scope, the Windows S/W seems to be essential. Overall, the scope greatly exceeds my needs and works great!

Simon
« Last Edit: June 30, 2021, 03:45:30 pm by SimonM »
 
The following users thanked this post: Performa01, tautech

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #2 on: June 20, 2021, 09:09:42 pm »
Random bits reply......

Installing options generated online by whatever means is so much easier when the scope is connected to a PC of any flavor if instead of EasyScopeX you use the inbuilt webserver. Copy/Paste> Enter, yes it's that simple !

To configure the scope for the webserver unless you know your LAN IP you need turn ON DCHP and let your modem assign an IP for the scope to use or better still have a known block of IP addresses reserved for instruments maybe even on their own subnet.
When LAN or a WiFi connection is enabled just enter the scopes IP into your PC browser and if all is well you will arrive at the scopes Welcome page where you can select Info, Instrument control and SCPI pages.
All that can be done in the scopes menus or front panel control can also be done remotely from the Instrument control page using the webserver.

USB A sockets
While these scopes have a very welcome 2 A type sockets sadly that's not enough at times to power the likes of USB drives, STB3, active probes, deskew fixtures and the SAG1021I AWG so some juggling does happen.
USB sticks, WiFi dongle and SAG1021I all must be plugged directly into the scope whereas other devices can be powered from any USB supply.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 
The following users thanked this post: SimonM

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #3 on: June 21, 2021, 09:58:40 pm »
Installing options generated online by whatever means is so much easier when the scope is connected to a PC of any flavor if instead of EasyScopeX you use the inbuilt webserver. Copy/Paste> Enter, yes it's that simple!

Instrument Control isn't simply READ ONLY
Thanks, it's a big help to know that the webserver should be interactive and that the Instrument Control isn't simply READ ONLY !!! (User Error).

I did try EasyScopeX - having installed the Visa runtime and Windows .net prerequisites. I did get it to briefly display a picture of the scope (the rotary controls are round) but it's certainly NOT easy. It takes control of the scope (locking it) and it can lose the link (forever). There may be occasions where EasyScopeX is a good thing - but I haven't found them - so I might uninstall Visa and EasyScopeX.  |O

I wouldn't disagree with some of these descriptions (from EEVblog):

"Has anyone got EasyScopeX working with their Siglent scope at all? Darned if we can get it working, pesky thing".

"Visa seems to work without any problems but EasyScopeX just crashes as soon as we try to run it".

Webserver allows the scope to be remotely used - it works very well
Using the webserver that is built-in to the scope provides a better user experience, by far! I used it previously to issue SCPI commands but my initial impression is that the Instrument Control is READ-ONLY. It continuously updates but the scope controls that can be enabled on the RHS are not visible and making it full screen doesn't help. The Siglent User's Manual describes setting up the webserver e.g. with DHCP but doesn't go to describe how to use it. Fortunately, there is a Siglent NA "demo" video that demonstrates "the click" that enables the controls to be made visible.

Thereafter it works just like using the scope, except the rotary controls are separated out as an LH rotate, click and RH rotate buttons. It really is like using the scope and everything works as you might expect. Using the scope directly is perhaps easier, but it would be OK for remote working.

Unlike EasyScopeX, which takes control of the scope, the real scope can be used at the same time when accessing via the webserver. The scope seems unaware that it isn't being used directly. Saving a picture, saves to the scope e.g. to a USB stick and I haven't found a way to save remotely - probably a limitation of using a browser.

Here are some shots from my Mac of remote operation. The FFT of the scope's 1K square wave updates in real-time and operation of the interface is like the real thing using the webserver (and not EasyScopeX). The scope identifies and labels the peaks, sorts them and what you get matches the theory e.g a square wave is a base with harmonics of 1/3 amplitude at 3x base, 1/5 at 5x, etc. Looking at the FFT, one might realize that the base frequency is what the scope FFT calculates: 954Hz vs real (see RHS displayed: 1KHz).

I could use the "Run Stop" button like the real scope, switch from displaying "peaks", to "markers" and adjust the points to allow measurements at the harmonic frequencies, and use show "Show Delta On" to add extra info to the table. Being remote has its disadvantages e.g. I can add channel 2, but I've not found a way to "remotely" attach the probe... :-BROKE

Finally, looking at the table, the graph, etc., and comparing it to a known square wave FFT, shows it has done an excellent job!

The only way I found to get the data back on the remote Mac, is to take a screenshot (like the two attachments). Clicking the print button is also able to save on the scope e.g. to a USB stick. In a day or so, I will have the WiFi set up, so no 25' cable through the house. I never realized, but modern, ultra-thin laptops no longer have ethernet ports. Perhaps future scopes will automatically include WiFi and not need a dongle?

Simon
« Last Edit: June 22, 2021, 09:14:46 pm by SimonM »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #4 on: June 21, 2021, 10:15:35 pm »
Saving a picture, saves to the scope e.g. to a USB stick and I haven't found a way to save remotely - probably a limitation of using a browser.
Quote
The only way I found to get the data back on the remote Mac, is to take a screenshot (like the two attachments). Clicking the print button is also able to save on the scope e.g. to a USB stick.
You missed the Save Waveform virtual button at the foot of the webserver page. That grabs a screenshot just like the blue Print button but instead drops it directly into your browsers Download folder. Same small file type PNG screenshot as when using the Print button.

Quote
Being remote has its disadvantages e.g. I can add channel 2, but I've not found a way to "remotely" attach the probe... :-BROKE
:-DD  :-+
When you find out the way to do that please post it here !  ;D
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 
The following users thanked this post: SimonM

Offline tttonyyy

  • Regular Contributor
  • *
  • Posts: 58
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #5 on: June 21, 2021, 10:57:36 pm »
EasyScopeX works fine for me. I can still take control of the scope front panel and ask the software to refresh regularly to reflect what changes. I can see this being useful when doing a Teams conf call to demonstrate something - much quicker than writing an image to usb stick, transferring to a pc and attaching to a chat.
 
The following users thanked this post: SimonM

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #6 on: June 22, 2021, 10:36:27 pm »
You missed the Save Waveform virtual button at the foot of the webserver page. That grabs a screenshot just like the blue Print button but instead drops it directly into your browser's Download folder. Same small file type PNG screenshot as when using the Print button.

I hadn't had a chance to explore these additional buttons:

1. Screen Save - saves a .png file locally;
2. Waveform Save - saves a binary file of the waveform;
3. Bin_2_CSV_Tool - provides a Windows utility to convert the binary file.

Screen Save was the virtual button I needed!

Bin_2_CSV_Tool - interesting but not universally executable
The Waveform Save virtual button also saves the waveform as a compressed binary file (13MB). I also used the Bin-2_CSV_Tool Windows Application to convert it to a big .csv file (260MB). I had mixed results with the Bin_2_CSV_Tool. On one Windows machine, it worked fine but on another, I had the usual cryptic Windows messages about compatibility problems. The Bin_2_CSV_Tool also includes (for reference) a Python script. Since I'm mostly on a Mac, using the script would be preferable to running the Windows Application under VMWare Fusion. When I tried to run it, Python complains about uninitialized variables (line 106). The file is interesting and since Siglent also publishes the format of the binary file, I may have another look at the script. Easier to work with it than starting over with just the spec.

Small .bin converts to an enormous .csv file
With the Windows App (on another machine), loading the generated .csv highlights just how much data the scope is working with, which when you are looking at the scope and adjusting it, you mostly take for granted - since it all works. One way to cut back the amount of data that the program produces is to limit the collection e.g. memory depth. There are some Siglent NA videos on how to use it with MS Excel.

MS "solution" to this "compatibility" problem, is to suggest reinstalling the application and trying again, with predictable results. I will probably get a better "solution" by undoing the prerequisites for the Visi program that were introduced - some old .net 2 & 3 libraries. One of the reasons why I try to avoid using Windows.

SDS1000X-E Service Manual - verifying scope's pass/failure criteria
The Service Manual has an interesting (to me) chapter on verifying the mainboard including how to set it up e.g. with/without a 50-ohm termination load (scope doesn't have a switchable input). Measuring the pass/failure criteria e.g. DC gain, offsets, time base, trigger delays, external triggers, noise floor, bandwidth, etc. everything being based on the SDS1202X-E or SDS1204X-E - one (only?) H/W difference being the input capacitance of the 4-channel scope.

I assume a 50-ohm termination is simply a 50-ohm added across the input terminals to ensure that equipment is currently set up to avoid reflections (distortion) and scaling issues (impedance mismatching).

Maintaining the sample rate
The Service Manual repeats a statement: "When a single channel per pair is active, that channel has a sample rate of 1 GSa/s". I had interpreted this to mean that if channel 1 was used without channel 2, the 1 GSa/s would apply to channel 1 when channel 3 and channel 4 were both active. You can see from enabling "dot" rather than "vector" with a Run/Stop (== stopped) and a 2ns/Div timebase, that all the channels drop back to 500 MSa/s (just join/count the dots). So, it is worth remembering that using channel 1 alone, channel 2 alone, channel 1 and channel 3 together, or channel 2 and channel 4 together, provides the highest sampling rate for the 4-channel scopes.

Simon
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #7 on: June 23, 2021, 11:21:02 am »
You missed the Save Waveform virtual button at the foot of the webserver page. That grabs a screenshot just like the blue Print button but instead drops it directly into your browser's Download folder. Same small file type PNG screenshot as when using the Print button.

I hadn't had a chance to explore these additional buttons:

1. Screen Save - saves a .png file locally;
2. Waveform Save - saves a binary file of the waveform;
3. Bin_2_CSV_Tool - provides a Windows utility to convert the binary file.

Screen Save was the virtual button I needed!

Bin_2_CSV_Tool - interesting but not universally executable
The Waveform Save virtual button also saves the waveform as a compressed binary file (13MB). I also used the Bin-2_CSV_Tool Windows Application to convert it to a big .csv file (260MB). I had mixed results with the Bin_2_CSV_Tool. On one Windows machine, it worked fine but on another, I had the usual cryptic Windows messages about compatibility problems. The Bin_2_CSV_Tool also includes (for reference) a Python script. Since I'm mostly on a Mac, using the script would be preferable to running the Windows Application under VMWare Fusion. When I tried to run it, Python complains about uninitialized variables (line 106). The file is interesting and since Siglent also publishes the format of the binary file, I may have another look at the script. Easier to work with it than starting over with just the spec.

Small .bin converts to an enormous .csv file
With the Windows App (on another machine), loading the generated .csv highlights just how much data the scope is working with, which when you are looking at the scope and adjusting it, you mostly take for granted - since it all works. One way to cut back the amount of data that the program produces is to limit the collection e.g. memory depth. There are some Siglent NA videos on how to use it with MS Excel.
Yes data files get big quick unless you set trigger conditions to control the size and just work with that capture.

Quote
SDS1000X-E Service Manual - verifying scope's pass/failure criteria
The Service Manual has an interesting (to me) chapter on verifying the mainboard including how to set it up e.g. with/without a 50-ohm termination load (scope doesn't have a switchable input). Measuring the pass/failure criteria e.g. DC gain, offsets, time base, trigger delays, external triggers, noise floor, bandwidth, etc. everything being based on the SDS1202X-E or SDS1204X-E - one (only?) H/W difference being the input capacitance of the 4-channel scope.

I assume a 50-ohm termination is simply a 50-ohm added across the input terminals to ensure that equipment is currently set up to avoid reflections (distortion) and scaling issues (impedance mismatching).
Yes Performance verification uses an external signal sources which are typically from a 50 Ohm source so therefore they must be correctly matched with a 50 Ohm pass through termination otherwise scope performance can't be accurately judged.

Quote
Maintaining the sample rate
The Service Manual repeats a statement: "When a single channel per pair is active, that channel has a sample rate of 1 GSa/s". I had interpreted this to mean that if channel 1 was used without channel 2, the 1 GSa/s would apply to channel 1 when channel 3 and channel 4 were both active. You can see from enabling "dot" rather than "vector" with a Run/Stop (== stopped) and a 2ns/Div timebase, that all the channels drop back to 500 MSa/s (just join/count the dots). So, it is worth remembering that using channel 1 alone, channel 2 alone, channel 1 and channel 3 together, or channel 2 and channel 4 together, provides the highest sampling rate for the 4-channel scopes.
4ch X-E are one of few in this class/price point that use 2 ADC's each 1 GSa/s so channel use management is required to maintain best sampling rates for when it might matter.
1&3, 1&4, 2&3 or 2&4 channel usage combinations all return 1 GSa/s max sampling rates by using one channel on each 1 GSa/s ADC whereas if using any 2 channels on either ADC reduces the max sampling rate to 500 MSa/s.
The user should also be aware each ADC has 14 Mpts of memory support that is halved when any 2 channels are used on one ADC however if just one channel is used on each ADC memory max depth remains at 14 Mpts/ch.

Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline DEV001

  • Contributor
  • Posts: 47
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #8 on: June 23, 2021, 04:58:19 pm »
Another tip for new 1104X-E users, RF-Loop's various posts on how the unit works and the linked post from Performa01 which contains several PDF documents that detail almost every function the 1104x-E has.

IMO, the official Siglent user manual while adequate doesn't provide nearly as much detail as the analysis below.

https://www.eevblog.com/forum/testgear/siglent-sds1104x-e-in-depth-review/

Just one of the many write-ups from RF-Loop on the 1104x-E.
https://www.eevblog.com/forum/testgear/siglent-sds1x04x-e-bodeplot-ii-(sfra)-features-and-testing-(coming)/


 
The following users thanked this post: SimonM

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #9 on: June 24, 2021, 12:05:54 am »
I really have to change my SSID just to configure and use the TL-WN725N WiFi dongle?
If I read it correctly (Jason from SiglentNA) advises: "XE firmware being incompatible with SSIDs and PSKs that have spaces or special characters" and the solution is just to not use them. :--

Perhaps I don't read the 802.11 standards correctly, because using English-like characters (other languages too) is a "convenience" but not a "requirement" and an SSID can be up to 32 octets of just about "anything", so "do not have spaces or special characters" doesn't begin to describe the problem...

Simon
« Last Edit: June 26, 2021, 10:24:44 am by SimonM »
 

Offline DEV001

  • Contributor
  • Posts: 47
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #10 on: June 24, 2021, 12:50:48 am »
I really have to change my SSID just to configure and use the TL-WN725N WiFi dongle?
If I read it correctly (Jason from SiglentNA) advises: "XE firmware being incompatible with SSIDs and PSKs that have spaces or special characters" and the solution is just to not use them. :--

Perhaps I don't read the 802.11 standards correctly, because using English-like characters (other languages too) is a "convenience" but not a "requirement" and an SSID can be up to 32 octets of just about "anything", so "do not have spaces or special characters" doesn't begin to describe the problem...

The TL-WN725N was supposed to arrive today. Tomorrow's task - explain why the home WiFi might be broken and how to "fix it" for a small array of WiFi widgets we have that was working just fine until now.

Simon

Another inexpensive option is to use an wifi router/extender/wireless bridge to authenticate to your wireless network and connect the scope ethernet port to one of the ethernet ports on the access point. You can use many inexpensive wifi routers to do this or flash OpenWRT/DD-WRT/Tomato, etc... to an older wifi router and you should be good to go until Siglent corrects the issue in firmware.

It isn't ideal but if all you have is WIFI access and you don't have the ability to change the SSID this may be an alternative option.

An example, this router looks like it supports wireless bridging for $19.98
https://www.amazon.com/TP-Link-TL-WA901ND-Wireless-450Mbps-Repeater/dp/B0857K4KVZ
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #11 on: June 25, 2021, 08:28:08 am »
Dropping a link to this post with an excellent illustration of LAN connection and explanation of DCHP principles for reliable connections:
https://www.eevblog.com/forum/testgear/siglent-ethernet-and-wifi-connection/msg3594083/#msg3594083
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #12 on: June 26, 2021, 11:11:31 am »
I really have to change my SSID just to configure and use the TL-WN725N WiFi dongle?
If I read it correctly (Jason from SiglentNA) advises: "XE firmware being incompatible with SSIDs and PSKs that have spaces or special characters" and the solution is just to not use them.

Hyphens are OK but spaces and a few other characters are not)
My SSID includes by default hyphens e.g. BT-XXXXXXX. That's OK and the scope and WiFi adapter connects OK. If I change it, spaces and some of the other restricted characters won't allow WiFi to connect. The router has its own set of restrictions e.g. spaces at either end of the SSID are excluded and well documented by the BT Help.

So, it turned out to be a non-event (for me) because hyphens are accepted and the scope WiFi connection works great. With five WLANs at home (and each defaulting to hyphens (-) in the SSID), that's good.

Scope and WiFi adapter seem to "play well" with a home router
I'm only using a "cheap" tp-link TL-WN725N adapter (about £8 from Amazon) that is restricted to the 2.4GHz band, with excellent range and speeds up to 150 Mbps that works perfectly. tp-link also has a dual-channel adapter TP-Link AC600 (Archer T2U Nano), but no real advantage using the 5GHz band that can be more "range restrictive" and some users report needing a separate Windows driver that's not required with a TL-WN725N adapter, so is most likely incompatible with the scope. Confirmed: TP-Link AC600 (Archer T2U Nano) is 2.4GHz/5GHz is definitely NOT compatible with the scope.

The shielding on the scope doesn't seem to be a problem e.g. it works well enough, with the adapter in the rear USB slot and any shielding by the metalwork of the scope doesn't obviously impact the performance, unduly.

Mostly I will try to use a LAN cable, but it's good to know that, if required, it also has a good WiFi range and I can now say "goodbye" to a 25' trailing LAN cable that goes between two rooms at home - that seems to be unpopular, for some unknown reason!

Can you keep a secret?
I noticed that the pre-shared keyword (PSK) secret password isn't quite so secret :-// - see the screenshot:

1230866-0

WiFi connect (and future reconnects after a power cut are good)
Once the WiFi is ON (connected) the scope will be ON (connected) the next time the scope is powered up. That's a good thing e.g. if the scope is configured to start after a power cut, then the WiFi connection is restored automatically, even if there is a delay as the router restarts and it keeps retrying the connection until it succeeds.  :-+

Simon
« Last Edit: June 30, 2021, 03:46:51 pm by SimonM »
 

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #13 on: June 29, 2021, 10:14:40 pm »
Adding SDG1032X Function Generator - phase angles
Having just bought the Siglent SDS1104X-E scope, and being pleased with it, I decided that I needed a 2 channel signal generator...

I have just bought and started to use a Siglent SDG1032X Function Generator.

We all make mistakes
Being somewhat new to having a signal generator, I connected Function Generator to the scopes using a BNC cable and a 50-ohm terminator at the scope end for channel 1.

I don't have a second BNC cable, so I have the second channel of the Function Generator with another 50-ohm terminator and a probe to scope 2nd channel.

With both channels on and triggering on channel 1, I can only "see" channel 2. The Function Generator defaults to locking both outputs.

Recognizing my "rookie" type mistake, I realize that the output from channel 1 is being exactly overlaid by channel 2.

How to fix a phase angle?
I now have changed the setup to have channel 1 and channel 2 of the Function Generator locked so that only channel 1 can be varied.

Both channel outputs are terminated with a 50-ohm terminator and the FG is set on both channels to 50-ohm.

I have two scope probes in 10x connected from channel 1 and channel 2 of the scope to the output of both 50-ohm terminators.

The Function Generator is now set at 10MHz with a sine wave. That's about the upper end of what I want to use the FG for, hence the purchase of an SDG1032X.

I see that the signal from channel 1 is being exactly overlaid (almost exactly) by channel 2.

Knowing that there are two groups of channels on the SDS1104X-E, I have started to use channel 1 and channel 3 as my two inputs. With both probes, are set to 10X and calibrated to the 1KHz square wave test. The scope has been reset to factory defaults and then all four probes set to 10X. The scope has been switched on for at least an hour and calibrated (takes about 5 minutes).

Triggering on channel 1 and connecting the "same input" to channel 2 where it's in phase:

1231664-0

Triggering on channel 1 and connecting the "same input" to channel 3 or channel 4 it's not in phase:

1231666-1
1231668-2

I can repeat this, triggering on channel 3 and connecting the "same input" to channel 4 where it's in phase:

1231670-3

Triggering on channel 3 and connecting the "same input" to channel 1 or channel 2 it's not in phase:

1231672-4
1231674-5

When I say "same input", I had initially set up the scope like the 1KHz setup. Thinking that the FG might not produce a signal that really was in phase, I connected both probes to the same 50-ohm output. That makes no difference. I have also switched the probes around to see if there is a phase difference between any of the four probes - they are the same.

So I have come to the conclusion that for the "same input":

Scope shows, (triggering on channel 1):

1. Channel 1 and channel 2 in phase;
2. Channel 1 and channel 3 not in phase;
3. Channel 1 and channel 4 not in phase (with the same phase difference at 10MHz).

Similarly, scope show, (triggering on channel 3):

1. Channel 3 and channel 4 in phase;
2. Channel 3 and channel 1 not in phase;
3. Channel 3 and channel 2 not in phase (with the same phase difference at 10MHz).

I'm not reliant on probes or the FG outputs to introduce the phase problem.

Would I be wrong to think that the scope needs some extra adjustment to bring everything back "in phase", or am I expecting too much?

NB I can show that the probes and the FG provide the same or an identical in-phase signal back to the scope.

Simon
« Last Edit: June 29, 2021, 10:19:11 pm by SimonM »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #14 on: June 29, 2021, 10:31:47 pm »
Would I be wrong to think that the scope needs some extra adjustment to bring everything back "in phase", or am I expecting too much?
Yes
Channel Phase locking is done in the SDG.....investigate the settings.  ;)

Another tip if/when you need to trigger on a channel but don't need to see that waveform is to hide that channel and that functionality is in every channel's menu on P2.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline StillTrying

  • Super Contributor
  • ***
  • Posts: 2850
  • Country: se
  • Country: Broken Britain
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #15 on: June 29, 2021, 10:51:14 pm »
Are you saying that the scope's CH 1+2 are not in phase with CH 3+4. :-//

If that have a look at SKEW while you're there. :)
.  That took much longer than I thought it would.
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #16 on: June 29, 2021, 11:15:03 pm »
Keep it simple and use two probes, one for CH1 and one for CH3 and display only CH1 and CH3.  Offset them a bit if it helps.  Connect both probes to the same channel of your SDG, eliminating all doubt as to phase.  Pick one to trigger from and then observe which one leads.  Leave the probes connected at the SDG and swap them at the scope.  If the same channel still leads, your scopes two acquisition units are not aligned, which is appears to be the issue.  If the channels change relative positions, the issue is in the delay characteristics of your probes. 

Now press CH1 and then Next Page.  The second button should read DESKEW and a time indication.  If this reads 0 and the channels are that far off, IMO your scope is defective, unless there is some way to calibrate the base deskew number.  You can press the DESKEW button and turn the small universal knob to adjust the position of the traces, but you should not have to do that with identical probes, certainly not for a correction of several nanoseconds.  Do check the DESKEW setting on all 4 channels to make sure someone hasn't set them this way.

FWIW, a quick check on my SDG2042X-E and SDS1104X-E show that the scope has a 1-3 skew of about 90ps and the sig gen about 270ps when the channels are copied but not coupled.

EDIT:  After a 30-minute warmup and a self-cal, my scope's 1-3 skew seems to have gone away altogether, best match is at zero now.  The sig-gen still is 270ps, which is about 12 degrees phase difference at 100MHz.  I'm not sure why that would be.

« Last Edit: June 30, 2021, 12:05:37 am by bdunham7 »
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 
The following users thanked this post: Performa01

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #17 on: June 30, 2021, 12:05:47 am »
Would I be wrong to think that the scope needs some extra adjustment to bring everything back "in phase", or am I expecting too much?
Yes
Channel Phase locking is done in the SDG.....investigate the settings.  ;)

Another tip if/when you need to trigger on a channel but don't need to see that waveform is to hide that channel and that functionality is in every channel's menu on P2.
Thanks. I can confirm:

I have channel 2 locked to channel 1. So I cannot change channel 2.

If I probe to channel 1 and channel 2 of SDG using scope channel 1 and channel 2, then they are both "in phase".

I can, of course, turn channels on and off, but I left them all on to show that the settings are the same and only switching inputs, to show a phase change.

I now have channel 2 of SDG off and both probes 1 & 3 on SDG channel 1 after a 50-ohm terminator. I get a phase shift.

So with two probes on the same source, I wasn't expecting any phase shift.

I can also replace probes 1 & 3 with probes 2 & 4 (all set to 10x) to show the same phase shift. Or swapping 2 & 4 to get the same phase shift.

So, I'm sure it's not the probes and there is only one channel of the SDG being used.

FWIW, SDG with one or two channels works very well - that's why I bought it!

Simon
 

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #18 on: June 30, 2021, 12:17:10 am »
Are you saying that the scope's CH 1+2 are not in phase with CH 3+4. :-//

If that have a look at SKEW while you're there. :)
Thanks.

Yes, I'm saying 1 & 2 of scope are in phase and 1 & 3 or 1 & 4 are not in phase.

I didn't know about the skew... until now.

I can get 1 & 3 to be in phase by adjusting the skew of channel 3.

The two traces appear to be completely aligned when it is set to -29.35ns.

I can also measure the skew difference, which averages about 400ps.

The skew can be set to a number by depressing the adjust button. There is also a default, or clear, that puts the skewback to 0.

On the function generator, I can increase the frequency e.g. from 10MHz, by 1MHz increments, to 15MHz.

When I change the frequency, the scope trigger still functions and the sine waves get closer together, as you would expect.

They also stay in phase (with skew set to -29.35ns) across different frequencies (10MHz, or lower and up to 15MHz).

If I switch to using channels 1 & 4, I can set the skew of 4 to stay in phase and it's also -29.35ns to get them in phase.

Simon
 

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #19 on: June 30, 2021, 12:31:11 am »
Keep it simple and use two probes, one for CH1 and one for CH3 and display only CH1 and CH3.  Offset them a bit if it helps.  Connect both probes to the same channel of your SDG, eliminating all doubt as to phase.  Pick one to trigger from and then observe which one leads.  Leave the probes connected at the SDG and swap them at the scope.  If the same channel still leads, your scopes two acquisition units are not aligned, which is appears to be the issue.  If the channels change relative positions, the issue is in the delay characteristics of your probes. 

Now press CH1 and then Next Page.  The second button should read DESKEW and a time indication.  If this reads 0 and the channels are that far off, IMO your scope is defective, unless there is some way to calibrate the base deskew number.  You can press the DESKEW button and turn the small universal knob to adjust the position of the traces, but you should not have to do that with identical probes, certainly not for a correction of several nanoseconds.  Do check the DESKEW setting on all 4 channels to make sure someone hasn't set them this way.

FWIW, a quick check on my SDG2042X-E and SDS1104X-E show that the scope has a 1-3 skew of about 90ps and the sig gen about 270ps when the channels are copied but not coupled.

EDIT:  After a 30-minute warmup and a self-cal, my scope's 1-3 skew seems to have gone away altogether, the best match is at zero now.  The sig-gen still is 270ps, which is about 12 degrees phase difference at 100MHz.  I'm not sure why that would be.

Thanks. I do have it all simplified down to only using one source which is SDG channel 1. channel 2 is off.

I was using both channels of the SDG to avoid having to put both probes into one BNC and hold them there.

With only one SDG output enabled, I can manually de-skew (is that a verb?) to align them.

I can swap the probes to show that it is independent of a phase shift in the probes.

I can switch off unused channels.

-29.35ns seems a lot to adjust something. A few 10's ps would be acceptable.

So, is there a way to "calibrate" the skew, so that the channels are in phase with a skew on all channels set to 0?

That's also how I would expect it to come from the factory.

There is also a "Quick Cal" in Utility, that is set to "On". Setting it to "Off" doesn't seem to make a difference.

Simon
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #20 on: June 30, 2021, 12:42:12 am »
-29.35ns seems a lot to adjust something. A few 10's ps would be acceptable.

Agreed, that sounds broken.  I have no idea how it is broken or what malfunction would cause that.  29ns is roughly the equivalent of 20 feet of coax.  It is possible that there is some calibration setting that is grossly off, so contact Siglent to see if they have some method of changing that or if they want to exchange it.

Quote
de-skew (is that a verb?)

delime, delouse, deskew....sounds good to me.
« Last Edit: June 30, 2021, 12:45:11 am by bdunham7 »
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline StillTrying

  • Super Contributor
  • ***
  • Posts: 2850
  • Country: se
  • Country: Broken Britain
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #21 on: June 30, 2021, 12:46:53 am »
It's odd, we've never seen that one before, with 4 near identical probes all the 4 channel skews on 0 should do.
Did you remove all 4 probes for the warmed up self-cal.
.  That took much longer than I thought it would.
 

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #22 on: June 30, 2021, 01:04:14 am »
It's odd, we've never seen that one before, with 4 near identical probes all the 4 channel skews on 0 should do.
Did you remove all 4 probes for the warmed-up self-cal?
Thanks. All the probes are the ones supplied and the type is pp510. For the warmed-up self-cal, I followed the instructions to remove all the probes. I would assume that, if the skew between different channel pairs can be set up and measured, then it can also be calibrated down to close to zero during a self-cal.

Simon
« Last Edit: July 02, 2021, 07:44:50 am by SimonM »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #23 on: June 30, 2021, 04:35:06 am »
Channel skew is specified on P9 in the datasheet:
Channel Skew <100 ps
https://www.siglenteu.com/wp-content/uploads/dlm_uploads/2020/12/SDS1000X-E_DataSheet_DS0101E-E04C.pdf

Whenever you strike something that doesn't look right use Default to bring settings back to factory then run Self Cal and revisit the measurement.
For BNC connections an unmatched pair of cables can often produce different results.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 
The following users thanked this post: SimonM

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #24 on: June 30, 2021, 05:41:28 pm »
Channel skew is specified on P9 in the datasheet:
Channel Skew <100 ps
https://www.siglenteu.com/wp-content/uploads/dlm_uploads/2020/12/SDS1000X-E_DataSheet_DS0101E-E04C.pdf

Whenever you strike something that doesn't look right use Default to bring settings back to factory then run Self Cal and revisit the measurement.
For BNC connections an unmatched pair of cables can often produce different results.

Thanks.

I am returning the scope to the factory setup and running Self Cal before revisiting the measurements.

The BNC connections are with/without the 50-ohm termination. Results are unchanged.

The cables are pretty well matched - they are the 4x PP510 probes. Swapping the probes around, the results are unchanged.

Here is what I did to simplify the setup (one channel of FG directly connected via two probes to the scope)

I have removed the 50-ohm terminator and setting the FG to Hi-Z. The setup is about as simple as it can be.

There is a setting for Quick-Cal on the scope. I hadn’t noticed it before and it was switched ON. I switched it OFF because the condition of varying temperature doesn’t apply.

I have reset the scope as follows:

1. Reset to factory defaults;
2. Use the scope Self Cal calibrate operation (takes about 5 minutes);
3. Press default button;
4. Set all four probes to 10X;
5. Set all four channels to the same 2V/div, 0V offsets;

Without anything connected, the four channels are overlain at 0V.

Set the FG to 10MHz and Hi-Z. Channel 2 is off. There is no 50-ohm terminator.

1. Channel 1 is triggered and displays the waveform with a probe point directly off the FG;
2. Channel 2 is added and the probe is connected to the same FG output. Both channels are in phase;
3. Move the 2nd probe from channel 2 to channel 3. The scope is out of phase;
4. Move the 2nd probe from channel 3 to channel 4. The scope is out of phase;
5. Move the 2nd probe from channel 4 back to channel 1. The scope is in phase;

When the scope is in phase the 2nd trace covers the first trace, otherwise, they are out of phase.

I can manually skew a second channel to overlay the first trace. -29.35ns is required.

NB without the 50-ohm terminator the results are as before, there is one difference that is expected: displayed trace doubles (Vp-p) because there is a change in the load impedance when connecting/disconnecting a second probe.

At 1MHz and probes set to 1X, timebase altered and scope input channels set to 1X, 2V/div, 0V offsets, it shows the same waveform.

Adding the second channel, overlays in phase, moving 2nd probe to channel 3 and then to channel 4 also shows the channels out of phase.

At 1MHz the amount of phase shift on the screen is reduced e.g. to 1/10. Skew to correct this is again best at -29.35ns.

Switching between probes doesn’t vary/alter the result.

Observations and conclusions

1. Removing the 50-ohm terminator doesn’t vary the out of phase outcome;
2. Varying which probes are used doesn’t vary the outcome;
3. Changing frequency e.g. between 10MHz and 1MHz and using probes 1X doesn’t vary the outcome;
4. The scope doesn’t show phase alignment between channels (1 & 2) with channels (3 & 4) after calibration;
5. The scope is remarkably aligned within the channels (1 & 2) or channels (3 & 4);
6. The scope can measure skew and offsets between channels, it just treats them separately e.g. scope is 2 + 2 channels.

rf_loop provided a review of the scope and his analysis of skew indicated a very high degree of alignment (see page 76):

1231987-0

NB the only mismatch is with attenuation (might be expected) and not with skew - each channel is very well aligned.

Simon
« Last Edit: July 01, 2021, 12:02:33 am by SimonM »
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #25 on: June 30, 2021, 05:46:53 pm »
6. The scope can measure skew and offsets between channels, it just treats them separately e.g. scope is 2 + 2 channels.

The scope is 2+2 channels as far as its ADC architecture goes, but they should be aligned.  This would be a huge issue if you were looking at 4 logic channels, for example.  Did you contact Siglent?
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #26 on: June 30, 2021, 06:07:11 pm »
6. The scope can measure skew and offsets between channels, it just treats them separately e.g. scope is 2 + 2 channels.

The scope is 2+2 channels as far as its ADC architecture goes, but they should be aligned.  This would be a huge issue if you were looking at 4 logic channels, for example.  Did you contact Siglent?
Agreed, they should be aligned and the spec for horizontal alignment is < 100ps across the scope (page 9 of the spec sheet). Agreed, it would be a huge issue for looking at anything that's 4 channel and would imply phase shifts that don't exist. It might also affect the decoding of a serial bus if the clock and data were split across channels 1 and 3.

As I get to know the scope, I'm starting to use channels 1 and 3 for dual channel use. This provides the full 1GSa/s sampling rate. If it's not aligned between the first and second pair of channels e.g. separate ADC chips, then the extra sampling can't be exploited. One of the reasons I bought the SDG1032X FG was the relative ease of evaluating bode plots. The outcome of this is to show frequency and phase shift, so I would like the measured phase to be accurate. I haven't tried it so far, but I would expect to use the best pair of channels e.g. 1 & 3, and not simply 1 & 2.

Simon
« Last Edit: July 02, 2021, 07:46:30 am by SimonM »
 

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4086
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #27 on: July 01, 2021, 07:24:18 am »
Channel skew is specified on P9 in the datasheet:
Channel Skew <100 ps
https://www.siglenteu.com/wp-content/uploads/dlm_uploads/2020/12/SDS1000X-E_DataSheet_DS0101E-E04C.pdf

Whenever you strike something that doesn't look right use Default to bring settings back to factory then run Self Cal and revisit the measurement.
For BNC connections an unmatched pair of cables can often produce different results.

Thanks.

I am returning the scope to the factory setup and running Self Cal before revisiting the measurements.

The BNC connections are with/without the 50-ohm termination. Results are unchanged.

The cables are pretty well matched - they are the 4x PP510 probes. Swapping the probes around, the results are unchanged.

Here is what I did to simplify the setup (one channel of FG directly connected via two probes to the scope)

I have removed the 50-ohm terminator and setting the FG to Hi-Z. The setup is about as simple as it can be.

There is a setting for Quick-Cal on the scope. I hadn’t noticed it before and it was switched ON. I switched it OFF because the condition of varying temperature doesn’t apply.

I have reset the scope as follows:

1. Reset to factory defaults;
2. Use the scope Self Cal calibrate operation (takes about 5 minutes);
3. Press default button;
4. Set all four probes to 10X;
5. Set all four channels to the same 2V/div, 0V offsets;

Without anything connected, the four channels are overlain at 0V.

Set the FG to 10MHz and Hi-Z. Channel 2 is off. There is no 50-ohm terminator.

1. Channel 1 is triggered and displays the waveform with a probe point directly off the FG;
2. Channel 2 is added and the probe is connected to the same FG output. Both channels are in phase;
3. Move the 2nd probe from channel 2 to channel 3. The scope is out of phase;
4. Move the 2nd probe from channel 3 to channel 4. The scope is out of phase;
5. Move the 2nd probe from channel 4 back to channel 1. The scope is in phase;

When the scope is in phase the 2nd trace covers the first trace, otherwise, they are out of phase.

I can manually skew a second channel to overlay the first trace. -29.35ns is required.

NB without the 50-ohm terminator the results are as before, there is one difference that is expected: displayed trace doubles (Vp-p) because there is a change in the load impedance when connecting/disconnecting a second probe.

At 1MHz and probes set to 1X, timebase altered and scope input channels set to 1X, 2V/div, 0V offsets, it shows the same waveform.

Adding the second channel, overlays in phase, moving 2nd probe to channel 3 and then to channel 4 also shows the channels out of phase.

At 1MHz the amount of phase shift on the screen is reduced e.g. to 1/10. Skew to correct this is again best at -29.35ns.

Switching between probes doesn’t vary/alter the result.

Observations and conclusions

1. Removing the 50-ohm terminator doesn’t vary the out of phase outcome;
2. Varying which probes are used doesn’t vary the outcome;
3. Changing frequency e.g. between 10MHz and 1MHz and using probes 1X doesn’t vary the outcome;
4. The scope doesn’t show phase alignment between channels (1 & 2) with channels (3 & 4) after calibration;
5. The scope is remarkably aligned within the channels (1 & 2) or channels (3 & 4);
6. The scope can measure skew and offsets between channels, it just treats them separately e.g. scope is 2 + 2 channels.

rf_loop provided a review of the scope and his analysis of skew indicated a very high degree of alignment (see page 76):

(Attachment Link)

NB the only mismatch is with attenuation (might be expected) and not with skew - each channel is very well aligned.

Simon

Not me. I think it was @Performa01

I do not remember/know with most new FW if there is still total deep reset.
During whole power up boot time  toggling very fast Math button. If it works you know it because you can see it have also reset calibration. If not work first time try again... As far as I remember factory reset (via menu) do not deep reset so that also all cal data is reset. (@tautech can perhaps confirm if this method is still there today)

And of course then calibration and -nothing- must be connected to input BNC's. Nothing - it really mean nothing. (how it affect it also depends scope model, in some old model whole internal cal signal come to BNC centers but do not know SDS1004X-E, I do not have this model available here for tests, these are in my homeland, not here in China)

Now if I understand right you have skev betwee (Ch1/Ch2 pair, (ADC1)) and (Ch3/Ch4 pair (ADC2))  roughly near 30ps.  Ande this is really lot of. Example 6 meter  RG316/U  have roughly 29ns travel time.


There is very poor simple and dirty method. "Wilderness and desert method. Just for bullet proof verify the problem is or is not."
ETA After seen your video:
Yoy can Connect your BNC T to channel 2 and trigger to CH2. Connect just one your coaxial from SDG to this T other side. Then take just very short copper wire and connect it from BNC T other side center as short as possible to CH3 input center. Now you have only one coax from SDG and very small, and roughly known,  signal travel time difference between CH 2 and Ch3 input.  Just 1kHz is signal where is fast edge is enough (square, pulse or 0% triangle). With this, because you do not have better things (special time matched cables and good splitter etc) available, you can infinitely proof if there is error between channels (when they are different ADC).

« Last Edit: July 03, 2021, 09:12:31 am by rf-loop »
I drive a LEC (low el. consumption) BEV car. Smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the wises gone?
 
The following users thanked this post: SimonM

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #28 on: July 01, 2021, 08:21:49 am »
I do not remember/know with most new FW if there is still total deep reset.
During whole power up boot time  toggling very fast Math button. If it works you know it because you can see it have also reset calibration. If not work first time try again... As far as I remember factory reset (via menu) do not deep reset so that also all cal data is reset. (@tautech can perhaps confirm if this method is still there today)
Those are enough as we should not share the other dealer secret Reset's we have.  ;)
However there still is Secure Erase in the Save/Recall menu used for deleting User settings such as LAN/WiFi and Reference waveforms.

Quote
If with this desert method there still is 10's of ns difference, if even ns difference, it is clear your scope have some severe problem what need Siglent service.
Soon to unpack my old prerelease SN#12 SDS1104X-E and check for channel skew and will post screenshots......

Channel skew tests.
Source SDG6202X square wave terminated with Tek feedthrough to Tee and to scope inputs with Siglent 1 GHz rated 1m BNC cables.
Dot and Vector displays.
« Last Edit: July 01, 2021, 08:50:50 am by tautech »
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #29 on: July 01, 2021, 08:19:36 pm »
Those are enough as we should not share the other dealer secret Reset's we have.  ;)
However, there still is a Secure Erase in the Save/Recall menu used for deleting User settings such as LAN/WiFi and Reference waveforms.

Thanks, I have used "Secure Erase" that removes user settings and the "Hard Reset" documented here: https://siglentna.com/operating-tip/oscilloscope-hardware-reset/. "Secure Erase" is suggested to remove the user settings e.g. if you were about to sell the scope. The "Hard Reset" is recommended for issues with start-up and whilst it certainly delays the bootup, it doesn't affect the ~ 30ns skew between pairs of channels (separate ADCs).

I sent a follow-up video (using DropBox) to my Siglent dealer showing the setup of the FG and channels on the scope:

https://www.dropbox.com/s/lv0rczm5bcguz9k/IMG_1965.MOV?dl=0

The movie shows all channels show deskew 0. At 1MHz the two traces separate (widen) using input channel 1 and moving 2 -> 3 or 2 -> 4. At 15 MHz the out of phase is obvious as input is moved across to 3 or 4. Emailed the dealer screen prints. With timebase set up for 15MHz sine wave, the 1MHz or 15MHz traces show a shift of 1 division when 2nd ADC is used. Ideally, this wouldn't need to be manually corrected by remembering to set a deskew of ~ -30ns on channels 3 & 4. Hopefully, there is an easy re-calibrate fix, or I (aka the 'rookie') overlooked something.  ;)

Simon
« Last Edit: July 04, 2021, 10:58:30 pm by SimonM »
 

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #30 on: July 04, 2021, 11:24:25 pm »
The more I look at my scope, the less confident I think I know anything about what it's doing...

I have two cables, one was delivered as RG58 50 ohm and the second is RG59 75 ohm. I bought the RG59 that was described as RG58 - when it arrived, I noticed a difference and checked the identification printed on the cable. The vendor had made a mistake because it is RG 59! I am looking to replace both BNC cables with some better and balanced RG58 50-ohm cables, as buy cheap, buy twice...

I'm looking at the differences using a 15MHz square wave out of SDG1032X set for 50 ohms) on both channels with 2nd channel locked, feeding 50-ohm terminators into the scope. I'm using channels 1 & 3 of the scope and the waveforms are slightly different, with RG58 providing a squarer waveform, especially evident at 30MHz, compared to the RG59.

Both these channels should be in phase - that's a known problem with the scope that the dealer is looking into and using channels 1 & 2 they can be shown to be in phase. Here's a print, where they are out of phase using channels 1 & 3:

1233548-0

As well as being out of phase, I get an interesting effect when I turn on either of the two other channels. The 3rd channel is phase-shifted, see the attached print. At 30MHz this is also particularly noticed, but for the cables, I was using a lower frequency signal.

1233550-1

Channels 2 & 4 aren't connected to anything. Since I'm triggering on channel 1, this stays exactly centred around the trigger point, but channel 3 moves back. Unless I haven't thought it through, I wouldn't expect adding another channel to the display, to make any difference to existing channels. Perhaps it just highlights again that the 2nd ADC isn't (particularly) aligned with anything? I get a similar shift if I trigger on another channel and add either of the other two channels.

Simon

 

Offline StillTrying

  • Super Contributor
  • ***
  • Posts: 2850
  • Country: se
  • Country: Broken Britain
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #31 on: July 05, 2021, 11:54:42 pm »
It shouldn't do that either.

"that's a known problem with the scope that the dealer is looking into"

Have they told you anything such as tested one of their current stock? The scope's own 120ns rise time 1kHz would be enough to see the 29ns error.

All I can think of is reinstall the latest firmware, hopefully tautech will know if that's a worth trying not.

I suspect it will be going back myself.
.  That took much longer than I thought it would.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #32 on: July 06, 2021, 02:16:04 am »
It shouldn't do that either.

"that's a known problem with the scope that the dealer is looking into"

Have they told you anything such as tested one of their current stock? The scope's own 120ns rise time 1kHz would be enough to see the 29ns error.

All I can think of is reinstall the latest firmware, hopefully tautech will know if that's a worth trying not.

I suspect it will be going back myself.
Yeah I know......been busy....

Problem is most likely with settings in SDG1032X where outputs are not perfectly edge aligned by default so settings must be tweaked to ensure edges are perfectly in phase.

Adjust as follows:
SDG>Utility>P2>Phase Mode = Locked
Utility>P1>Channel Copy Coupling>Track = ON

This SDG1032X requires a Ch2-Ch1 Phase Deviation of -2o

Screenshots below SDS1104X-E where SDG Ch1 always = Ch4 on SDS.
First 2 on each scope ADC and last 2 on same ADC.
The scope is NOT the problem only the phase deviation in the SDG.
It should also be noted that once channel phase differences have been trimmed out that a Power ON =Last setting be made so adjustments to phase differences will be saved.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #33 on: July 06, 2021, 02:25:20 am »
The scope is NOT the problem only the phase deviation in the SDG.

I'm not clear on what you are saying here...

At one point he split a single channel and observed 29ns of difference on the scope.  How could the scope not be the problem?
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #34 on: July 06, 2021, 03:03:46 am »
The scope is NOT the problem only the phase deviation in the SDG.

I'm not clear on what you are saying here...

At one point he split a single channel and observed 29ns of difference on the scope.  How could the scope not be the problem?
Where, which post ?
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #35 on: July 06, 2021, 03:14:18 am »
Read my post #16 and his reply #17.  If he did what he said he did--both probes to one channel, switching the connections at the scope and observing no change or reversal of the phase shift--that scope is borked.

I have no idea what could possibly be going on with that scope, but something isn't right.  29ns is huge for any physical delay and the only thing I can think of is that there is a factory-only calibration constant (or at least not done during self-cal) that has a range greater than necessary and that value has accidentally been set to the min or max somehow.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #36 on: July 06, 2021, 03:26:29 am »
Read my post #16 and his reply #17.  If he did what he said he did--both probes to one channel, switching the connections at the scope and observing no change or reversal of the phase shift--that scope is borked.

I have no idea what could possibly be going on with that scope, but something isn't right.  29ns is huge for any physical delay and the only thing I can think of is that there is a factory-only calibration constant (or at least not done during self-cal) that has a range greater than necessary and that value has accidentally been set to the min or max somehow.
You may be correct but I'm not convinced as yet.
SimonM needs to do a factory reset on his AWG so to eliminate any possibility the SDG1032X is not the source of phase difference.
His last screenshots show ~180o phase difference @ 15 MHz and there's no way it should be that far out.
I believe something has been overlooked.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline StillTrying

  • Super Contributor
  • ***
  • Posts: 2850
  • Country: se
  • Country: Broken Britain
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #37 on: July 06, 2021, 03:35:06 am »
Yeah I know......been busy....

I/we mean SimonM's dealer in the UK.

Definitely the scope, first 2 images in Reply#13

https://www.eevblog.com/forum/testgear/using-(and-living-with)-siglent-sds1000x-e-series-oscilloscopes/msg3597615/#msg3597615

CH2 overwrites CH1, but CH3 doesn't overwrite CH1

"His last screenshots show ~180o phase difference @ 15 MHz and there's no way it should be that far out"

1/15M = 66.6ns, half a cycle is 33ns which is the 29ns error, getting close to 180 deg for 15MHz.
.  That took much longer than I thought it would.
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #38 on: July 06, 2021, 03:39:10 am »
SimonM needs to do a factory reset on his AWG so to eliminate any possibility the SDG1032X is not the source of phase difference.
His last screenshots show ~180o phase difference @ 15 MHz and there's no way it should be that far out.

How could the SDG be the issue if he only used  one channel??

The phase difference just seems to be about the same 29ns lag in the 3/4 ADC, although it seems to reduce a bit at 500MSa/s. 

Quote
I believe something has been overlooked.

It certainly seems like something that you eventually figure out and slap yourself in the head, "DOH!" and wonder what you were thinking.  I just can't think of any user setting that would cause this.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline StillTrying

  • Super Contributor
  • ***
  • Posts: 2850
  • Country: se
  • Country: Broken Britain
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #39 on: July 06, 2021, 03:48:15 am »
It certainly seems like something that you eventually figure out and slap yourself in the head, "DOH!" and wonder what you were thinking.  I just can't think of any user setting that would cause this.

I had to read Reply#13 3 times the first time, and now again so 4 times.

"29ns is huge for any physical delay"

I could imagine the PCB needs a fixed 29ps delay somewhere and the firmware is using ns instead. :)
.  That took much longer than I thought it would.
 

Offline Orange

  • Frequent Contributor
  • **
  • Posts: 346
  • Country: nl
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #40 on: July 06, 2021, 05:19:49 pm »
There have been other reports of phase error's with this scope. see for instance this posting

https://www.eevblog.com/forum/testgear/siglent-sds1204x-e-released-for-domestic-markets-in-china/msg3233194/#msg3233194

Forget about all the noise of the SDS generator or wrong cabeling or Bode plot crap... It's the scope

I suspect that this is an error in the FW or in the calibration process, which needs to correct the timing of the 2 ADCs

You can actually watch the process on screen if you put the scope into single shot mode.
setup:
TB at 50ns/div. Generator at 10MHz, maximum sample memory (14 meg), sinx/x interpolation, vector mode.
sequence:
First only enable channel 1.
single shot, one channel visible.
enable channel 3.
now watch carefully what happens if you do a single shot.
you first see a misalignment of two sines, and immediately, a time shift on channel 3, aligning channel 1 and 3

I have witnessed this error myself with older FW. Now with the latest FW, the error has been reduced to 600 ps
Still not 100% ok, and certainly not within the Siglent spec of 100 ps

What version of FW on the scope are you using SimonM ?
 

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4086
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #41 on: July 07, 2021, 06:12:19 am »
Very simple and robust method for check and measure this level of oscilloscopes channels time difference.
For avoid all hassle and suspects use ONLY one signal. Take this same signal with Time matched cables to scope channels under test. You do also not need play with different frequencies etc or Phase angles. Just measure time, repeat, measure time.
Simple method is Y cable. Yes, just cables centers connected and shields connected. Do not think impedance matching here, this is simple split. Now it  is also easy to fine adjust both output sides perfect time if want.

Third cable in this Y, input. This length do not matter, 1cm  to 1 meter is just ok.

Connect it to gen out and then these Y cable time mached sides to scope channels under test.
Use what ever signal what give fast edge as possible. It can be pulse, square, triangle0% or even sine but twith sine you need high frequency. Using just fast edge waveforms you can use 1kHz or 1MHz, no matter... but keep this base freq so that period is least well over 2 times max time difference in channels under test... so even first time noob can not go to wrong.

Also I have seen many og SDS1000X-E scopes, 2 and 4 channels. I have never ever seen 10 or 30ns time skew between any channel. Also I ask how it can be FW problem if there 1000 scope without this problem and 1 scope with this problems. Just wake up and  start  thinking...

Also it is amazing what kind of hassle this looks when this is tested in this thread...  how it can be this kind of total hassle.

Even if can not have/diy this kind of Y cable... I have told to OP how he can make very very easy and robust test just with all  what he have... but no... just continuing this hassle like walking in circle where nothing is clear at once.

Here attached one image.
There is also simpliest possible DIY Y cable.
Done so that cable time skew is roughly 2.5ps so I just multiply it with 2 so that I can always trust it is <5ps when BNC are connected perfectly. I have hand adjusted lengths using scope when connect this Y center.
Note that RG316/U  travel time 4.856ps/mm (ref. Belden)  (yes I did many iterations for get as perfect match what I can easy detect with my tools. Note that 1cm difference mean still under 50ps error, so you can do this kind of cable just measuring length)
With this kind of simple split it is very important that cable outputs load impedance/reactance are as equal as possible. Still naturally it is not at all impedance matched signal transmission. But, this do not matter because only important is  Time difference and just roughly.

No need hassle with 100MHz or 1MHz or 15MHz. Just what ever from 1s to  1us period edge is enough and measure channels time skev (including amount of Y cable time error).  Do not play with probes or other things what can way or other suspect. Just reduce all unknowns to minimum.

If scope have roughly 29ns enormous time skew between channel pairs (different ADC), as previously told, no need anything but return it as DOA and get it changed or repaired. Have you contacted Siglent technical support.
This kind of thing need just one or two very simple and easy measure and thats all. Result is failed or ok. Until start playing "this and that".

« Last Edit: July 08, 2021, 04:03:09 am by rf-loop »
I drive a LEC (low el. consumption) BEV car. Smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the wises gone?
 
The following users thanked this post: Performa01, 2N3055

Offline StillTrying

  • Super Contributor
  • ***
  • Posts: 2850
  • Country: se
  • Country: Broken Britain
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #42 on: July 09, 2021, 02:48:17 pm »
Has your replacement SDS1104X-E arrived yet. :)
.  That took much longer than I thought it would.
 

Offline MathWizard

  • Super Contributor
  • ***
  • Posts: 1411
  • Country: ca
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #43 on: July 09, 2021, 09:48:21 pm »
Why is it sometimes a horizontal line is shown as you set the trigger level, and other times, it won't show it ? I haven't noticed a pattern to why it stop's working, I guess some software bug

I am very pleased tho with my siglent equipment in general so far. I'd like to try their handheld Oscope/DMM, if only I had a need to take 1 on a roadtrip repair.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #44 on: July 10, 2021, 01:40:46 am »
Why is it sometimes a horizontal line is shown as you set the trigger level, and other times, it won't show it ? I haven't noticed a pattern to why it stop's working, I guess some software bug
Level lines only show when adjustments are made and for a short period afterwards and then only the level marker is displayed as reference to the newly set level.

Quote
I am very pleased tho with my siglent equipment in general so far. I'd like to try their handheld Oscope/DMM, if only I had a need to take 1 on a roadtrip repair.
If you can wait there’s a new SHS series coming......dunno much more sorry other than they will have a new GUI and be more aligned to the later X and X-E model DSO’s.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline skempf

  • Newbie
  • Posts: 5
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #45 on: July 13, 2021, 12:12:35 pm »
Is there a summary checklist of all the (non trivial) items you should check or test when unboxing a new scope? (I just got a hold of an 1104x-e).
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #46 on: July 13, 2021, 12:26:43 pm »
Is there a summary checklist of all the (non trivial) items you should check or test when unboxing a new scope? (I just got a hold of an 1104x-e).
Not really.
In the process of compensating the probes check each probes 1x/10x switch for clean and correct attenuation of the probe compensation output.....very occasionally there have been dud probes in the past....of late, rare.
Set all probes and channel attenuation to 10x.
You may find the button beeps sound annoying so disable it in Utility>Sound.
Do all that ^ then reboot.

Then enjoy.  :)
« Last Edit: July 13, 2021, 12:29:07 pm by tautech »
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #47 on: July 14, 2021, 10:20:21 am »
Code: [Select]
Has your replacement SDS1104X-E arrived yet. :)

Yes - the UK dealer replaced the scope with a new one from their current stock
The new scope has the same (current) firmware, was manufactured (calibrated) earlier this year, and has a serial number that indicates it is from the same batch (the dealer had 38 in stock).

No noticeable skew between channels
(ADC 1 = channel 1 & 2) and (ADC 2 = channel 3 & 4) have no noticeable skew that I could detect.

I'm using:
  • SDG1032X Function Generator with one output at 30MHz square wave (50-ohm output) on channel 1 (channel 2 is OFF)
  • T Splitter and two (near identical) 1m BNC cables
  • 50Ω Impedance Adapter on each of the two scope inputs
I have let the scope warm-up, calibrated with "Self Cal" whilst all probes and BNC cables are unconnected, and then restarted the scope.

I'm using the fastest square wave source I have and the fastest rise time available. With the impedance adapters, the signal is very clean.

My tests - confirm all is OK
On all the channels (triggered + 2nd): 1 + 2, 1 + 3, 1 + 4, 2 + 1, 2 + 3, 2 + 4, 3 + 1, 3 + 2, 3 + 4, etc.

I can observe the traces being in phase and the measured skew averages 20ps across all combinations.

I have added the extra channels, hidden them, and repeated all the combinations I could think of without any issues. I'm manually triggering and getting readings of 10ps, 20ps, or 30ps using the measured skew channel A and channel B. The Siglent dealer had previously set up the scope and triggered it with a 60MHz sine wave with the triggering and measurement of skew for channels 1 & 3 still active when I received it. Initially, the scope showed two power cycles.

Conclusion
I don't have access to the same level of test equipment as my Siglent dealer. I'm getting a good result using the fastest rise time available using my own function generator. Any variations measured by the scope are more than likely an indication of small differences between the BNC cables, which are acting as a filter on the output signal from the FG.

I'm pleased with the outcome and delighted with my Siglent scope and also my function generator.

Simon
« Last Edit: July 14, 2021, 12:57:46 pm by SimonM »
 
The following users thanked this post: tautech, StillTrying, Andrew_Debbie

Offline Orange

  • Frequent Contributor
  • **
  • Posts: 346
  • Country: nl
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #48 on: July 14, 2021, 12:06:56 pm »
Has your replacement SDS1104X-E arrived yet. :)

Yes - the UK dealer replaced the scope with a new one from their current stock
The new scope has the same (current) firmware, was manufactured (calibrated) earlier this year, and has a serial number that indicates it is from the same batch (the dealer had 38 in stock).

No noticeable skew between channels
(ADC 1 = channel 1 & 2) and (ADC 2 = channel 3 & 4) have no noticeable skew that I could detect.

I'm using:
  • SDG1032X Function Generator with one output at 30MHz square wave (50-ohm output) on channel 1 (channel 2 is OFF)
  • T Splitter and two (near identical) 1m BNC cables
  • 50Ω Impedance Adapter on each of the two scope inputs
I have let the scope warm-up, calibrated with "Self Cal" whilst all probes and BNC cables are unconnected, and then restarted the scope.

I'm using the fastest square wave source I have and the fastest rise time available. With the impedance adapters, the signal is very clean.

My tests - confirm all is OK
On all the channels (triggered + 2nd): 1 + 2, 1 + 3, 1 + 4, 2 + 1, 2 + 3, 2 + 4, 3 + 1, 3 + 2, 3 + 4, etc.

I can observe the traces being in phase and the measured skew averages 20ps across all combinations.

I have added the extra channels, hidden them, and repeated all the combinations I could think of without any issues. I'm manually triggering and getting readings of 10ps, 20ps, or 30ps using the measured skew channel A and channel B. The Siglent dealer had previously set up the scope and triggered it with a 60MHz sine wave with the triggering and measurement of skew for channels 1 & 3 still active when I received it. Initially, the scope showed two power cycles.

Conclusion
I don't have access to the same level of test equipment as my Siglent dealer. I'm getting a good result using the fastest rise time available using my own function generator. Any variations measured by the scope are more than likely an indication of small differences between the BNC cables, which are acting as a filter on the output signal from the FG.

I'm pleased with the outcome and delighted with My Siglent scope and also my function generator.

Simon

So you got it sorted, good to hear.

After all I was right that the scope was the problem.

From rf-loop:
Also I have seen many og SDS1000X-E scopes, 2 and 4 channels. I have never ever seen 10 or 30ns time skew between any channel. Also I ask how it can be FW problem if there 1000 scope without this problem and 1 scope with this problems. Just wake up and  start  thinking...
;)
« Last Edit: July 14, 2021, 12:10:18 pm by Orange »
 

Offline SimonMTopic starter

  • Contributor
  • Posts: 34
  • Country: gb
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #49 on: July 14, 2021, 02:12:28 pm »
I do not remember/know with most new FW if there is still a total deep reset.
During the whole power-up boot time, toggling very fast Math button.

Yes, I had previously followed the instructions for the Oscilloscope Hardware Reset that are on the EU Siglent website. I also had an email from Siglent in Germany to confirm that a Hardware Reset "sorted" the problem until powering off (it didn't make any difference to the ~29ns skew, except delaying the startup and afterward, suggest a "Self Cal").

There is a very poor simple and dirty method. "Wilderness and desert method. Just for bulletproof verify the problem is or is not."

ETA After saw your video:
You can Connect your BNC T to channel 2 and trigger to CH2. Connect just one of your coaxial from SDG to this T other side. Then take just a very short copper wire and connect it from BNC T other side center as short as possible to the CH3 input center.

Now you have only one coax from SDG and very small, and roughly known, the signal travel time difference between CH 2 and Ch3 input.

With this, because you do not have better things (special time matched cables and good splitter etc) available, you can infinitely prove if there is an error between channels (when they are different ADC).

I appreciate the thought. As you say, we don't all have (so far) all the right test gear. On the new SDS1104X-E scope, I used one BNC cable, with a T splitter and two short lengths of copper hook-up wire. I'm putting the FG output (1 MHz square wave) into channels 1 & 3 and using channel 2 as a ground (see the picture).



The result of using it is also shown and the error is ~-5ps.



The scope traces are intertwined/overlain which is also a good sign.  :-+

Simon



PS one "useful" feature of the scope: after it has triggered, many of the measurements can be determined afterward e.g. I realized that the skew was not measured from the triggered channel (easily corrected) and I added some other useful info e.g. the Pk=Pk voltage and the rise time of the square wave. In the second scope picture, it's the exact same waveform with only the measurements updated.
« Last Edit: July 14, 2021, 02:47:01 pm by SimonM »
 

Offline skempf

  • Newbie
  • Posts: 5
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #50 on: July 15, 2021, 03:49:40 pm »
I don't have a 50Ω terminator, so I'm just using a probe. The FG (1032x) is set to 1MHz square wave, 50% duty, 1Vpp. Is this what one would expect to read on the scope?

 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #51 on: July 15, 2021, 03:54:55 pm »
I don't have a 50Ω terminator, so I'm just using a probe. The FG (1032x) is set to 1MHz square wave, 50% duty, 1Vpp. Is this what one would expect to read on the scope?

Did you compensate the probe?  Are you using the BNC-probe adapter directly onto the SDG1032X?
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline skempf

  • Newbie
  • Posts: 5
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #52 on: July 15, 2021, 04:12:00 pm »
I'm using a probe that came with the 1104x-e, on 10x setting, and checked the compensation prior to the above reading. I'm just holding the probe to the exposed BNC male connector on the 1032x.
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #53 on: July 15, 2021, 04:21:49 pm »
I'm using a probe that came with the 1104x-e, on 10x setting, and checked the compensation prior to the above reading. I'm just holding the probe to the exposed BNC male connector on the 1032x.

The ringing is probably due to ground lead inductance.  The PP510 probes apparently don't come with a BNC adapter, but they should come with a ground spring--a little coiled up thing.  Remove your ground lead and slide the ground spring over the probe, then make sure the spring touches the BNC shell of the SDG1032X when you are probing it.  The spring should connect the shell to the exposed foil on the probe.  That should reduce the ringing somewhat.  A BNC-probe adapter is a collar that goes between the BNC and probe and provides a matched, fully enclosed ground and is what you need to properly use a probe this way.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline skempf

  • Newbie
  • Posts: 5
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #54 on: July 15, 2021, 06:19:57 pm »
See attached using the ground spring on the probe. Is it safe to assume that all of the "ringing" in the waveform is ground lead inductance?
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #55 on: July 15, 2021, 08:32:01 pm »
I don't have a 50Ω terminator, so I'm just using a probe. The FG (1032x) is set to 1MHz square wave, 50% duty, 1Vpp. Is this what one would expect to read on the scope?
Precisely what you expect without 50 Ohm termination.

Risetime meets and betters the 7ns spec and the overshoot is the result mismatched termination on the SDG which BTW always has a 50 Ohm source impedance.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #56 on: July 16, 2021, 01:44:10 am »
Precisely what you expect without 50 Ohm termination.

Risetime meets and betters the 7ns spec and the overshoot is the result mismatched termination on the SDG which BTW always has a 50 Ohm source impedance.

Where is the mismatch???
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #57 on: July 16, 2021, 01:56:36 am »
Precisely what you expect without 50 Ohm termination.

Risetime meets and betters the 7ns spec and the overshoot is the result mismatched termination on the SDG which BTW always has a 50 Ohm source impedance.

Where is the mismatch???
50 Ohm source to 10x probe.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #58 on: July 16, 2021, 02:36:49 am »
Precisely what you expect without 50 Ohm termination.

Risetime meets and betters the 7ns spec and the overshoot is the result mismatched termination on the SDG which BTW always has a 50 Ohm source impedance.

Where is the mismatch???
50 Ohm source to 10x probe.

The source is it's own termination, there's no need to add another when there's no transmission line.  I don't know the exact distance between the 50R pad in the SDG and the end of the BNC connector, but λ/4 would be in the GHz range for the likely distances of a few cm.  You can add a 50R pass-thru to make it a 25R output if you like, but there's still a distance between that 50R and the terminators BNC connector....
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #59 on: July 16, 2021, 02:43:35 am »
Knock yourself out looking at properly terminated sources for square waves here:
https://www.eevblog.com/forum/testgear/show-us-your-square-wave/
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #60 on: July 16, 2021, 03:46:46 am »
Knock yourself out looking at properly terminated sources for square waves here:
https://www.eevblog.com/forum/testgear/show-us-your-square-wave/

I have my own--SDG2042X with PP215 probe and BNC adapter, Leo Bodnar fast edge with the same plus a BNC-to-BNC extension.  Each has issues, but not ringing.

You need to terminate the characteristic impedance of a transmission line, not a source.



A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline Orion33

  • Contributor
  • Posts: 14
  • Country: kz
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #61 on: October 02, 2022, 12:22:46 pm »
So, I got my Siglent 1104X-E under the AKIP brand. Success :)
However, I can not calibrate the scope qualitatively. After each iteration, I get different DC offset values. Moreover, these values change over time.
What deviation is acceptable? Now after 2 hours of warm-up I get 400uV on the first channel
Is it possible to calibrate each channel individually using service commands?
If I activate options, for example, AWG, how do I then turn it back off for a warranty claim?
« Last Edit: October 02, 2022, 02:02:30 pm by Orion33 »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #62 on: October 02, 2022, 07:27:59 pm »
So, I got my Siglent 1104X-E under the AKIP brand. Success :)
However, I can not calibrate the scope qualitatively. After each iteration, I get different DC offset values. Moreover, these values change over time.
What deviation is acceptable? Now after 2 hours of warm-up I get 400uV on the first channel
Is it possible to calibrate each channel individually using service commands?
If I activate options, for example, AWG, how do I then turn it back off for a warranty claim?
Which version firmware is installed ?

We can not see what you do, please post some screenshots with relevant menus visible, preferably in English.
Also, describe your signal connections if any and be sure to run the Self Cal only after ~30 minutes of warm up time.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Orion33

  • Contributor
  • Posts: 14
  • Country: kz
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #63 on: October 02, 2022, 08:45:01 pm »
Which version firmware is installed ?

We can not see what you do, please post some screenshots with relevant menus visible, preferably in English.
Also, describe your signal connections if any and be sure to run the Self Cal only after ~30 minutes of warm up time.
V6.1.37R8
All day long I just run self-calibration. The device worked for almost 9 hours. The result of the last calibration in the photo. For some reason, the device does not recognize the flash drive (exFAT). During the day, the best result was +/-80uV. After that, it was not possible to repeat it, the calibration of one of the channels is always unsatisfactory, as in the photo. The device works in a separate room where other devices are turned off.
« Last Edit: October 02, 2022, 08:50:10 pm by Orion33 »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #64 on: October 02, 2022, 08:50:01 pm »
[For some reason, the device does not recognize the flash drive.
It needs be 8GB max FAT32 and 4k clusters.
Some brands of 16GB sticks work.

Your stick might be faulty or formatted incorrectly.
Get this fixed first.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Orion33

  • Contributor
  • Posts: 14
  • Country: kz
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #65 on: October 02, 2022, 08:52:03 pm »
Yes, I just made sure my flash drive has exFAT. Unfortunately, I can't format it, and can't find another one. Everything is clearly visible in the photo.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28314
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #66 on: October 02, 2022, 09:01:17 pm »
Which version firmware is installed ?

We can not see what you do, please post some screenshots with relevant menus visible, preferably in English.
Also, describe your signal connections if any and be sure to run the Self Cal only after ~30 minutes of warm up time.
V6.1.37R8
Please install V6.1.37R9
https://int.siglent.com/download/firmwares/?ProId=12
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #67 on: October 02, 2022, 09:55:51 pm »
All day long I just run self-calibration. The device worked for almost 9 hours. The result of the last calibration in the photo. For some reason, the device does not recognize the flash drive (exFAT). During the day, the best result was +/-80uV. After that, it was not possible to repeat it, the calibration of one of the channels is always unsatisfactory, as in the photo. The device works in a separate room where other devices are turned off.

That doesn't look broken or unsatisfactory to me. 
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline Orion33

  • Contributor
  • Posts: 14
  • Country: kz
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #68 on: October 03, 2022, 09:18:49 am »
Please install V6.1.37R9
https://int.siglent.com/download/firmwares/?ProId=12
I asked the distributor if this would void the warranty.
That doesn't look broken or unsatisfactory to me. 
≤±4.0%: ≤2 mV/div
For 500uV/div 4% is 500*0.04 = 20uV.
 

Offline Orion33

  • Contributor
  • Posts: 14
  • Country: kz
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #69 on: October 03, 2022, 09:34:14 am »
With open terminals I have this picture. The noise from the AC mains is large enough to make a correct calibration. How do I get rid of it?
« Last Edit: October 03, 2022, 09:36:42 am by Orion33 »
 

Offline Orion33

  • Contributor
  • Posts: 14
  • Country: kz
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #70 on: October 03, 2022, 10:33:29 am »
How to save the calibration results to a file so that I can apply them later?
 

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4086
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #71 on: October 03, 2022, 11:12:23 am »
Please install V6.1.37R9
https://int.siglent.com/download/firmwares/?ProId=12
I asked the distributor if this would void the warranty.
That doesn't look broken or unsatisfactory to me. 
≤±4.0%: ≤2 mV/div
For 500uV/div 4% is 500*0.04 = 20uV.

It is DCGain accuracy! Where you have measured or even looked from screen DC Gain using any acceptable method?

DC Offset accuracy is totally different thing. Note there is in data sheet formula also this ".....+500µV"

And if you can not even format USB stick to native FAT32 format please do not even try FW update.
I drive a LEC (low el. consumption) BEV car. Smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the wises gone?
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7814
  • Country: us
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #72 on: October 03, 2022, 12:32:55 pm »
≤±4.0%: ≤2 mV/div
For 500uV/div 4% is 500*0.04 = 20uV.

That's doubly incorrect.  The DC Gain spec you've quoted would mean relative to the input.  If you have 0V input, the stated tolerance for DC Gain error would be 0V * 4%, which is of course zero.

What you are seeing is not a DC Gain error, it is offset.  The offset spec for zero offset and zero input boils down to 1.5% of 8 divisions (not 1) + 500uV, which is 1.5% of 4mV + 500uV or ~560uV.  Your errors are all less than 500uV, so I'd say it is considered acceptable by the specifications.  Also, your traces look very quiet--was this done with all open BNC connectors?  The calibration procedure should be done with nothing connected, although in practice I'm not sure how much difference this makes. 

Here is a screenshot of my scope after running self-cal after 30 minutes of warmup.  It seems better than yours, perhaps, but still with ~100uV offsets. 

A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline Orion33

  • Contributor
  • Posts: 14
  • Country: kz
Re: Using (and living with) Siglent SDS1000X-E Series Oscilloscopes
« Reply #73 on: October 04, 2022, 11:01:19 am »
Also, your traces look very quiet--was this done with all open BNC connectors?
Yes, with open but with averaging by 1024 to good see DC offset.
Here is a screenshot of my scope after running self-cal after 30 minutes of warmup.  It seems better than yours, perhaps, but still with ~100uV offsets. 
Yesterday I also achieved 160uV deflection. I asked how I can save this calibration to a file and apply it on subsequent failures, but so far no one has answered.
Yes, the distributor said that I can install the latest firmware myself. I'm going to do it now.
 