Author Topic: Using a Prema 5017  (Read 1169 times)

0 Members and 1 Guest are viewing this topic.

Offline ResonantTopic starter

  • Contributor
  • Posts: 13
  • Country: gb
Using a Prema 5017
« on: October 04, 2019, 07:13:53 pm »
I have a question about the use of the Prema 5017. I purchased it about 3 years ago and it came from the estate of a deceased voltnut and it was in good cosmetic condition. It has been calibrated by Prema twice recently last time April 19.
When turned on and allowed to warm up for a couple hours it seems accurate when checked against a calibrated source. However, if it is left on in 7.5 digit mode on 30V range using a 10V standard for more than a day it starts to drift a lot, around 0.000171V. Using my own 34401a and borrowed Keithley 2002 they shows almost no drift over the same period against the same Standard.
I have checked the internal reference LM399 over the same period and that also seems stable. The big difference seems to be the Prema shows a large offset, if you short the leads.
My previous experience has been using US made meters like 34401a and Keithley which don’t develop much of an offset and can be connected up and left for long periods.
So do I have a faulty meter or is a characteristic that the Prema needs regular zeroing to be accurate?

Resonant
 

Online Kosmic

  • Super Contributor
  • ***
  • Posts: 2529
  • Country: ca
Re: Using a Prema 5017
« Reply #1 on: October 04, 2019, 07:58:49 pm »
Mine is pretty stable over multiple days. You should check the electrolytic capacitors. Had to change all of them on the analog board and some on the digital board. With the old cap my unit was drifting badly.
 

Offline maginnovision

  • Super Contributor
  • ***
  • Posts: 1963
  • Country: us
Re: Using a Prema 5017
« Reply #2 on: October 04, 2019, 08:26:00 pm »
Yea, I had the same issue. It was very noticeable because the bad cap was getting much warmer than the others.

I added a screenshot of the graphed data from my issue(reading a short). As you see it got really bad but now it's stable again.
« Last Edit: October 04, 2019, 09:11:48 pm by maginnovision »
 

Offline ResonantTopic starter

  • Contributor
  • Posts: 13
  • Country: gb
Re: Using a Prema 5017
« Reply #3 on: October 05, 2019, 04:20:04 pm »
Thanks for the advice, I have new Caps on order and I'll report back on progress next week. I was slightly surprised to find CapXon and Jamicon Caps in a piece of expensive test equipment whilst none tested faulty the ESR was poor compared with premium brands like Panasonic or Nichicon.
Thanks
 

Offline ResonantTopic starter

  • Contributor
  • Posts: 13
  • Country: gb
Re: Using a Prema 5017
« Reply #4 on: February 07, 2020, 12:28:48 pm »
Well I finally got the new Caps installed, I faced a bit of a delay with a problem on the connections on C15 which relies on the though hole plating. This stopped the unit booting correctly but was fixed by adding a link C15 to L2 and no further problems and it seems much more stable its no been on for 24hrs. For reference I used Panasonic FC  range 1050C from RS.
Resonant.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf