The clamp type probe has a good change to work, maybe marginally of due to the different load resistor, but one may not notice the difference.
The high voltage probel would likely need an extra 1 M (or 1.1 M if the meter has 10 M input impedance) in parallel to make it lokk more like the scope. AC would be limited (low frequency only) as the DMM usually has higher input capacitance than the scope. For the high voltage type, keep in mind that the DMM is usually not grounded like a scope. It may need an extra ground link.