| Very slow measurement from Rigol MC3065 DMM |
| (1/1) |
| stevessp:
Taking a voltage reading on this instrument is taking over 3 seconds which is an eternity. Haven't been able to figure out a way to speed it it up from the documentation. Anyone have any ideas? Thanks. |
| TurboTom:
You can set the integration time of the DMM module. Search the manual for "PLC" (Power Line Cycles) or see page 2-24 for further information on that topic. If this doesn't solve the problem, you may check if the measurement uses a trigger function to take readings at predetermined intervals. Setting up an M300 system can be a real PITA... Confusing to tell the least. :-// |
| stevessp:
Thanks. Doesn't that only apply to DC though? |
| TurboTom:
I'm not sure. But since there's the RMS/DC converter contained in the DMM and thus the AC measurement is internally also a DC measurement, the integration time should apply here as well. |