Author Topic: VFC on a multmeter  (Read 10664 times)
Offline ButchTopic starter
VFC on a multmeter
« on: April 08, 2020, 07:27:08 pm »
The other day i bought a multimeter (Mustek DMC100c) that has a V.F.C function (Voltage Frequency Converter).
From the manual i can not really understand what this function adds to the multimer and what the benefit is.
Can somebody explain this function (in theory and what the benefit is to the accuracy of an AC-measurement)?
Regards
Butch
Offline ogden
  • Country: lv
Re: VFC on a multmeter
« Reply #1 on: April 08, 2020, 08:37:51 pm »
Voltage to frequency converter most likely is internal function of said obscure multimeter that adds nothing to function or accuracy of said device. Seems like marketing BS. Note that google search returns exactly zero hits for "Mustek DMC100c".
« Reply #2 on: April 08, 2020, 09:47:38 pm »
Thanks for the response.
How does this answer my question? "most likely", "obscure", "Seems", "No google return".
Regards
Offline bob91343
« Reply #3 on: April 08, 2020, 10:02:40 pm »
If indeed the acronym means Voltage to Frequency Converter, I don't see its application for a voltmeter.  If it were Frequency to Voltage, then I'd guess it would turn it into a frequency meter.
Searches on the brand name and on the model number bring up nothing useful.
What do you get when you select vfc?  What is the info in the manual?
Online 2N3055
  • Country: hr
« Reply #4 on: April 08, 2020, 10:03:00 pm »
That is probably filter to filter out high frequencies when measuring VFC (frequency converter motor controllers) that use high frequency PWM.  And it is Mestek ..
Dr. Richard W. Hamming
The following users thanked this post: ogden
Offline ogden
  • Country: lv
« Reply #5 on: April 08, 2020, 10:24:33 pm »
Answer to your question was: "adds nothing to function or accuracy of said device"
And it is Mestek ..
Seems you (OP) can't even properly tell what you are talking about.
« Last Edit: April 08, 2020, 10:39:02 pm by ogden »
Online 2N3055
  • Country: hr
« Reply #6 on: April 08, 2020, 10:45:07 pm »
And it is Mestek ..
Seems you can't even properly tell what you are talking about.
It is Mestek DMC100c not Mustek. It is obvious what I meant. (BTW Mustek makes scanners)..
And what you said was wrong. A low-pass filter for measurement of voltage on motors connected to frequency converters is a common function and I'we seen it called VFD and VFC on metes. For instance, Extech calls it VFC, Bymen and Fluke VFD.
It filters out PWM frequency from VFD controller and leaves only low frequency to be able to accurately measure RMS voltage/current  going into motor.. It's not marketing wank and is absolutely useful if you work with motor control
The following users thanked this post: gtv
Offline ogden
  • Country: lv
« Reply #7 on: April 08, 2020, 11:01:34 pm »
I was not addressing you. Sorry if you did not get it :-\ BTW note that I clarified who I am talking to, long before your post.
Thank you for explaining, in your 2nd try. Initially you were not so sure:
Offline joeqsmith
« Reply #8 on: April 09, 2020, 01:23:38 pm »
Regards
Butch
Welcome to the forum.  The Variable Frequency Drives (Controllers) can be a problem to measure.   A while back I was playing around with a Fluke non-contact type voltage meter and was curious if it would have troubles with something more complex than AC coming into the house.  I setup a test using a scope to provide a visual idea of what the waveforms I applied looked like.  Testing begins with a small VFD.
BTW, later I took the Fluke with me to try on some other controllers and it was hit and miss if it would work or not, mostly depending on the carrier frequency the VFD was set to.
Offline Wytnucls
« Reply #9 on: April 10, 2020, 12:44:38 pm »
Does not seem to have anything to do with voltage conversion: