| twoexem:
Sorry for the bump, but this was the only usable thread I found. What config (data bytes, stop bytes, parity, baud etc.) did you all use to connect to the scope? I'm currently trying to connect to my DSO5202P and no matter which settings I choose, I always get garbled junk data. It might be an issue with my USB serial connector, though, but it worked fine before so I can't imagine it's that. |
| airwolf:
I used 115.200 baud, 8 data bits, 1 stop bit no parity and no flow control. As USB-serial converter I use a CP2102 based device, this kind always seems to work while the PL2303 based ones don't. I acutally installed the 2102 device in the scope (it has a standard USB A connector), and now I can connect the scope to a PC with a terminal program any time very easy. Maybe you should try to exchange the USB-serial adapter, could save you a lot of time and headache (I should have done it as well). |