Hi,
1) Power on the Siglent and allow it to warm up for at least a half hour.
2) Set the VNA to sweep over the entire frequency range using an IFBW of somewhere around 100Hz (make a note of this). Use something around 800 to 1600 points.
2) Using your best cables and standards, perform a SOLT calibration
3) Attach a load to each port and measure S21.
4) Save the data to a Touchstone file and upload it here. You may need to rename it or ZIP it up. Maybe name the file to include the model, IFBW, range. Something meaningful.
I tried, but I cannot set the IFBW to anything other than 10kHz. The setting is greyed out and locked at 10kHz. I do have the capability to change the sweep time, I tried increasing it significantly, but it does not change the IFBW.
In the specs of the SVA1032X from Tequipment, they do write that the IFBW for the VNA feature is 10kHz:
https://www.tequipment.net/Siglent/SVA1032X/Spectrum-Analyzers/However, on the website of Siglent NA, they write "Minimum RBW 1Hz" but I suspect this applies to the Spectrum Analyzer feature and not the VNA feature.
The SVA1032X I have has all options enabled so I doubt it's an issue about software options.
I looked into the Siglent Programming Guide for this instrument. The corresponding SCPI command seems to be ":BWIDth?", I tried and it returns 10000.
I tried issuing ":BWIDth 100", it returns success, but when querying the bandwidth again it's still stuck at 10000, and it does not change on the GUI either.
So... it looks like the IFBW is locked at 10kHz.
The datasheet confirms:

So, I proceeded to still capture the data for you, with that 10kHz IFBW. See attached.
I believe the most valuable measurement is the one where I averaged 100 readings, it displays the true noise floor of the Siglent SVA1032X:
I attached all measurement raw data along with a file describing all measurement conditions.