Hi

I couldn't find much info about the blind time and update rate of analog scopes and how they compare to the latest digital scopes with Max. >100kHz waverform updates.

In summary of this measurement I would say:
-This analog scope kept going until 1µs/div while the DS2000 starts breaking down earlier at 20µs/div.
-This analog scope has a max. of 511kHz waveform updates, that is less than the new one from Agilent with 1000kHz. (The DS2000 has a max. of 50kHz 1/10 of the Analog one)
So it would be interesting to add an 1000kHz Digital Scope to this graph.
-This analog scope is setup so the blind time in never getting close to 0% (sweep of 12 div, 10.2 div visible)


Notes:
About the DS2000:
The waveform update rate is from the excel file from marmad. (smallest memory depth)
About the Tektronix 2465:
The blind time of 14% is about 1.8 divisions, or with other words one division left and one right of the screen (visible part is 10.2div)
« Last Edit: June 03, 2014, 07:38:27 am by KedasProbe »
Not everything that counts can be measured. Not everything that can be measured counts.
[W. Bruce Cameron]
 
The following users thanked this post: djnz

Offline SAUL BRITTO

  • Frequent Contributor
  • **
  • Posts: 446
  • Country: br
  • Don't Make a Mistake
Re: Waveform Update Rate Digital vs Analog (DS2000 vs Tek 2465)
« Reply #1 on: June 02, 2014, 02:24:28 am »
You want to know what is a "digital" scope,go to XY mode... AARRRGHH!!
Thank You, for all earth.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf