Author Topic: Waveforms on Fluke ESA612 Electrical Safety Analyzer  (Read 265 times)

0 Members and 1 Guest are viewing this topic.

Offline Roberts_CuriosityTopic starter

  • Contributor
  • Posts: 14
  • Country: us
Waveforms on Fluke ESA612 Electrical Safety Analyzer
« on: December 18, 2022, 06:06:15 pm »
Question: Anyone know how to connect an oscilloscope to an ESA612 to prove its 10 Hz sine wave is working?

Background:
The Fluke ESA612 Electrical Safety Analyzer, is meant to test EKG/ECG monitors.
One of its features is to output a 10 Hz sine waveform.
Its documentation ( https://www.flukebiomedical.com/sites/default/files/esa612_umeng0300.pdf  at https://www.flukebiomedical.com/products/biomedical-test-equipment/electrical-safety-analyzers/esa612-electrical-safety-analyzer ) shows how to connect a patient monitor to the ESA612 to see this waveform.

But, when I connect an oscilloscope to the connections, and set the ESA612 to 10 Hz (or any other waveform), all I get is 60Hz and high-frequency noise.
This is a recently calibrated unit;  so, I figure the error is likely my understanding of how to do this measurement.
 
All advice and criticisms are welcome.

Thanks, Bob
 