Products > Test Equipment
Waveforms on Fluke ESA612 Electrical Safety Analyzer
(1/1)
Roberts_Curiosity:
Question: Anyone know how to connect an oscilloscope to an ESA612 to prove its 10 Hz sine wave is working?

Background:
The Fluke ESA612 Electrical Safety Analyzer, is meant to test EKG/ECG monitors.
One of its features is to output a 10 Hz sine waveform.
Its documentation ( https://www.flukebiomedical.com/sites/default/files/esa612_umeng0300.pdf  at https://www.flukebiomedical.com/products/biomedical-test-equipment/electrical-safety-analyzers/esa612-electrical-safety-analyzer ) shows how to connect a patient monitor to the ESA612 to see this waveform.

But, when I connect an oscilloscope to the connections, and set the ESA612 to 10 Hz (or any other waveform), all I get is 60Hz and high-frequency noise.
This is a recently calibrated unit;  so, I figure the error is likely my understanding of how to do this measurement.
 
All advice and criticisms are welcome.

Thanks, Bob
Navigation
Message Index
There was an error while thanking
Thanking...

Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod