Grandchuck:
This unit just arrived so needs more testing. As a preliminary report, all seems to work as expected. Amplitude accuracy is nothing to rave about but is acceptable; the frequency accuracy is very good. The operation range of frequencies is as advertised.
Channel 2 can be modulated by Channel 1, as shown below.
Diabolo:
Hello,
Ok.
The CH1 frequency of 100 kHz will modulate the CH2 frequency (1080 MHz) by this value (100 kHz) if the "MOD" function of CH2 is ON.
The WB-SGI generator is sold +/- 145 $ / 135 € on Aliexpress, or Ebay.
Serious seller and fast shipping.
- https://fr.aliexpress.com/item/4000468601984.html?spm=a2g0w.12010615.8148356.22.11f0d5c7pfHHq1
------
The name WB-SG1 would suggest that a version # 2 "WB-SG2" could appear with all the options that are absent on the WB-SG1, but it would be more expensive to buy.
The FA-1 frequency counter was followed by the FA-2 frequency counter.
Diabolo
Diabolo:
Hello,
It would be interesting to see the level of harmonics with a spectrum analyzer that I do not have available. If anyone could make these measurements they would be welcome.
----
What about an FM modulated signal on CH2 ?
For example, a fundamental signal modulated in FM with a deviation of 25 kHz is not likely to resemble an FM signal with a deviation of 100 kHz.
Regards
Diabolo
Grandchuck:
Here is the spectrum of channel 2 from 35 MHz to 1.5 GHz. How does one obtain FM?
Diabolo:
Thank you for the answer.
-----
The manual says that CH2 modulated by frequency CH1 is a CH2 signal modulated in FM.