Hey folks - looking for something of a sanity check here...
I recently had to teardown & repair my SPD3303X because of what appears to be an absurd design oversight.
I was charging a large lithium ion battery (yes, through a BMS & active balancer) when we had a brownout in my building.
(Before someone shouts at me - it wasn't just banal charging, I was performing charging tests which utilized some features of the PSU).
When the PSU re-powered it went through an internal relay cycling which immediately flooded the room with the stench of conflagrated FR4.
Upon inspection, the supply had shorted the inputs on power-up (which seems an intended part of the process) and turned two PCB traces into fusible links and took out a triac (UB19, a BTA08-600C).
I've never had any other bench PSU do something like this. That includes 4 Agilents, 2 Tenmas, and a host of other cheapie uprights (I'm not saying it couldn't happen, I'm just saying it hasn't).
What on earth is the point of this shorting? Some kind of self test for the relays? Why would it not check for voltage on the output first? Why wouldn't there be fuses somewhere? (Crowbar circuits typically have fuses, no?)
Am I crazy? Is this a normal implementation and I should always externally fuse or Schottky any high-current capable sources connected to the output? Would it be silly to modify the circuit to include some chunky SMT fuses (higher than the output rating, but below "blowing traces off the board" current)? I get that adding components to a calibrated output is usually a no-no, but some beefy fuses right at the output lugs aren't likely to change the specs, I'd wager...
This seems like madness to me, and cost me a day of repairs - and I'm frankly afraid to use this thing for anything with batteries or big capacitors connected to the output.
Forgive the somewhat scrubby repairs, I went back and cleaned them up after I verified the supply was again functional, but I didn't take any pics.