Author Topic: How to check a 230VAC sine/modified wave with an oscilloscope (FNIRISI 1013D)  (Read 5873 times)
Offline juan3211Topic starter
Hi guys, I am newbie with oscilloscopes.
I have an 1013D between Aliexpres and my home.
I have a doubt about this.
I want to check sine waves from several inverters. (230VAC)
I have read that with the 10x probe, this oscilloscope is capable of measure up to 400v
In oscilloscopes' specs, are these values peak or peak to peak (I mean, from -400 to 400 or -200 to 200)?
I have found different answers
This is important. If 400v specification is from -400 to 400, then I could use 10x probe to check pure sine inverter wave form
If 400v is peak, then I can't
Another question is: imagine that the previous answer is that I can measure only 400v (-200 to 200). Is it better a 100x probe or a 10x little transformer and use the included 10x probe ? Does transformer change/disturb/smooth the output sine wave of an inverter ? Even with load connected to inverter ?
Could you guys please ask me first answer? And if I need a 100x probe, for my use may be better/cheaper a little transformer that, besides, I already have ?
OF COURSE, i will use it in battery mode, without connecting anything to USB port.
I also know about risks of such high voltages.
Thanks a lot.
SPECS:
« Last Edit: March 19, 2023, 01:42:43 pm by juan3211 »
Offline wasedadoc
Re: Who to check a 230VAC wave with a FNIRISI 1013D
« Reply #1 on: March 18, 2023, 11:03:49 pm »
i will use it in battery mode, with connecting anything to USB port.
...withOUT connecting anything to USB port.
At such low frequencies you can get adequate performance by using a few suitable resistors to make a voltage divider.
« Last Edit: March 18, 2023, 11:06:53 pm by wasedadoc »
Offline alm
« Reply #2 on: March 18, 2023, 11:21:01 pm »
I'd expect 400V max to be 400V rms, but on these scopes I wouldn't want to push them to their maximum ratings, so I wouldn't use 10x probes for probing mains.
I'd say a transformer is better from safety point of view (it will provide galvanic isolation in addition to attenuation), but a 100x probe will give better bandwidth. I'd expect a small mains transformer to have a bandwidth up to a few kHz, so I would expect that to be fine to get a decent number of harmonics. If you're not using a transformer, then I'd make sure you are not connecting the ground lead to anything other than ground. Not neutral, not phase, just ground. With a transformer you can clip the ground lead on either side of the secondary of the transformer. Without transformer, you may have to use two probes in subtract (CH2-CH1) mode if you need to prove both phase and neutral, although you may get a good enough signal if you just measure between ground and phase.
Even if you have a 100x probe I think a transformer would still be useful for the galvanic isolation, though you would probably just use a 10x probe with the transformer.
Offline wasedadoc
« Reply #3 on: March 18, 2023, 11:40:50 pm »
I would not expect that.  Scopes are used for looking at all kinds of waveforms not just sine waves.  If looking at a waveform that is alternating between +400v and -400v (both "flat tops") its RMS is 400v.  The sqrt(2) factor does apply in that case or more generally.
My interpretation of the spec is 400v peak from zero in either direction.  INCORRECT - See update below.
Also my interpretation of the opening post is that the 230 v is not AC mains and the concepts of Live/Neutral or Phase/Ground etc may not apply.  The OP could clarify that.
Update:  Read the extract from the 1013D's manual.
« Last Edit: March 18, 2023, 11:55:28 pm by wasedadoc »
Offline alm
« Reply #4 on: March 19, 2023, 12:17:18 am »
That's true, and maximum sensitivity is normally about peak voltages, but absolute maximum voltages, similar to CAT ratings, are normally RMS. Look at the ratings of the more reputable scope vendors. Clearly as you've shown they deviate from this practice. But then I would take the absolute maximum ratings for these scopes with a bucket of salt anyway.
An inverter will still produce single phase power, and I'd hope that for safety reasons one side is tied to ground inside the inverter. But more clarification would be really helpful indeed.
Saying things like don't connect anything to USB makes me worry, because to me that would mean touching the scope could shock you.
« Reply #5 on: March 19, 2023, 01:01:16 am »
Sorry, original post corrected.
So there is a new option: simple voltage dividir made with resistors. Thanks.
Do you think it will be better a transformer (I already have a little one) ?
« Reply #6 on: March 19, 2023, 01:02:52 am »
Thanks, so you proposal is use transformer and 10x probe.
Thanks a lot for your suggestions.
« Reply #7 on: March 19, 2023, 01:07:20 am »
It is incredible.
 I haven't still recieved it but two different YouTubers have confirmed me that they have already measured 230VAC mains waves.
Thanks a lot for your information. When I receive the unit I will read manual and I hope read this.  :clap:
So no 10x probe measuring directly main's circuit. Thanks.
« Last Edit: March 19, 2023, 01:26:50 am by juan3211 »
« Reply #8 on: March 19, 2023, 01:16:03 am »
So if I want to check my home circuit (ground +neutral+phase, generally neutral is connected to ground in grid's substation) how do I need to connect transformer secondary outputs ? What is the correct option?
A. One transformer output to ch1, second output to ch2, main's ground to ground clips of the two channels. (Probably I will see one channel as a plain signal, as ground is connected to neutral)
B. One transformer output to ch1 and ground clip to the second output (it doesn't matter nuetral or phase).
And in a case of neutral+phase of an inverter ?
A. Like A, clips connected to home grid grounds
B. Only connect to transformer secondary outputs. Only one channel is used.
Could you help me? May be there is an better option C ?
Thanks.
Offline nctnico
« Reply #9 on: March 19, 2023, 01:39:43 am »
For as long as you clip the ground clip of the probe to ground (the actual ground!) you can use a 100x high voltage probe (with a CAT rating for mains voltages) to measure between ground and phase. If you want to measure between neutral and phase, then the only safe solution is to use a CAT rated, high voltage differential probe. The Fnirsi 1013D is not designed to be used as a floating / isolated oscilloscope. It does not provide the necessary isolation barrier to be safe to use / touch / operate even when running from it's internal batteries. Battery operated oscilloscopes are typically NOT suitable to measure dangerous voltages. Only when the models that have dedicated, high voltage CAT rated isolated inputs (IOW: a high voltage differential probe built in), they are safe to use on mains directly. The tell tale sign is the price; oscilloscopes with isolated inputs are very expensive (starting from ball park US $800).
« Last Edit: March 19, 2023, 01:46:54 am by nctnico »
« Reply #10 on: March 19, 2023, 06:06:20 am »
Thanks a lot
With all your suggestions, it is clear for me that I can't use the 1013D "directly".
I will use with a transformer.
What do you know about the options that I have proposed in my previous message ?  (A, B, ...C ?)
I already have a transformer, so I don't need to buy an expensive 100x probe or differential probe
Offline modoran
« Reply #11 on: March 19, 2023, 10:55:38 am »
Specs you linked are fake, already proven by various eevblog members. Do not trust what fnirsi says in their advertisements.
Yes, using a isolated transformer and measuring only secondary output will be fine ( except the signal amplitude, of course ).   But beware that not all "inverters" output sine waves, most of them don't.
Offline wasedadoc
« Reply #12 on: March 19, 2023, 11:51:41 am »
If you only want to determine the waveform shape you only need a few pieces of insulated wire. Wrap 10 cm of one around a wire from one pole of the inverter output  Likewise wrap another around a wire from the other pole. Connect the 10x probe tip to one of the wires - not an inverter output.  Likewise connect probe clip to the other wire.  There will be enough capacitive coupling between inverter output and probe to safely give a small signal to the scope.
If you need to know the effective amplitude use a true RMS voltmeter.
Most low price inverters give "modified sine wave".  Just a few discrete levels.  Fine for switch mode PSUs and purely resistive loads.  May not be good for 50/60Hz transformers, motors etc. Heed any such warnings in the inverter's user manual.
« Last Edit: March 19, 2023, 11:57:20 am by wasedadoc »
Offline WackyGerman
« Reply #13 on: March 19, 2023, 12:12:34 pm »
In the manual you get all the information you need . You have to to prepare your own 100x probe for doing this measurement . Go to "http://www.fnirsi.cn/support" and then choose FNIRSI 1013 D English manual.pdf
« Reply #14 on: March 19, 2023, 12:16:40 pm »
That is precisely what I want to check. Pure sine waves, or modify ones.
What about connection to the secondary of transformer?
Thanks
« Reply #15 on: March 19, 2023, 12:18:51 pm »
Thanks this is a new option, but as I already have the transformer, how do I need to connect the secondary? Thanks
« Reply #16 on: March 19, 2023, 12:23:19 pm »
This is the transformer what I have. Tyco
Offline wasedadoc
« Reply #17 on: March 19, 2023, 12:29:28 pm »
I already attached a relevant extract from the manual in Reply #3.
« Last Edit: March 19, 2023, 12:32:34 pm by wasedadoc »
Offline WackyGerman
« Reply #18 on: March 19, 2023, 12:40:06 pm »
Oops , you`re right  :-DD
Offline Fungus
« Reply #19 on: March 19, 2023, 01:11:23 pm »
A transformer can change the shape of the wave. The most accurate way is with a 100x probe... or make a box with a resistor divider in it.
« Last Edit: March 19, 2023, 01:13:46 pm by Fungus »
« Reply #20 on: March 19, 2023, 01:41:53 pm »
Thanks a lot, here, in a previous post  I asked for this topic , but someone has ansemwer me that I could lost few armónicos but nothing about changing the shape.
So then the use of transformer is not an option for you, is it ?
I have changed the topic of the post to help.other with same question.
« Last Edit: March 19, 2023, 01:43:24 pm by juan3211 »
Offline wasedadoc
You will easily distinguish between pure and modified sine wave by just placing the scope probe tip close to an insulated cable from the inverter output.  No connection required.
Or two multimeters, one true RMS and one not.  (Most analogue meters are not).  If they read the same within meter tolerances it is pure sine.
« Last Edit: March 19, 2023, 01:56:17 pm by wasedadoc »
But as you know we always want more, and I want to see the curve and compare curves of different inverters, ....
Hi, what do you think is better ?
Make my own 10x divider with 10 X 1 Mohm resistors and use the 10x probe (total 100x, and I don't care of I measure 232 or 235 because of different values in resistors)
Or a transformer 230/48 and measure with 10x probe ?
I think I will use one of these options. What do you think?
I hope you understand that I don't want to buy an expensive 100x probe afte buying a cheap oscilloscope, and only for a few test.
Offline Fungus
I would use 200k resistors so your 'scope doesn't affect the circuit so much.
Just don't cheap out on the precautions. Be aware of polarities, etc. It might even be wise to measure across a resistor in the middle of the chain so there's always a megaohm or so between you and the electrons. Everything else should be in a box with just that one resistor exposed.