Why does mains transient tester only get up to 1600V when set for 2kV?

(1/3) > >>

Faringdon:
Hi,
The attached shows a burst mode transient voltage applied to our mains connected equipment.
This comes from a Schaffner BEST  burst mode tester.
I set it to 2kV Peak...but as can be seen, its only 1600Vpk. (the zero ref for the "yellow"
channel is at -1600Vdc).
Why did it not get up to 2kV?
I measured it with a 100:1 probe connected from true mains earth (not the earth coming out of the Schaffner)
and Line. (Line that  comes out of the Schaffner)
The Schaffner was set up as follows....
_2kv peak
_No of burst pulses = 1
_Phase angle of burst = 90 degrees
_Transient coupled to Line only.

So why did it only get up to 1600V?
Was it due to the MOV etc?

joeqsmith:
That's some beginner stuff there.  The generators are spec'ed for open circuit voltage and short circuit current waveforms (combo).    They have a specified source impedance as well.  Drive it into a dead short and the voltage will be less.   

You really would be better off renting some time at a lab. 

Faringdon:

--- Quote ---The generators are spec'ed for open circuit voltage and short circuit current waveforms (combo).
--- End quote ---
Thanks,  say we have a Boost PFC'd power supply with 800uF of electrolytics after the booster. Then i take it that we can set 2kV, but will get way, way under that.
As you kindky discuss, it will clamp at the current limit. I must admit, i didnt see from the manual what that was. I think those pages are missing.
(theres no manual online for the Schaffner BEST)
Thanks but We cant afford to pay for lab time.

joeqsmith:

--- Quote ---(theres no manual online for the Schaffner BEST)
--- End quote ---

As you stated in your previous thread:
https://www.eevblog.com/forum/testgear/schaffner-best-transient-tester/


--- Quote ---Thanks but We cant afford to pay for lab time.
--- End quote ---
Can you afford not to?

I would just run it open circuit, and then with some resistive loads until you at least understand some of the basics.  With it being so old an unsupported, are you sure it even meets the standards you are required to adhere to?  Or is this more just for fun and you really are not trying to meet any of the current standards?

floobydust:
I am familiar with Teseq ones, super easy to use so techs can do the testing. It's a very high cost of ownership with commercial mains transient generators.
A new standard rolls out, or an update to a standard and you have to send it in for hardware mods- or the unit just goes obsolete. Steady churn of these units on the surplus market.
Calibration is a constant cash drain, for test-lab requirements. New, they are the price of a car.
That's why I wanted joeq to give details about the insides of his homebrew one ;)

If output voltage is low, make sure the unit is capable of that in the first place. It could be for older (standards).
It could need repairs, have some damage and only be making 1.6kV not 2kV, or have high output resistance. Or there is an arc in the DUT.
I'd test it with a simple resisitive load to make sure it's working properly.

Most important is safe test leads and the blast shield, on an insulated bench. You have to know what you are doing. We wouldn't want a dead Treez.