The TekVPI oscilloscope probe connector, the BNC connector with a 2 by 3 grid of pogo pins, has a 250 kΩ mode in addition to 50 Ω and 1 MΩ. Does anyone know why?
It seems to be a hidden feature in oscilloscopes with the TekVPI interface. The feature can only be activated with certain TekVPI probes. The only place where I found a reference to it is in the specifications for the 4 Series MSO
https://www.tek.com/en/datasheet/4-series-mso. The spec for DC gain accuracy says, "1 MΩ and 250 kΩ: ±1.0%".
Probes that make use of said feature include some TekVPI probes with with 10 MΩ // 4 pF input impedance, for instance the TPP0500 and TPP1000. The manufacturer does not seem to say they are 10X. TekWiki says they are 10X and classifies them as 1 MΩ probes, but there is no citation. By measuring them, I found that they have 40X attenuation, and that the oscilloscope would switch its input resistance to 250 kΩ when one is attached. When not connected, they measure 9.75 MΩ end to end, and open from the tip to ground. When connected, there is 10 MΩ from the tip to ground. When the replaceable tip hybrid is removed and a voltage is fed to the socket, 40 times the voltage shows up on the scope.
It makes sense that they want a 40 times attenuation for less input capacitance. It also makes sense that they may want to make input hybrids with a similar resistance to what they already know how to make instead of figuring out how to make 39 MΩ ones. But when the TekVPI connector has plenty of space for a 330 kΩ resistor, the only reason I can think of for putting it in the scope is maybe they did not think figuring out how to switch to 1 MΩ without compromising the 50 Ω path was enough to show off their competence.