| Why doesn't my RMS multimeter agree with my Rigol 1054Z scope on the AC VRMS? |
| << < (4/5) > >> |
| 2N3055:
Quite a discussion about something without basic data: type of sig-gen and details on settings, and complete lack of ANY information on scope and it's settings. |
| joeqsmith:
--- Quote from: 2N3055 on August 10, 2023, 01:31:45 pm ---Quite a discussion about something without basic data: type of sig-gen and details on settings, and complete lack of ANY information on scope and it's settings. --- End quote --- I was guessing AI generated on many of the 1st time posters like this one. |
| 2N3055:
--- Quote from: joeqsmith on August 10, 2023, 02:01:54 pm --- --- Quote from: 2N3055 on August 10, 2023, 01:31:45 pm ---Quite a discussion about something without basic data: type of sig-gen and details on settings, and complete lack of ANY information on scope and it's settings. --- End quote --- I was guessing AI generated on many of the 1st time posters like this one. --- End quote --- You're probably right.. |
| Datman:
A month ago I wrote to Rigol: "On the DS1054Z I noticed that RMS value is wrong if I apply a square wave and duty cycle if different from 50%! If I select AC for the input, it seems to be OK, but if I select DC the value is wrong and it doesn't change when I adjust the duty cycle! In AC, the value changes if I change the frequency of the square wave, displaying 2 or 10 or 20 or more waves. The scope should calculate RMS only on complete waves... Can you verify?" They asked Zhou Lihui to verify... |
| Datman:
I found something from 2015: :( http://www.sonsivri.to/forum/index.php?topic=60363.0 (...) "Not sure if I will be able to convince them about changing the way calculations are done to match what tektronix and agilent do, but I'll try" |