Author Topic: why Fluke use bandpass filter in their DMMs?  (Read 2369 times)
Offline emm386Topic starter
« on: October 04, 2023, 05:46:54 am »
Hi all,
While I was skimming over the service manual of Fluke's 87-V and 189-II (figures attached), I realized that they have a bandpass-looking filter in their custom IC.
According to the Fluke 87-V service manual, it is referred as a low-pass filter, but its feedback structure makes it looking like an active bandpass filter (reference: https://cheever.domains.swarthmore.edu/Ref/FilterBkgrnd/Filters.html).
Can anyone please explain me the purpose of this filter and why this is configured as a bandpass filter ?
Plus, where would the 'FAMP_O' pin in the first attachment be connected?
« Last Edit: October 04, 2023, 05:52:14 am by emm386 »
Online Kleinstein
Re: why Fluke use bandpass filter in their DMMs?
« Reply #1 on: October 04, 2023, 07:28:13 am »
The filter as 2nd order low pass filters. It is a configuration often found in DMMs, but not so much in the standard books.
In the first schematics C27+C36+R78 and the Filter op amp form a kind of "super (2nd order)" capacitor.
With a cross over at some 10 Hz (crude estimate) this filter is likely there to suppress mains hum (as part of the signal) before the gain stage.
« Reply #2 on: October 04, 2023, 07:41:00 am »
Could you explain me how this forms a second order filter? Even in my LTSpice simulation this shows BP response...
Online Kleinstein
« Reply #3 on: October 04, 2023, 08:20:57 am »
The low pass ouput is not from the output of the OP-amp, but from the right of R2.
The following users thanked this post: robert.rozee, emm386
Offline MegaVolt
  • Country: by
« Reply #4 on: October 04, 2023, 12:18:46 pm »
Perhaps these images will help you understand these filters.
The following users thanked this post: Exosia, robert.rozee, tridac, emm386
« Reply #5 on: October 04, 2023, 05:39:56 pm »
MegaVolt and Kleinstein, thanks for the clarification and answering my question! this is very helpful
Online David Hess
  • DavidH
« Reply #6 on: October 04, 2023, 09:11:55 pm »
The advantage of this filter configuration is that it is DC accurate; the DC offset voltage of the amplifier does not contribute to the output.  There is a similar configuration for a DC accurate precision rectifier which is AC coupled on its input and output.
« Last Edit: October 04, 2023, 09:25:43 pm by David Hess »
Offline tszaboo
« Reply #7 on: October 04, 2023, 10:36:32 pm »
Looking at the second picture, my question is rather:
Why do they have so many lightbulbs in a multimeter?
:-DD
Online David Hess
  • DavidH
« Reply #8 on: October 05, 2023, 07:08:56 am »
As shown below, what light bulbs in multimeters might look like.
Online Kleinstein
« Reply #9 on: October 05, 2023, 07:48:25 am »
In a few cases light bulbs are used as a low current fuse.
In the plan from the start the circle with x symbol is not for light bulbs but for CMOS switches.  So no light bulbs but chip internal CMOS switches for the signal.
P.s. : A light bulb is not a good fuse and may fail with an arc flash. It is not uncommon for a failing mains bulb to take out the breaker from that.
« Last Edit: October 05, 2023, 07:50:26 am by Kleinstein »
Offline MegaVolt
  • Country: by
« Reply #10 on: October 05, 2023, 10:19:40 am »
Do you have an example circuit diagram?
Offline tridac
« Reply #11 on: October 05, 2023, 12:15:09 pm »
Been trying to find more info on this filter design. Asked in another thread and pointed towards a Burr Brown third gen op amp techniques book, since found on ABE books. Still interested in a more thorough treatment, so which book or source did you find the above 5th order example ?.
Thanks, still collecting data...
« Last Edit: October 05, 2023, 12:20:12 pm by tridac »
Test gear restoration, hardware and software projects...
Offline MegaVolt
  • Country: by
« Reply #12 on: October 05, 2023, 12:26:45 pm »
The following users thanked this post: tridac, emm386
Online David Hess
  • DavidH
« Reply #13 on: October 05, 2023, 12:40:06 pm »
The Tektronix DM501 and DM502 use it with a 318 type of operational amplifier which otherwise would contribute large DC errors.
The following users thanked this post: MegaVolt, tridac
Offline tszaboo
« Reply #14 on: October 05, 2023, 03:40:47 pm »
Come on, take a look at the picture, there are at least 20 light bulbs in that multimeter according to the schematic.
Or someone used that symbol as the selector switch's switching part as a lightbulb. If you look at it, 29-30-31-32 clearly sets a gain of an amplifier.
Offline joeqsmith
« Reply #15 on: October 05, 2023, 04:04:38 pm »
internal switches, not lamps.
check here for example:
http://www.cyrustek.com.tw/en/wp-content/uploads/ES51997.pdf
« Last Edit: October 05, 2023, 04:08:15 pm by joeqsmith »
Online David Hess
  • DavidH
« Reply #16 on: October 05, 2023, 05:05:17 pm »
Any light bulb is a switch, if driven hard enough, and the reverse.
Online TimFox
« Reply #17 on: October 05, 2023, 05:18:51 pm »
Symbolism can be difficult.
Offline tridac
« Reply #18 on: October 06, 2023, 03:45:50 pm »
There are references in that article, one (9), being Analog Filter Design by Val Valkenburg. Looked that up and lloks like it may have material on the filter topology of interest. Found a copy on ABE Books for less than $10, soon to be on it's way to me...
