Products => Test Equipment => Topic started by: matt09 on December 12, 2018, 09:51:33 pm
-
Hi all,
Recently acquired this bit of kit, not really my area of expertise so unsure where to go with it.
I have substituted the dead NiCad battery but I still get the error:
TEST FAILED
(A4) warning - ramp not calibrated
CH2 LED flashes on power up which does indicate an A4 - sweeper interface error.
Does anyone have experience with kit of this era?
Have glanced through the service manual but it doesn't mention this and cannot find a copy of the regular manual.
Cheers
-
Anyone?
-
bump!
-
New years bump, someone must have used these?
-
Please guys :-//
-
Ramp is a sawtooth frequency sweep signal (usually 0-10V) generated either by analyzer itself or by external frequency sweeper.
In your case there are 2 variants:
1. Analyzer is configured to use external ramp and misses it.
2. Internal ramp generator is faulty.
Look through the user and service manuals (both are freely available - Google for them).
-
Yes I know the board is for sawtooth generation, I've looked at both manuals but neither seem to give specific information regarding a procedure to verify the unit via test points. This would imply there is a third type of test and validation manual, though I can find no mention of it.
-
Ramp generation is a part of the scan cycle.
So at least is the scan cycle itself working?
Any LED blinks each cycle?
Any blank/channel_switch strobes on the outputs?
Will analyzer accept external ramp signal (0-10V with potentiometer)?