I have an opportunity to buy a Keithley 196 for U$ $175,00 with freight is gonna cost me U$ $250,00
It's calibrated, the only issue is that the latch mechanism in the switch button does not work, so only the read plugs are working. to be fully functional the switch must be changed.
Should I go for it?
Thanks,
Vini
« Last Edit: February 28, 2019, 12:50:10 pm by vmiguel »
Offline TiN
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: Working Keithley 196 U$$250 is a good deal?
« Reply #1 on: February 28, 2019, 12:46:36 pm »
For 250$ you can find K2000 or HP34401, if you patient and willing to take some risks. :)
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
The following users thanked this post: vmiguel
Offline pigrew
« Reply #2 on: February 28, 2019, 01:17:13 pm »
I agree with TiN that your money is probably better spent on a newer unit.
However, the 196 does have offset compensated ohms measurement, which the other meters lack.
Offline LapTop006
« Reply #3 on: February 28, 2019, 01:56:54 pm »
I bought one for the work lab a while back, works fine.
It makes a nice meter that won't walk. Sure it's no Keithley 2000/6500, but I like it.
(Yes I was somewhat cheating for this photo)
Offline EEVblog
« Reply #4 on: February 28, 2019, 02:03:03 pm »
They go for a fair bit cheaper:
Offline coromonadalix
« Reply #5 on: March 01, 2019, 02:49:49 am »
Snatched the 115$ one, but the top case is broken, miss a big piece,  want to find a donnor
They have the same case in many models ...
Offline kawal
« Reply #6 on: March 15, 2019, 04:25:29 pm »
I was looking at that one myself.  Glad you got it. Post a review when you get it.
« Reply #7 on: March 15, 2019, 04:48:19 pm »
Backed the calibration chip and re-write it a few times, it`s on the b14 firmware, i will upgrade it to b17
Searching for precision resistors  for a full range calibration IE :  the manual ask for 190, 1.9k 19k, 190k, 1,9 meg, 19 and 190 megs, but
Im going for: 300 ohms, 3k, 30k, 300k, 3meg, 30 meg and 300 megs, since the dmm will top at 3,030 value display    Boy they are priceyyyyyyyy  8-10$ a resistor at 0.02%, i do know you can calibrate it, with known resistors values and adjust the dmm accordingly.
Checking some Vishay foil resistors ...
Doing a Full capacitors recap,  there`s only 1 chimical cap and 1 tantalum cap on the meter board,  very nice :)
Sad to be unable to reverse the calibration chip coding loll ,  i`ll push the 196 against my recently calibrated 34401a i`m selling
The front rear switch is not good, i`ll remove it and the rear input terminal xformer coil(s)  and set it permanently for a ``front`` use.
Finally  I need to find a top case  (wtb post), since it miss a big part on the rear, thats why i haven`t paid a lot for it  loll       I`m the final user  loll
Online Kleinstein
« Reply #8 on: March 15, 2019, 05:22:46 pm »
The 196 Ohms modes don't use constant current sources, but a kind of bridge circuit with the DUT and reference resistor in series. So they will not necessary have the best accuracy at the full scale range. It's more like best accuracy near the center of the range, or the value of the reference resistors.
A value near 300 Ohms and 3 K may be OK, but 300 K ... 300 Meg are likely not such a good idea.
For the 300 M range I would even prefer some 50 M.
Offline Conrad Hoffman
    • The Messy Basement
« Reply #9 on: March 15, 2019, 07:03:43 pm »
Can't remember which one I had, but that sure looks like the one where the power supply pass element runs at the surface temperature of the sun. Even with a DIY heat sink it ran hot. The unit wasn't all that stable either. Sold it cheap. Might not be the same unit, but be sure to check how hot the supply runs.
« Reply #10 on: March 15, 2019, 08:44:04 pm »
okay   i should use the same values in the manual ...   thks
Offline macboy
« Reply #11 on: March 21, 2019, 04:09:40 pm »
Also keep an eye out for HP/Agilent/Keysight 34970A. This is a "data logger" which means it is a DMM with built in switch/mux system (OK, technically it is a switch/mux with a built in DMM). The DMM was an option, so make sure it (and a 34901A mux card) is present before buying. Most have it. The DMM is a 6.5 digit one, and is almost identical to the 34401A in both functionality and specifications, but it adds things like offset compensated Ohms and temperature measurements (with thermocouples, RTDs, etc.). The current measurement might be slightly limited (1 A max?) but check the specs yourself if it's important.  These are usually cheaper than 34401A DMMs because of lower demand, even though their original cost was much higher.
One good (?) thing about the 196 is the through-hole construction which generally makes repair easier. And few people have ever cursed at a big bright 7-segment LED display. That is usually reserved for poorly lit LCDs, like what HP liked to use in the late 70's-80's. sigh.
« Reply #12 on: March 21, 2019, 07:11:38 pm »
Just checked on Fleabay, they sell pretty high,  even with no card installed ????
« Reply #13 on: March 27, 2019, 09:01:27 am »
the 34970a  dmm section goes to 1 amp, 300vdc, 300vac,  100meg ohms, and rated at 0.004%  precision
the dmm section is optional,   have to be careful if you buy theses units