1 3 When calibrating fusion or focus using 2 targets, the focus motor position for the far target has to represent a further position than motor position for the near target. The check for this has failed. Likely Cause: Failed autofocus operation. One or both measurements are missing. If manual focusing, bad focusing by operator.
2 When calibrating fusion or focus using 2 targets, the metric distance given for the further target must be larger than the distance given for the closer target. Likely Cause: Bad input of metric distances.
4 Failed to create focus branch in calibration. Likely Cause: Bad camera configuration. Software bug.
6 Computed calibration constants for focus ( C2 and/or C3 ) is outside spec. Likely Cause: Failed autofocus operation. Bad optics HW. If manual focusing, bad focusing by operator.
7 The focus motor position for the far target represents a position at the furthest possible position. Likely Cause: Bad focus calibration in "Core work center". Failed autofocus operation. Bad optics HW. If manual focusing, bad focusing by operator.
8 The focus motor position for the near target represents a position at the nearest possible position. Likely Cause: Bad focus calibration in "Core work center". Failed autofocus operation. Bad optics HW. If manual focusing, bad focusing by operator.
9 No or bad name read from lens. Likely Cause: Bad connection to lens. Bad data in lens.
10 Failed to save calibration after focus or fusion calibration. Likely Cause: Bad camera configuration( /FlashFS/system/appcore.d/config.d/conf.cfg ). Software bug.
11 When calibrating fusion using 2 targets: IR to visual
camera rotation error is inconsistent. Likely Cause: Bad image(s) of
fusion targets. Check: IRImageNear, IRImageFar VisImageNear, VisImageFar. If
manual calibration, bad positioning of spots by operator.
12
When fusion or focus calibration: (Far) measurement missing.
13
When fusion or focus calibration: Lens replaced during calibration.
14 When calibrating fusion using 2 targets: IR to visual camera scale factor
error is inconsistent. Likely Cause: Bad image(s) of fusion targets. Check:
IRImageNear, IRImageFar VisImageNear, VisImageFar. If manual calibration, bad
positioning of spots by operator.
15
Laser positions are NOT measured. Likely Cause: Operator mistake.
16
Far Laser measurement missing. Likely Cause: Operator mistake.
17
Near Laser measurement missing. Likely Cause: Operator mistake.
18
Failed to take visual image. Likely Cause: Bad or missing vis camera. Vis
camera not enabled. Image frozen
19
Fail when detecting spots in visual image when calibration fusion. Likely
Cause: Bad image(s) of fusion targets. No illumination of fusion target. Check: VisImage, VisImage2.
20
Failed to take IR image. Likely Cause: Image frozen.
21
Fail when detecting spots in IR image in fusion calibration: . Likely Cause: Bad
image(s) of fusion targets. Spots outside active areas due to pointing error.
Check: IRImageNear, IRImageFar.
22
Fail when detecting laser spot in laser calibration . Likely Cause: Bad laser.
Bad visual camera. Laser pointing err. Visual camera pointing error. Check image
LaserNear, LaserNearSpot, LaserFar, LaserFarSpot.
23 Position measurements missing in dual visual camera alignment calibration. Likely Cause: Operator mistake.
24 Scale factor outside spec in dual visual camera alignment calibration. Likely Cause: Bad visual camera(s). Visual camera(s) pointing error. Bad target. Check images VisImage1, VisImage2. If manual calibration, bad positioning of spots by operator.
25 Rotation angle outside spec in dual visual camera alignment calibration: Likely Cause: Visual camera rotated. Bad visual camera(s). Visual camera(s) pointing error. Bad target. Check images VisImage1, VisImage2. If manual calibration, bad positioning of spots by operator.
26 Vis to IR camera scale factor outside spec in fusion calibration. Likely Cause: Bad visual camera. Bad IR image. Vis or IR spots outside active areas due to pointing error. Bad target. Bad Shutter. Check images IRImage, VisImage (OrgIRwAreas). If manual calibration, bad positioning of spots by operator.
27 Rotation angle outside spec in fusion calibration. Likely Cause: Visual camera rotated. Visual camera pointing error. Bad target. Bad Shutter. Check images: IRImage, VisImage. If manual calibration, bad positioning of spots by operator.
28 Scale factor outside spec in fusion calibration. Likely Cause: Visual camera rotated.. Visual camera pointing error. Bad target. Check images: IRImage, VisImage. If manual calibration, bad positioning of spots by operator.
29 Storage card is bad, missing or write protected. Replace, insert or unprotect storage card.
30 Failed to copy calib.rsc from backupFromBAL or backupFromKaKa. Likely Cause: Camera not run in BAL. Camera not run with standard lens if equipped with extra lens
31 Bad calibration. Likely Cause: Missing or corrupt calib.rsc file.
32 Storage of radiometric images not enabled. Likely Cause: Wrong or corrupt configuration in camera.
33 MBox measurements not enabled. Likely Cause: Wrong or corrupt configuration in camera.
34 Error when reading GSK from detector. Likely Cause: HW problem
35 Unable to find GSK. Likely Cause: HW problem
36 Pixel clusters with irreplaceable bad pixels. Likely Cause: HW problem. Bad detector. Check: BadDeadMap
37 Bad pixels where only 1 suitable replacement pixel was found. Likely Cause: Large clusters of bad pix( not a reason to fail detector( yet?)
38 Large number of bad pixels. Might cause "Time slot exceed error" in KaKa Likely Cause: HW problem. Bad detector. Check static map for the case
39 No lens detected. Likely Cause: Bad program in lens. Lens not in place
40 When creating gain map. The cold image was not significantly colder than the warm image. Likely Cause: Bad shutter?. If run in KaKa: Camera failed to detect trigBB. Check images: ColdImage, WarmImage, TriggerBB. Also check temperature and response from MBox 1 ( image.sysimg.measureFuncs.mbox.1.avgT )
41 Failed to read prodspec.rsc file. Likely Cause: Missing or corrupt file. Bad checksum in file
42 Too many bad rows Likely Cause: ?
43 Too many bad columns Likely Cause: ?
44 Too few visual spots detected( warning ), Will cause problems if wide FoV lens used. Likely Cause: Vis camera pointing error. Check VisImage
45 Too few visual spots detected, Likely Cause: Vis camera pointing error. Bad visual camera. Check VisImage
46 Failed to detect spots in vis camera 1. Likely Cause: Bad visual camera. Insufficient illumination of fusion target. Check VisImage1
47 Failed to detect spots in vis camera 2. Likely Cause: Bad visual camera. Insufficient illumination of fusion target. Check VisImage2
48 A focus operation time out. Likely Cause: ?
49 Failed to set live image. Likely Cause: HW error
50 Result of calibration control measurement outside spec. Possible lens/aperture problem. Check results.xml
51 Result of operability control outside spec. Likely Cause: Detector or optics HW problem . Check results.xml, BadOpMap
52 Camera not failed despite result of operability control outside spec. Check results.xml, BadOpMap
53 Result of NETD control measurement outside spec. Check results.xml
54 FpaGain outside spec. Possible detector or optics HW problems.
55 This camera has been configured for customer. It is not possible to run it in KaKa without restoring directory /FlashFS/tmp
56 Bad optics. Likely Cause: Optics out of specification. Operator mistake. For some problems check XYMap.gan
57 Date not set. Check that battery backup for dataandtime is working
58 Failed to freeze image when checking MRTD Likely Cause: ?
59 Failed to read image when checking MRTD Likely Cause: ?
60 Bad background or contrast in image when checking MRTD Likely Cause: ?
61 Result from dual cameras or fusion calibration is outside specification
62 Time slot exceed reported from camtorrent.
63 Unknown command or Bad resource value sent to prodapp. Likely Cause: Syntax error in kaka-script. Obsolete prodapp or prodspec
64 Calculation of RBF contstants failed. Likely Cause: F constant not optimal. Bad measurements for radiometric calibration. Check files: caldata.log and prod.log
66 Resource \"rawPosition\" not implemented. Update FVD and/or FPGA
67 Fusion distance range outside specification. Might cause bad fusion at far or near distances. Likely Cause: Visual camera pointing error
68 Fusion not calibrated. Reason unknown
69 Result of dual cameras calibration outside spec. Likely Cause: Visual camera pointing error
70 Fusion calibration outside spec. Likely Cause: Bad focusing in KaKa or By operator if service web )
71 Lens failed to locate target in BAR test. Likely Cause: Bad optics
72 Unable to "fusion" at close distance. Likely Cause: Visual or IR camera pointing error
73 Bad image defect "Blob". Likely Cause: Dust or defect on detector window
74 Bad tdrift image( too many saturated pixels ). Likely Cause: Bad BAL run
75 Focus max/min position not found. Likely Cause: Function GetFocusLimits() not run
76 Dual visual cameras calibrated. Reason unknown
100 BAL Run terminated OK. Ready for KaKa
101 KaKa Run terminated OK. Ready for image control
102 KaKa Run terminated OK with special lens.
103 Image control after KaKa Run terminated OK
G01 The camera is in the wrong state for the requested operation. Likely Cause: The "Init" command was not executed for the sub-system
G02 Resource access error. A requested resource was not found or not accessible. Check prod.log for further information.
G03 Memory error. Likely Cause: SW or HW bug.
G04 Unknown command. Likely Cause: "typo".
G08 Failed to create a coarse map. Likely Cause: Detector or HW problem. Check image: BadCoarse.
G09 A generic calculation error. Result of a calculation is outside spec or bad of unknown reason.
G10 Computation of calibration constants during "onChip" or radiometric constants failed.
G11 File access error. Reading or writing of a file failed. Check prod.log for further information.
G13 Failed to grab a IR or visual image. Likely Cause: Camera in vis image mode. Image frozen. HW or SW bug.
G16 A selected calibration case is missing. Failed to enter the selected calibration case. Likely Cause: Camera in vis image mode. Image frozen. HW or SW bug.
G17 Failed to created a gain map. Likely Cause: Too may bad pixels. Input cold and/or warm image bad. File read/write error. Check GainMap, WarmImage, ColdImage.
G18 Failed to create a dead pixels map. Too may bad pixels. Irreplaceable pixels. Check DeadMap, KillPixelsBad
G20 Failed to set a measure area. Likely Cause: Bad camera configuration.
G22 Prodspec error. File prodspec.rsc is corrupt, or has missing entries.
G25 A measured value is outside spec from prodspec.rsc.
G27 Temperature sensor error. Likely Cause: I2C HW problem.
G29 Some operation hit on a limitation is SW.
G30 Failed to Ulis detector VSK/ GSK. Likely Cause: Detector HW problem.
G34 Wrong BB temperature during radiometric calibration verification.
G37 TBD. PROD_MAP_SLOW_PIX_ERR
G38 Error during focus calibration or failed autofocus operation.
G39 Error during fusion calibration( alignment ).
G42 Error detected on I2C bus.
G43 Error during NUC operation.
G44 Fail to aim camera or measure area on calibration BB during radiometric calibration. Check AimImage.
G45 Laser position measurement fail or missing. More info
G46 Failed to detect IR spot during fusion calibration. More info
G47 Failed to detect visual spot during fusion calibration. More info
G48 Missing lens data. More info
G49 Radiometric measurement error. Measure count not updated. Likely Cause: Bad camera configuration, SW bug.
G50 Unknown BB requested during radiometric calibration.
G51 Wrong production state. Likely Cause: Standard lens not calibrated when calibrating extra lens.
G52 TBD. PROD_BAD_POWER_DOWN_ERR
G53 IR to visual camera rotation. More info
G54 Camera not ready for KaKa run. Likely Cause: Bal calibration missing. No storage card. Bad camera configuration. Bad calib.rsc file. Missing lens
G55 TBD. PROD_CONFIGURATION_ERR
G56 Visual cameras align error. More info
G58 Failed to find map center offset.
G59 Failed to find GFID when checking detector dynamic range. Likely Cause: HW problem
G61 Camera not ready for BAL. Likely Cause: No storage card. Bad camera configuration. Bad calib.rsc file.
G62 Failed to find center of collimator target.
G63 Focus calibration failed C3 out of spec..
G64 Unable to read image of bar target.
G65 Unable to freeze image of bar target.
G66 Failed to locate bat target. Check image mrtdbad.fff
G67 Dual cameras not calibrated.
G69 Result of dual cameras calibration outside spec.
G70 Fusion calibration outside spec
G72 Unable to "fusion" at close distance.
G74 Bad tdrift image( too many saturated pixels )
G75 Focus max/min position not found