Recently I got a damaged FLIR TAU320. And working 9Hz VUE 336 and a 30/60Hz VUE 336 PRO. I made some measurements on the circuit and timing of the detector. The relevant information is included in the attachment.
Notes:
1.The circuit of the detector is drawn on FLIR TAU320. This camera uses FLIR ISC0601B detector, which is also used in products such as NV2, E40, A320. The effective resolution of the detector may be 324*256px and 25um pixel pitch.
2.The detector timing is measured on FLIR VUE, and the camera's detector "maybe" ISC0901. This detector is used in NV3, E4, QUARK, etc. in CSP package. However, the FLIR TAU2 336 (equivalent to VUE 336) of the ceramic vacuum packaged detector, the die inside the detector is very similar to the CSP packaged ISC0901. I used another thermal imager to observe the die inside the ceramic vacuum package through the silicon window. The PAD on the die and the surrounding test points are the same as the detector in the CSP package. I don’t have a CSP packaged detector, so I can’t know if the internal structure is the same. As an assumption here, it is believed that FLIR uses the same die to produce two different packaged detectors.
3. FLIR TAU2 (VUE) and TAU320 have the same detector peripheral circuit and the same pin definition. But the flexible circuit board connecting the detector to the main board is slightly different. There is an additional thermistor socket on the TAU2 flexible circuit board, which is used to measure the temperature of the lens mount.
4.I measured some detector waveforms and attached them to google drive. These waveforms include the detector's readout timing, power-on timing, power-down timing, and camera soft reset timing. I use TLA5202B to measure the timing of the detector, regardless of whether the 9Hz or 30/60Hz camera detector works at the same clock speed. You need to install TEKTRONIX TLA v6.1 software to open these files. The software can be downloaded for free and can run without TLA hardware. At present, I do not fully understand the functions of these signals. Some of the functions have been analyzed and I marked them on the schematic.
https://drive.google.com/drive/folders/1Y86siS_A-smqSURbPRdMCSJpgdKCVOX4?usp=sharing Have fun. BI4LBK 73