Finetuned,
Earlier generations of FLIR thermal camera contained Service Menu’s that offered both dead pixel map and calibration table updating. Those service menus were protected by a password that is now well known. FLIR removed the Service Menu from cameras Circa 2014, after the Ex series upgrade became known. They did not want the Service Menu to be accessible by users so removed it rather than improved its protection. When we look at the FLIR C5 and its kin, what we have is a LEPTON core with a host electronics package. The Radiometric version of the LEPTON is calibrated at factory and defects in calibration are normally detected at that point. The Lepton does not offer any option for a user, or 3rd party manufacturer to update its calibration table. As with many thermal imaging cameras, the accuracy of measurement remains pretty close to factory unless a component fails or ages over time. Most 3rd party Calibration checks are just that, a check, not a re-calibration as they do not hold the required FLIR Factory calibration utilities.
The LEPTON radiometric core can suffer from issues with the ambient temperature thermistor that is attached to the front of it and connected to the electronics via conductive paint. Damage or corrosion to the conductive paint can cause wildly inaccurate measurements. The Lepton is not really intended to be a core that will be re-calibrated… if the core fails measurement calibration checks, a new Lepton will be fitted into the host. The host electronics, in this case, have no involvement in calibration errors.
LEPTON is an inexpensive core. It can suffer quality issues.
Fraser