Author Topic: Is it possible to trigger from the math channel on the Rigol DS1054Z?  (Read 2236 times)
Offline ChumanistaTopic starter
Hi!
I have a really noisy input that needs a low-pass filter before it becomes usable (a lot of noise).
Because of external rules I cannot use any external parts.
Is there any way to do it? I googled but couldn't find anything.
Online JPortici
Re: Is it possible to trigger from the math channel on the Rigol DS1054Z?
« Reply #1 on: November 06, 2016, 02:15:51 pm »
you wish! you can however apply a filter to the trigger itself. in your case, select HF rejection on the trigger coupling
« Reply #2 on: November 06, 2016, 02:17:57 pm »
I do wish indeed. The math option is so nice. I don't need a lot of resolution, so I could just record 5s to memory and filter it in Excel, but triggering of the already working function would be SO NICE. Why is it impossible? Calculation limitations?
Online JPortici
« Reply #3 on: November 06, 2016, 02:26:47 pm »
math is calculated from on screen data, not from acquisition memory (even on much better scopes, unfortunately)
Online JPortici
« Reply #4 on: November 06, 2016, 02:28:45 pm »
depending on your actual signal you can work out different ways to produce a more stable trigger. though i'm used working only with edge and pulse/window and by adjusting the holdoff
