Strange that Agilent doesn't specify it. Do they specify to only use the correct Agilent part, or specific parts from other manufacturers like Bussmann?
You could check check the actual fuse, it's often printed on them. For example, on a 440mA fuse I have on hand (DM-44/100) I think it's the same as the one you link to), IR: 10KA in fine print. The data sheet from the second Farnell link actually mentions the IR, 20kA for the 11A fuse. The datasheet for the first link is useless, but if you go to
the Fluke fuse selection guide (since they specify a Fluke part number), and select for example the 87-V (uses those same fuses I think), you'll see that Fluke p/n 943121 is 440mA/10kA/1000V.
Page 20 of the user/service manual (from the Agilent website) mentions that both fuses should be 30kA, so the fuses you link to are inferior to the original fuses. Although they are used by Fluke, I wouldn't trust the meter to meet its CAT III/IV specs after replacing a safety device with an inferior part (maybe their shunt has a lower impedance, or Fluke took other measures to limit the max. current). Page 103 specifies specific Agilent parts, although it wouldn't hurt to give the full specifications again. The specified Agilent parts will no doubt meet the specs, not sure if there are any standard Bussmann parts that meet it. The Agilent price for both fuses is $4.35, so it shouldn't be that expensive.