There is not even a commercial ESD module to protect electrodes? Since there is also no stand alone commercial Isolated module that works with all amplifiers. Why don't you build them and sell them as finished products? I won't try building them in breadboard.
It's because to be able to sell for medical uses, you need that certification, and it is impossible to certify a device to work with all amplifiers or all electrodes.
I want to know if it is better to have a separate power supply at the ISO-Z instead of getting it from the main amplifier.
The amplifier requires so little power, it does not actually matter. It makes things more straightforward if the power is supplied by the main amplifier.
Actually, I don't like the design of these devices much. Above, I tried to explain that I'd prefer the entire sampling front-end to be in the preamplifier: only
digital information would pass. Even with discrete components, say AD7766 and a precision amplifier frontend with high-impedance inputs, you could get 16-bit data from an isolated sampling frontend that only involves very low voltages and currents, with double or triple safety features.
So the ISO-Z are not useful for anything above 100Hz. I wonder if a component got damaged or it is really designed that way.
From
this article:
"The heartbeat has frequency from 0.1 Hz to 150 Hz. The electrical signal which is resulted from the heartbeat has amplitude of 100μV - 4 mV."So, could very well be designed that way.
Why do you say LF411CP is the initial buffer of the signal? It is the last buffer? Here is the initial trace I did.
Because I was wrong; I looked at the ISO-Z upside-down (as if signal came in at the top, and came out of the jack near the power connectors).
if the circuit is really safe. The product is not really certified.
In that case, I would use something else. It looks fine in theory, but with the issues you're seeing, a self-made PCB could be made better and inherently safer (by limiting the power available to the isolated side, and with low-leakage ESD diodes).
When you push calibration which suppresses the input. It seems you only disconnect the input while running the 741CN which always have the 0.5mV 10 Hz signal on?
Two things happen: the button pulls the sensor input to ground via a resistor, and connect the clipped 10Hz signal to the output.
The circuit looks simple enough. Can it be run in Pspice or other simulators and can it produce the output and frequency? In the software, can one trace the source of the mV level high frequency noise by maybe removing each capacitor at a time? I want to know if the noise is part of the circuit or due to broken capacitor.
I don't know, and knowing the device isn't really certified, I wouldn't bother: the design is old, and we have better components now.
A fully digital front-end wouldn't need a calibration signal generation. If needed or desired, a completely separate signal synthesizer (that could play back recorded signals for educational purposes) would be much better; and that's just a microcontroller-driven DAC with attenuation and opamp buffering.
your theory of how it works
I thought the red connector on the bottom of the board image is where the amplified signal comes out, and the red six-pin Molex connector at the top was the input where the electrodes are connected to. It's obviously the other way around, the Molex connects to the bioamplifier, and the electrodes to the connectors at the bottom.
So, the electrode 0V reference is the white connector at the bottom (black wire), and the red and black are the differential signals. Based on the silkscreen, the button connection are
4 1 Normally 1 is connected to 3, and 4 connected to 6
5 2 6 3 When pressed, 2 is connected to 3, and 5 connected to 6
If so, the electrode positive and negative go to pins 1 and 4 of the switch, and from pins 3 and 6 through the two blue resistors to the inverting and non-inverting inputs of the INA114AP amplifier. (Note that the positive and negative get swapped here, and you should normally have excellent continuity with no noise between the right side of the two bottom blue resistors and the electrode input jacks, when the button is not pressed; and none when the button is pressed.)
INA114AP is configured as an inverting amplifier with a gain of 10. Its output goes to the isolation amplifier ISO122P, crossing the isolation barrier between the electrode and the bioamplifier. ISO122P is in unity gain configuration. Its output goes to the inverting input of the final LF411CP buffer amplifier, also in unity gain configuration. The non-inverting input to LF411CP is from the blue DC offset potentiometer.
That is the full electrode signal path. I would suspect the noise is in the tracks and the switch, before the first resistors.
When the button is pressed, the electrode positive and negative are disconnected, and instead of the positive electrode, the inverting input to INA114AP is from the third pin of RP-1. The electrode negative connects to the solder point just above the push button contacts, and from there to the resistor closest to the green wire that connects it to the electrode positive: essentially, when you push the button (even without any power), you should see that resistor between the electrode positive and negative inputs, and that only.
L7805CP regulates the positive supply from the bioamplifier to a 5V used by both the LED (the light gray 680 Ω resistor top center is its current-limiting resistor) and the DCP010512DBP DC-DC converter. The copper area on the other side is the common of its isolated output, and the component-side tracks are the derived +12V and -12V (unregulated) supply rails. These power the LM741CN in a twin-T oscillator configuration (as well as the INA114AP and the isolated side of ISO122P). I do believe you should see a large sine wave constantly on LM741CN pin 6 (top row, second from right). It is the wide track on the solder side, and although I can't see it fully, I think it comes up to the component side as the left one of the wide black tracks in the middle of the board. That is then connected to the diodes and an RC low-pass filter (two light brown resistors and two capacitors in the center) for each polarity. I think the second diode from the left is connected to one of the blue resistors at the bottom, which pass the electrode positive and negative signals when the button is not being pressed; and this is where the calibration signal gets injected.
I don't see exactly where the oscillator signal gets attenuated to 500 µVpp = 0.5 mVpp = 0.0005 Vpp.
If the 10Hz couples to the output signal always, then I'd look at the two bottom blue resistors and corrosion.
For noise in the electrode signal, I'd check the tracks and the button switch before the blue resistors, and maybe replace the INA114AP (10€ at Mouser, in stock) and LF411CP (1.6€ at Mouser, in stock) opamps, as they're already socketed. Also clean and check the sockets for corrosion, too. I don't know if it is the picture angle, but the LF411CP does not look like it is fully seated to me.